Author Topic: FS UK: Agilent 3458A 8.5 digit DMM  (Read 9139 times)

0 Members and 1 Guest are viewing this topic.

Offline voltsandjoltsTopic starter

  • Supporter
  • ****
  • Posts: 3532
  • Country: gb
FS UK: Agilent 3458A 8.5 digit DMM
« on: October 21, 2023, 01:23:49 pm »
This is not my auction, I have no connection with the seller, just pointing this out for the volt nuts.

Starting bid £2350
https://www.ebay.co.uk/itm/225827704316

Quote
Agilent 3458A 8.5 Digit Digital Multimeter
Instrument passes self test & performs ACAL correctly.
It has not been calibrated for many years so no guarantee of the accuracy or stability.
1 volt applied from standard that is known to be +2ppm, 1 mV applied via divider
no options fitted

I think somebody on this forum said (half-joking) the only 3458A's on ebay are the drifty ones no one wants, so buyer beware.
« Last Edit: October 21, 2023, 01:26:02 pm by voltsandjolts »
 

Offline coromonadalix

  • Super Contributor
  • ***
  • Posts: 8505
  • Country: ca
Re: FS UK: Agilent 3458A 8.5 digit DMM
« Reply #1 on: October 22, 2023, 01:10:45 am »
« Last Edit: October 22, 2023, 01:49:07 am by coromonadalix »
 

Offline alm

  • Super Contributor
  • ***
  • Posts: 2903
  • Country: 00
Re: FS UK: Agilent 3458A 8.5 digit DMM
« Reply #2 on: October 22, 2023, 07:57:17 am »
5 units are an anecdote, not data. There is certainly a substantial risk that it's drifty, and the problem is that you probably won't find out until the return period is over. But I bought two early units from 1989-1991 and both are stable (n=2) (knock on wood).

Offline voltsandjoltsTopic starter

  • Supporter
  • ****
  • Posts: 3532
  • Country: gb
Re: FS UK: Agilent 3458A 8.5 digit DMM
« Reply #3 on: October 22, 2023, 04:31:30 pm »
Is £2500 a fair price for a drifty example?

Not that I want to get into this volt nuttery, but it does intrigue me.
It feels like purchasing this would almost be like adopting a child, although easier to sell on  ;D
It needs care and investment for the long term. Constantly powered in a cosy environment.
 

Offline coromonadalix

  • Super Contributor
  • ***
  • Posts: 8505
  • Country: ca
Re: FS UK: Agilent 3458A 8.5 digit DMM
« Reply #4 on: October 23, 2023, 12:24:38 am »
absolutely not,       like having an formula 1 car with a 4 cylinder at 80 hp

if the a3 board is drifty   it will cost lots of $ to get one,  or try to send you meter to keysight  ...  and lots $$$ to get it repaired and calibrated

and sadly ebay sellers may or may not know if their meter is problematic without the proper tests
« Last Edit: October 23, 2023, 04:11:46 pm by coromonadalix »
 

Offline voltsandjoltsTopic starter

  • Supporter
  • ****
  • Posts: 3532
  • Country: gb
Re: FS UK: Agilent 3458A 8.5 digit DMM
« Reply #5 on: October 23, 2023, 03:13:23 pm »
It's probably worth £2500 broken up for parts.
And if it's not drifty, it's a bargain price 3458A.
But I've got a cluttered bench already.
No. Nope. No way. Hmm, well....
 

Offline chilternview

  • Frequent Contributor
  • **
  • Posts: 499
  • Country: gb
Re: FS UK: Agilent 3458A 8.5 digit DMM
« Reply #6 on: October 23, 2023, 04:00:25 pm »
I'd be tempted but I already have two ;)

Don't forget that a Keysight cal is going to set the buyer back nearly £1k - and why buy one if you're not going to be sure it's in cal?
 

Offline alm

  • Super Contributor
  • ***
  • Posts: 2903
  • Country: 00
Re: FS UK: Agilent 3458A 8.5 digit DMM
« Reply #7 on: October 23, 2023, 04:11:50 pm »
Don't forget that a Keysight cal is going to set the buyer back nearly £1k - and why buy one if you're not going to be sure it's in cal?
Because of the excellent linearity? For metrology-grade measurements I use mine purely as ratio devices. I have better voltage and resistance standards: the 3458A's built-in DCV and resistance reference are pretty good but not amazing. So for me the 3458A is generally just measuring the ratio between the DUT and an external reference. And thanks to the magic of the extremely linear ADC and autocal, it can accurately measure those ratios without any external calibration. The exception is high frequency ACV, but those specs aren't so hot anyway. The only thing that would make it even better is if the front/rear inputs could be switched under remote control.

If you want to rely on its absolute values then you might want to send it out for calibration, but I don't consider that essential for my use.
 
The following users thanked this post: Dr. Frank

Offline chilternview

  • Frequent Contributor
  • **
  • Posts: 499
  • Country: gb
Re: FS UK: Agilent 3458A 8.5 digit DMM
« Reply #8 on: October 23, 2023, 04:19:46 pm »
If you want to rely on its absolute values then you might want to send it out for calibration, but I don't consider that essential for my use.

Agreed, if the user has better standards that they can trust that's certainly a valid use model.
 

Offline coromonadalix

  • Super Contributor
  • ***
  • Posts: 8505
  • Country: ca
Re: FS UK: Agilent 3458A 8.5 digit DMM
« Reply #9 on: October 23, 2023, 05:32:40 pm »
i think the 3458  just needed 2 things

a very stable reference voltage and an precise resistor value to calibrate it ???
 

Offline alm

  • Super Contributor
  • ***
  • Posts: 2903
  • Country: 00
Re: FS UK: Agilent 3458A 8.5 digit DMM
« Reply #10 on: October 23, 2023, 10:16:49 pm »
To adjust everything but high frequency AC, you're right, you just need a very stable 10Vish DC voltage source and 10 kOhm ish resistor that you know the value of very accurately, plus a good short. The manual tells you how to make it out of a piece of copper wire. I'd argue just running this is good enough for most hobby use.

But for a proper calibration you should also verify all ranges and functions afterwards. How else can you verify uncertainties or get a report with data? That's what Keysight does if you send it to them.

Offline chilternview

  • Frequent Contributor
  • **
  • Posts: 499
  • Country: gb
Re: FS UK: Agilent 3458A 8.5 digit DMM
« Reply #11 on: October 24, 2023, 02:26:48 am »
Of course you need to know if your voltage and resistance reference is 'accurate'.... is  it calibrated, what is its uncertainty, what is it traceable to, how much may it have drifted due to temperature and/or ageing. And do you need an accredited calibration - all depends on what you want to use the 3458A for.
 

Offline deepfryed

  • Supporter
  • ****
  • Posts: 139
  • Country: au
Re: FS UK: Agilent 3458A 8.5 digit DMM
« Reply #12 on: October 25, 2023, 12:21:59 am »
I think somebody on this forum said (half-joking) the only 3458A's on ebay are the drifty ones no one wants, so buyer beware.

Out of curiosity, emailed the seller last night to ask if there's any paperwork on history / calibration & adjustments. No response  :-//

If someone on the forum buys it, I'd be interested to know how it turns out :)
 

Offline Oldtestgear

  • Regular Contributor
  • *
  • Posts: 184
  • Country: gb
Re: FS UK: Agilent 3458A 8.5 digit DMM
« Reply #13 on: October 26, 2023, 12:51:04 pm »
I have sold a couple of 3458A DMMs over the years. I always checked for drift before selling AND showed the data. Specifically excluding drift in a listing screams "I know it drifts".
Just my opinion but I would not pay anything like the starting price on what is  a huge gamble. 

Phil
 

Offline factory

  • Super Contributor
  • ***
  • Posts: 3666
  • Country: gb
Re: FS UK: Agilent 3458A 8.5 digit DMM
« Reply #14 on: October 26, 2023, 05:13:10 pm »
Or zero history and ERR on display... https://www.ppauctions.com/lot/185406/lot-1528
Not interested myself.

David
 

Offline alm

  • Super Contributor
  • ***
  • Posts: 2903
  • Country: 00
Re: FS UK: Agilent 3458A 8.5 digit DMM
« Reply #15 on: October 26, 2023, 08:42:52 pm »
Or zero history and ERR on display... https://www.ppauctions.com/lot/185406/lot-1528
Not interested myself.
It's always a risk, some more than others. For GBP 480 I would absolutely buy it, but I'm sure it won't stay like that. The ERR annunciator could be something innocent like the digital NVRAMs for reading stories having an empty battery, or something as big as needing a new A3 board. You have these risks with most used equipment in unknown condition, it's just that the stakes are higher here because of the repair costs. However, expect to pay a lot more for a unit with a documented drift history because obviously these are more desirable.

Offline veedub565

  • Frequent Contributor
  • **
  • Posts: 453
  • Country: gb
Re: FS UK: Agilent 3458A 8.5 digit DMM
« Reply #16 on: October 26, 2023, 09:13:42 pm »
Or zero history and ERR on display... https://www.ppauctions.com/lot/185406/lot-1528
Not interested myself.

David

I saw that and put a bid on around £400 (+VAT +fees). Not a bad price still, but I doubt it'll sell that cheap. And I already have a calibrated Datron 1281 so I don't desperately need the 3458a, although it would be nice to have.
 

Offline bastl_r

  • Regular Contributor
  • *
  • Posts: 160
  • Country: de
Re: FS UK: Agilent 3458A 8.5 digit DMM
« Reply #17 on: October 27, 2023, 06:57:13 am »
I cannot make an offer because there is "no shipping to germany" marked. :scared:
 

Offline Oldtestgear

  • Regular Contributor
  • *
  • Posts: 184
  • Country: gb
Re: FS UK: Agilent 3458A 8.5 digit DMM
« Reply #18 on: October 27, 2023, 09:18:31 am »
I have just looked at the 3458A being auctioned by PP Auctions. It was sent to Keysight in January this year -- see electrical safety check label -- but no sign of a calibration label.

Bit strange unless it was returned as "unrepaired"  or someone has peeled the calibration sticker off to try & get a great deal? Either way it is now too expensive for me to take a punt on...........

Phil
 

Offline chilternview

  • Frequent Contributor
  • **
  • Posts: 499
  • Country: gb
Re: FS UK: Agilent 3458A 8.5 digit DMM
« Reply #19 on: October 27, 2023, 10:34:10 am »
I have just looked at the 3458A being auctioned by PP Auctions. It was sent to Keysight in January this year -- see electrical safety check label -- but no sign of a calibration label.

Bit strange unless it was returned as "unrepaired"  or someone has peeled the calibration sticker off to try & get a great deal? Either way it is now too expensive for me to take a punt on...........

Phil

Yep. No cal label almost certainly means it failed, else why put it up for auction? At £1250 already its looking expensive as repair and cal is going to be in the £3k range.
 

Offline factory

  • Super Contributor
  • ***
  • Posts: 3666
  • Country: gb
Re: FS UK: Agilent 3458A 8.5 digit DMM
« Reply #20 on: October 27, 2023, 12:32:11 pm »
I'm suspicious that the high bidder is an alias of "reserve not met", think this will be appearing in quite a few future auctions.

Might give anyone seriously interested a chance to ask Keysight about the S/N and research if it's failed cal.

The company work outsourced calibration to still puts cal-due labels on failed items, even though it also states failed.

David
« Last Edit: October 27, 2023, 12:45:33 pm by factory »
 

Offline mendip_discovery

  • Super Contributor
  • ***
  • Posts: 1140
  • Country: gb
Re: FS UK: Agilent 3458A 8.5 digit DMM
« Reply #21 on: October 27, 2023, 01:18:09 pm »
I'm suspicious that the high bidder is an alias of "reserve not met", think this will be appearing in quite a few future auctions.
Either a punt for nothing or a shill to get it moving.

Quote
Might give anyone seriously interested a chance to ask Keysight about the S/N and research if it's failed cal.
:-DD They don't like customers let alone someone asking for a cert/details for free.

Quote
The company work outsourced calibration to still puts cal-due labels on failed items, even though it also states failed.

If I reject something it gets a cal date but where the due date is I put rejected. But I know from a 17025 (UKAS) that they feel it should be for the customer to decide so I have to use a "Limited Calibration see Certificate" label. But they also get funny about us putting due dates on things as it is for the customer to decide, the 12-month thing is a really lazy option and was mostly promoted by companies that wanted you to send the stuff back to them each year. If a customer doesn't state a frequency we just leave it blank.
Motorcyclist, Nerd, and I work in a Calibration Lab :-)
--
So everyone is clear, Calibration = Taking Measurement against a known source, Verification = Checking Calibration against Specification, Adjustment = Adjusting the unit to be within specifications.
 

Offline macaba

  • Regular Contributor
  • *
  • Posts: 227
  • Country: gb
Re: FS UK: Agilent 3458A 8.5 digit DMM
« Reply #22 on: October 27, 2023, 01:45:42 pm »
Or zero history and ERR on display... https://www.ppauctions.com/lot/185406/lot-1528

Approx. £3222.45 for a unit with a multislope rundown convergence error. Madness.
 

Offline chilternview

  • Frequent Contributor
  • **
  • Posts: 499
  • Country: gb
Re: FS UK: Agilent 3458A 8.5 digit DMM
« Reply #23 on: October 27, 2023, 02:01:03 pm »
a unit with a multislope rundown convergence error.
Just curious, but how do you know that? Did you go view it before auction?
 

Offline factory

  • Super Contributor
  • ***
  • Posts: 3666
  • Country: gb
Re: FS UK: Agilent 3458A 8.5 digit DMM
« Reply #24 on: October 27, 2023, 02:59:04 pm »
Or maybe asked the auction house to run the self test, or asked them for the error code, not an unreasonable request given what these make. Or asked Keyshite for a cal/repair quote against that S/N.  :-//

I'm 99.99% certain the house won it.

David
 


Share me

Digg  Facebook  SlashDot  Delicious  Technorati  Twitter  Google  Yahoo
Smf