Author Topic: Got to be woth £20 uk  (Read 119525 times)

0 Members and 5 Guests are viewing this topic.

Offline staxquad

  • Frequent Contributor
  • **
  • Posts: 417
  • Country: ca
  • Eye Candy
Re: Got to be woth £20 uk
« Reply #75 on: September 19, 2013, 09:33:02 pm »
mine died

Tried to turn it on, nothing.
Plug in 5v, and all 4 leds rapidly flashing dimly with no Volt Ref output, unplug.
Voltage low on the battery so I disconnect and connected another 3.7v Li-Po
All 4 leds still flashing dimly with no output.
Accidentally short the battery leads and the unit stops flashing dimly.
Turn it on and it works.
Checked the values and same as before, as the spec sheet.
Charged the battery externally before reattaching the leads.
Had to short the leads again to reset from all leads flashing dimly.
Volt Ref works again with same reattached battery.
Battery doesn't charge on the Volt Ref, will have to do it externally.
I always made sure I am sure to turn the Volt Ref off before putting it away, but maybe I missed the last time and it caused the failure.

Same day my Sony NEX 5N digital camera goes on the fritz.
Turn it on, no image, just the outlying info, stuck on 30" and ISO 1600, shutter or movie button will not activate.
Screen goes out completely most of the time.
Have to turn off the camera, pull out the battery, put back in the battery, turn on the camera,  plug in the USB just to get to the Menu screen, change things, but it reverts, it's a dead end.
Got the camera with 1.0 firmware and updated to 1.1 last year.  Can't reload the same firmware, but luckily Sony had 1.2 on their website, so the plan was to try to update the firmware to try to unlock the freeze.  Finally get USB to be in Mass Storage mode by uploading a few photos from the camera, it stayed in USB Mass Storage mode (wouldn't stay when I changed it in the menu and it won't allow a connection without), so clicked the firmware update program, clicked safely remove hardware for the camera in Windows so it could connect, and uploaded the new 1.2 firmware and the camera now works.

so what caused the 2 devices to lock out at the same time?

Aliens?
NSA, RCMP probing the house with electronic waves? (who knows what they are looking for)
Smart Meter just 10 feet away on the other side of the wall where the camera and the volt ref were parked?
coincidence?'



« Last Edit: September 20, 2013, 05:24:38 pm by staxquad »
"TEPCO Fukushima you long time"
You say Vegemite, I say Yosemite. (Ve-gem-mit-tee, Yo-zey-might)  
"For starters : you're Canadian...."
 

Offline KJDS

  • Super Contributor
  • ***
  • Posts: 2442
  • Country: gb
    • my website holding page
Re: Got to be woth £20 uk
« Reply #76 on: September 20, 2013, 07:12:37 am »
Mine arrived a couple of days ago. Need to find the right connector to charge them.

Offline M0BSW

  • Frequent Contributor
  • **
  • Posts: 673
  • Country: 00
  • Left this site 2013, they will not delete it ????
Re: Got to be woth £20 uk
« Reply #77 on: September 20, 2013, 10:33:19 am »
Got mine today, straight down to my friends at the calibration centre it checks out 100% OK with the supplied chart  :-+, so ,like a kid with a new toy, I tried it on my multimeter, remember nothing fancy here uni T61E out by 2 to 3 counts everywhere, Digitek again 2 to 3 everywhere ,  Ideal  Alphatek  1 sometimes flickering to 2 counts out lower end only, but the absolute STAR is 30 Years plus old Farnell TMS, English made locally, BANG ON EVERYWHERE go ENGLAND :-+ :-+ :-+ :-+ :-+ :-+
no one would or will tell me how to delete this account
 

Offline KJDS

  • Super Contributor
  • ***
  • Posts: 2442
  • Country: gb
    • my website holding page
Re: Got to be woth £20 uk
« Reply #78 on: September 20, 2013, 11:01:44 am »
I've got four. Just put the first one on charge. Once they've all had a while to charge up I'll compare them to a few Keithley 2015THD meters and post up the results, as well as an AVO 8 and some Fluke 25s and Wavetek 2015, 2020 and 2030. Will be interesting to see what they all read.

Offline M0BSW

  • Frequent Contributor
  • **
  • Posts: 673
  • Country: 00
  • Left this site 2013, they will not delete it ????
Re: Got to be woth £20 uk
« Reply #79 on: September 20, 2013, 11:47:11 am »
I've got four. Just put the first one on charge. Once they've all had a while to charge up I'll compare them to a few Keithley 2015THD meters and post up the results, as well as an AVO 8 and some Fluke 25s and Wavetek 2015, 2020 and 2030. Will be interesting to see what they all read.
I'm really looking forward to the AVO results, as I'm looking at one of these on ebay
no one would or will tell me how to delete this account
 

Offline KJDS

  • Super Contributor
  • ***
  • Posts: 2442
  • Country: gb
    • my website holding page
Re: Got to be woth £20 uk
« Reply #80 on: September 20, 2013, 02:28:09 pm »
ok, so here is what I found. I numbered the references 1 through 4.

With ref 4,

Wavetek 2020   wavetek 2015  wavetek 2020   Wavetek 2015  Fluke 25  Fluke 25         AVO 8   IsoTech 205   Isotech 205
2.500               2.496               2.500                   2.500          2.499         2.499       2.498        2.506          2.498
5.004               5.00                  5.002                  5.00            5.00           5.00         4.95         5.01             4.99
7.504                7.49                 7.500                   7.50           7.49           7.50         7.45          7.52            7.49
10.007             9.99                 10.002                   10.00         9.99          10.00        9.98          10.03          9.99

and on three Keithley 2015THDs          HP3478A                   values on the certificate
2.49942        2.49940       2.49939       2.49940                     2.49946
5.00208        5.00204       5.00202        5.0020                     5.00211
7.50057        7.50050       7.50049        7.5004                     7.500591
10.00266      10.00257      10.00255     10.0025                     10.00266


Reference 2 was faulty, providing ever changing readings.

Reference 1 on the three keithleys and the HP3478A

 on three Keithley 2015THDs          HP3478A                   values on the certificate
2.49965       2.49964       2.49963       2.49965                     2.49971
5.00213        5.00210       5.00208        5.0021                     5.00218
7.50225       7.50220       7.50218        7.5022                     7.50230
10.00427      10.00420      10.00417     10.0042                     10.00430


Ref 3 on three Keithley 2015THDs          HP3478A                   values on the certificate
2.49861        2.49861       2.49860       2.49861                     2.49868
5.00140        5.00130       5.00136        5.0013                     5.00137
7.49879        7.49873       7.49871        7.4987                     7.49866
10.00113      10.00105      10.00102     10.0010                     10.00089


I'm a little disappointed that number two is unstable on all ranges, but nowhere near as if I'd only bought one or two.

I'm happy that all the meters are giving reasonable readings. I only let the bench meters warm up for half an hour or so. The handheld meters were all tested straight after turning on, so typical of normal hobby use.

The three references that do work give results consistent with their certificates. I'll keep measuring a few over time to see how they drift.









Offline Fraser

  • Super Contributor
  • ***
  • Posts: 15232
  • Country: gb
Re: Got to be woth £20 uk
« Reply #81 on: September 20, 2013, 04:19:48 pm »
From what we have seen on this thread, the certificate that comes with the units is quite accurate. With that in mind your Reference number 2 was likely working when made. You may be lucky and find a dry joint on the PCB that has been disturbed by the journey to you.

I have four of these now and all are holding their 'calibration' so far.
« Last Edit: September 21, 2013, 09:04:00 am by Aurora »
If I have helped you please consider a donation : https://gofund.me/c86b0a2c
 

Offline staxquad

  • Frequent Contributor
  • **
  • Posts: 417
  • Country: ca
  • Eye Candy
Re: Got to be woth £20 uk
« Reply #82 on: September 20, 2013, 09:44:48 pm »
From what we have seen on this thread, the certificate that comes with the units is quite accurate. With that in mind your Reference number 2 was likely working when made. You may be luck and find a dry joint on the PCB that has been disturbed by the journey to you.

I have four of these now and all are holding their 'calibration' so far.

Printed on the certificate is the device used for the measurements, an HP 34401A (6 1/2 digit).  So if the 34401A is within calibration, the numbers should stack up with other devices of similar precision within calibration validating the values, for the stability period of the Volt Ref.
"TEPCO Fukushima you long time"
You say Vegemite, I say Yosemite. (Ve-gem-mit-tee, Yo-zey-might)  
"For starters : you're Canadian...."
 

Offline Noldir

  • Newbie
  • Posts: 8
  • Country: nl
Re: Got to be woth £20 uk
« Reply #83 on: October 04, 2013, 12:23:59 pm »
Got mine as well, the small sheets says:

2.500 = 2.69932
5.000 = 5.00118
7.500 = 7.69952
10.000 = 10.00096

Aren't the 2.5 and 7.5 values a bit...off? 0.2 is quite a bit of deviation..
« Last Edit: October 04, 2013, 12:25:40 pm by Noldir »
 

Offline mianchen

  • Regular Contributor
  • *
  • Posts: 225
  • Country: gb
Re: Re: Got to be woth £20 uk
« Reply #84 on: October 04, 2013, 12:41:18 pm »
Got mine as well, the small sheets says:

2.500 = 2.69932
5.000 = 5.00118
7.500 = 7.69952
10.000 = 10.00096

Aren't the 2.5 and 7.5 values a bit...off? 0.2 is quite a bit of deviation..

are u sure the 6s are 6s not 4s?

Sent from Mars

 

Offline Fraser

  • Super Contributor
  • ***
  • Posts: 15232
  • Country: gb
Re: Got to be woth £20 uk
« Reply #85 on: October 04, 2013, 03:08:55 pm »
I concur.

My certificate has some hard to read numbers, 4's being very close to 6 shaped !

The 2.5V sets the calibration for the whole unit so that should be accurate if the derivatives are.

Fraser
If I have helped you please consider a donation : https://gofund.me/c86b0a2c
 

Offline Noldir

  • Newbie
  • Posts: 8
  • Country: nl
Re: Got to be woth £20 uk
« Reply #86 on: October 04, 2013, 05:16:55 pm »
Quite positive they are sixes.... (as in 99% positive unless someone has a awfully weird 4)
 

Offline Noldir

  • Newbie
  • Posts: 8
  • Country: nl
Re: Got to be woth £20 uk
« Reply #87 on: October 04, 2013, 05:19:47 pm »
Here's a piccie :)
 

Offline pmcouto

  • Supporter
  • ****
  • Posts: 103
  • Country: pt
Re: Got to be woth £20 uk
« Reply #88 on: October 04, 2013, 05:29:13 pm »
Don't be so sure - That must be a Chinese 4  :-DD
 

Offline Kostas

  • Regular Contributor
  • *
  • Posts: 71
  • Country: gr
Re: Got to be woth £20 uk
« Reply #89 on: October 04, 2013, 05:35:59 pm »
Here's a piccie :)

Look at the header, the measurements were done at 26°C. If you look carefully, you'll see that this six is quite different than  the other "sixes". They have the ending line going downwards. I'd say they are fours.
 

Offline KJDS

  • Super Contributor
  • ***
  • Posts: 2442
  • Country: gb
    • my website holding page
Re: Got to be woth £20 uk
« Reply #90 on: October 04, 2013, 06:01:21 pm »
Here's a piccie :)

Look at the header, the measurements were done at 26°C. If you look carefully, you'll see that this six is quite different than  the other "sixes". They have the ending line going downwards. I'd say they are fours.

I just looked at three of mine and have to agree. Two of them have the 4s very clearly as 4s. The third has the 4s more like those in the image

Offline PA0PBZ

  • Super Contributor
  • ***
  • Posts: 5512
  • Country: nl
Re: Got to be woth £20 uk
« Reply #91 on: October 04, 2013, 07:19:37 pm »
Looks like my label was written by the same guy:
Keyboard error: Press F1 to continue.
 

Offline Fraser

  • Super Contributor
  • ***
  • Posts: 15232
  • Country: gb
Re: Got to be woth £20 uk
« Reply #92 on: October 04, 2013, 08:03:58 pm »
They are the number 4. As I stated, the 2.5V ref sets the others so it can't be a number 6 if they are close to right. I agree that they look like a 6 but this chap is writing in a numerical language not native to him, so could be forgiven for poor style, and I suspect he is doing the tests 'at speed' to keep up with demand.

My 4 units number 4's are a range of good, to the same as yours, purely due to poor style.
If I have helped you please consider a donation : https://gofund.me/c86b0a2c
 

Offline Noldir

  • Newbie
  • Posts: 8
  • Country: nl
Re: Got to be woth £20 uk
« Reply #93 on: October 04, 2013, 10:12:26 pm »
They are the number 4. As I stated, the 2.5V ref sets the others so it can't be a number 6 if they are close to right. I agree that they look like a 6 but this chap is writing in a numerical language not native to him, so could be forgiven for poor style, and I suspect he is doing the tests 'at speed' to keep up with demand.

My 4 units number 4's are a range of good, to the same as yours, purely due to poor style.

Ah yes, I'd almost forgot that he's not quite used to writing in western numerals  :-[

And of course I'm quite forgiving of his style, he just put me on the wrong footing by his otherwise good handwriting :)
 

Offline Noldir

  • Newbie
  • Posts: 8
  • Country: nl
Re: Got to be woth £20 uk
« Reply #94 on: October 04, 2013, 10:13:28 pm »
Looks like my label was written by the same guy:

Yup, looks the same to me! And, looking at yours I can definitely see how that "6" could be a 4 :)
 

Offline staxquad

  • Frequent Contributor
  • **
  • Posts: 417
  • Country: ca
  • Eye Candy
Re: Got to be woth £20 uk
« Reply #95 on: October 04, 2013, 11:44:11 pm »
Looks like my label was written by the same guy:


For 10.000V he has 10.00094, with the 4 looking more like a 4, but still showing that hook he uses to drop down to finish off the 4.   He uses one stroke to make a four while most people use 2.
« Last Edit: October 04, 2013, 11:46:34 pm by staxquad »
"TEPCO Fukushima you long time"
You say Vegemite, I say Yosemite. (Ve-gem-mit-tee, Yo-zey-might)  
"For starters : you're Canadian...."
 

Offline iloveelectronics

  • Frequent Contributor
  • **
  • Posts: 940
  • Country: hk
Re: Got to be woth £20 uk
« Reply #96 on: October 05, 2013, 07:18:51 am »
Being Chinese I can say with 100% certaint that it is a very typical “4” written by a Chinese person.

I have travelled to many diiferent places and I have always found it very interesting the ways different people in different parts of the world write their numbers. The “1” written in Europe, for example, is very interesting. It looks like a upside down “V” to me :)
My email address: franky @ 99centHobbies . com
My eBay store: http://stores.ebay.com/99centhobbies
 


Offline iloveelectronics

  • Frequent Contributor
  • **
  • Posts: 940
  • Country: hk
Re: Got to be woth £20 uk
« Reply #98 on: October 25, 2013, 03:31:32 am »
Their pricing is nuts.... I know because I have actually ordered a batch from the producer of these so I know the costs. Plus they are still trying to sneak the lithium batteries through HK Post, practically slowing the entire system down (due to screening and rejecting at least some of their packages) for everybody else :(
My email address: franky @ 99centHobbies . com
My eBay store: http://stores.ebay.com/99centhobbies
 

Offline mianchen

  • Regular Contributor
  • *
  • Posts: 225
  • Country: gb
Re: Got to be woth £20 uk
« Reply #99 on: October 26, 2013, 04:22:45 pm »
Got mine today:

   Volt Ref    UT-61E    Fluke 87III(1) Fluke 87III(2) HP 3468A
  2.500     2.49849    2.496 2.497 2.498  2.49817
  5.000     5.00233    4.9995.0015.001  5.0018
  7.500     7.49884    7.4967.4977.497  7.4980
10.000  10.00221    9.99910.0009.99910.0012


My little uncalibrated DIY AD588JQ voltage ref gives "better" results  :-DMM

   UT-61E    Fluke 87III(1) Fluke 87III(2) HP 3468A
    5.000      4.998  4.999 5.000  5.0003
  10.000      9.99810.0009.99910.0007


Which one should I trust? I don't need high precision measurement, but it's kind of nice to know which reference is more 'real'
« Last Edit: October 26, 2013, 04:34:13 pm by mianchen »
 


Share me

Digg  Facebook  SlashDot  Delicious  Technorati  Twitter  Google  Yahoo
Smf

 

- - -
-->