@18:55 hard pull-down and wimpy pull-up is how TTL is
Offline free_electron
    • SiliconValleyGarage
« Reply #2 on: December 15, 2025, 04:58:31 am »
TTL doesn't require totem pole. The inputs of a TTL gate are the emitters of a transistor with its base tied to vcc via a 400-ish ohm resistor.
When you are sending a logic 0 into a TTL gate you are actually pulling current OUT of the input.
That's where the concept of fan-in and fan-out comes from. You need to know how many inputs you will be driving LOW because you are pulling current OUT of them.
Offline MarginallyStable
« Reply #3 on: December 16, 2025, 04:41:12 pm »
Whats up with the power switch? (Down for on?)
« Reply #4 on: December 17, 2025, 02:02:02 am »
Correct, it doesn't require it, but that's how it's implemented, as a totem pole with the upper resistor dependent upon the series (7400,LS,S)
Online Kleinstein
« Reply #5 on: December 17, 2025, 09:20:58 am »
For an educational instrument the simple resistor + transitor to ground solution makes perfectly sense as it is short circuit proof (at least towards ground). The rise time is not great and there is some delay from the transistor driven to saturation. For the educational market this is not a problem, as it is not about highest performance.