I've just bought my first oscilloscope, and after some initial tests I would like to ask you some question.
I've tested build-in generator, and I noticed that when I use BNC cables, after some frequency, amplitude starts dropping a bit (no matter if I use 50ohm terminator or not), as you can see on attached images when I use BNC cables, at 25MHz amplitude drops around 5-7%.
I'm aware of transmission line theory, but thought that with 1m cable at 25MHz with 50 Ohm termination there shouldn't be any drops.
I would like to ask you if it is normal or I need to buy better cables/connectors.
BNC cable, connector and terminator are from Amphenol, so there are not the cheapest.
« Last Edit: January 27, 2023, 12:45:52 am by gawlas »
Offline JeremyC
Re: Rigol MSO5000 Generator <--> Scope Amplitude Drop
« Reply #1 on: January 27, 2023, 01:13:53 am »
When you make the test with 50 ohm terminator, you should change the generator output from "HighZ" to "50 ohm".
« Reply #2 on: January 27, 2023, 01:16:56 am »
As I understand it only changes what is displayed on screen (divide actual amplitude by 2), so doesn't really matter.
I know that, when using 50ohm terminator amplitude on scope should be half of generated (1V), I'm asking about difference (drop) between 1kHz vs 25MHz
« Last Edit: January 27, 2023, 01:20:47 am by gawlas »
Offline JeremyC
« Reply #3 on: January 27, 2023, 01:25:25 am »
It may matter, but I'm not sure. It doesn't hurt to change it and repeat the test. BTW, how it looks when frequency is 20MHz? 25MHz it's the max.
« Last Edit: January 27, 2023, 01:42:24 am by JeremyC »
« Reply #4 on: January 27, 2023, 01:30:18 am »
I've tried with '50Ohm' option in generator and results are the same.
Here some more screens from tests with BNC cables - 10MHz an 20Mhz, I didn't attach them before as there is limit per post.
As you can see amplitude is dropping with frequency:
BNC 0.5 meter (direct):
1kHz - 2.03 V
1MHz - 1.97 V
10Mhz - 1.97 V
20Mhz - 1.95 V
25Mhz - 1.93 V
BNC 0.5 meter (50 Ohm terminated):
1kHz - 0.99 V
1MHz - 0.98 V
10Mhz - 0.96 V
20Mhz - 0.94 V
25Mhz - 0.92 V
« Last Edit: January 27, 2023, 01:34:51 am by gawlas »
Offline thm_w
  • Non-expert
« Reply #5 on: January 27, 2023, 01:52:29 am »
10.1 dBm to 9.69 dBm thats not much of a drop. https://www.aetechron.com/calculator-Volt2dB.shtml
You can find the spec in the datasheet:
Frequency Range 100 mHz to 25 MHz
Flatness ±0.5 dB (relative to 1 kHz)
So its within spec.
If you need something better, get a dedicated function generator. The built in AWG is kind of a toy (although may be useful for certain automation applications).
Offline rstofer
« Reply #6 on: January 27, 2023, 01:54:39 am »
Almost everything rolls off with frequency.  A 100 MHz scope will likely show the trace -3dB down at 100 MHz.  You can expect yours to do this very thing.  At -3dB down, the voltage is 70.7% of the low frequency value.
The question is: How much should the scope roll off at 25 MHz.  See Summary point 1 and perhaps go back to the equations.
https://edadocs.software.keysight.com/kkbopen/how-does-gaussian-rolloff-factor-into-oscilloscope-bandwidth-selection-criteria-part-1-614517535.html
Then there is the issue of the AWG rolling off at 100% of bandwidth...
https://beyondmeasure.rigoltech.com/acton/attachment/1579/f-0907/1/-/-/-/-/MSO5000_datasheet.pdf
Offline xrunner
  • hp>Agilent>Keysight>???
« Reply #7 on: January 27, 2023, 01:56:29 am »
The manual looks like the waveform generator output is +/- (2% of setting + 1 mV) (freq. = 1 kHz)
So 2V setting would give +/- .041 V accuracy, from 1.959 to 2.041 V . Does that make more sense?
« Reply #8 on: January 27, 2023, 08:09:08 am »
Thanks for answers, I think I understand this a little bit better now, but have two more questions:
- 10x probe shows almost exactly 2V over entire range (drop only with BNC coaxial cable), so I don't think generator accurracy/flatness is an issue, am I right?
- If amplitude drops because I'm getting closer to my scope bandwidth, then why probe shows correct values? Does this 5 times rule does not apply to probe?
Offline xrunner
« Reply #9 on: January 27, 2023, 01:16:27 pm »
This isn't the same kit you are using, but for comparison I used my most "inferior" scope and a brand new scope to check a 2Vpp sine wave using a 1 m BNC cable. The signal came from a Rigol DG1032Z.
The poorest scope I have is the TDS-210 (60 MHz BW) and the best is a new Siglent SDS2202X-E opened to 400 MHz BW.
I checked the signal on both and got these Vpp measurements -
TDS-210
1 kHz   2.06
25 MHz   1.96
Siglent SDS2202X-E
1 kHz   2.06
25 MHz   2.00
They seem reasonable to me.
Offline thm_w
  • Non-expert
« Reply #10 on: January 27, 2023, 10:46:25 pm »
Look up the input impedance of the oscilloscope probe vs the oscilloscope itself.
You are essentially putting a large resistor in series with the input when you use the oscilloscope probe, so the loading will be less.