Dave, very interesting and informative video!
Some remarks, though:
FLUKE made these glass covered resistor networks for their high grade calibrators and references in first place.. You'll find these in the 732B, the 5720A and also in the 8508A 8 1/2 digit "Reference Multimeter".
These networks give these instruments their superior stability.
The reason, it's an open, glass covered design, is simple.. The LASER trimming is done after the sealing of the element, through the glass plate, and in some cases obviously in situ, in the finalized instrument at the final testing place.
Imagine, you use that network for final trimming of the 732B output, i.e. absolute 10.00000V and 1.00000V.
The High Voltage divider should have exactly 9.9M / 100k .. I assume that there was some protection circuitry inside the instrument, giving these few kOhm instead.
The reason, Fluke manufactured this special network, has two reasons or advantages over usual design ocompany Caddock, where they bond 2 distinct resistors back-to-back.
First, both resistors have identical T.C., as they have been sputtered from the same alloy target (NiCr or TaN), so the T.C. tracking is already near perfect.
But they also overcome the power difference T.C. effect in a usual HV divider..
The line width of each resistor seems to be the same everywhere.
The length of the 9.9M resistor path should be exactly 99 times the 100k resistor path.
So the power dissipation on each resistor element is equal.
That means, that the 9.9 M and the 100k resistor will heat up to the exactly same temperature when 1000V were applied. Additionally, the high thermal conductive ceramic substrate equalizes any temperature gradients, but on a slower time constant.
That means, that the 100:1 ratio will be constant under all circumstances, to < 1ppm..
Also dynamically, if you want to digitize an alternating 1000V level on a short time scale.
Most other 6 1/2 ... 8 1/2 DMM suffer from that power dissipation effect.
They need a quadratic compensation, by calibrating at 1000V, 500V, and maybe 100V, to compensate later numerically for this change in ratio over power.
The 3458A does not have this compensation, because maybe it is not feasible to calculate the quadratic compensation for 100kHz digitizing at these high voltages, but it therefore has a mediocre 12ppm additional error at 1000V.
So, this Fluke Thin Film network really is absolutely perfect.
Frank