Re: Siglent SDS2000X Plus
« Reply #3875 on: November 23, 2022, 11:26:27 pm »

-snip

But I saw you repeating one question about triggering and would like to know if I got it right:
you would like to be able to trigger to a serial digital pattern. Do you mean that in any particular protocol (UART, I2C, SPI, CAN) or some custom defined sequence?

Standard async serial (UART), trigger must take place on specific sequence of bytes taken from the decoder output.
Ok so you want to trigger on a string (a multibyte sequence) ?
2000X+ can trigger only on single character, as far as I know. I don't have one to check so someone with SDS2000X+ correct me if I'm wrong.
In some other protocols you can trigger on more detailed data.
 

Offline tautechTopic starter

  • Super Contributor
  • ***
  • Posts: 28471
  • Country: nz
  • Taupaki Technologies Ltd. Siglent Distributor NZ.
    • Taupaki Technologies Ltd.
Re: Siglent SDS2000X Plus
« Reply #3876 on: November 23, 2022, 11:35:10 pm »
I don't think HD protocol triggering is any different to X Plus or for that matter 5000X or 6000A models.

Some X Plus arrive in the next day or 2......  :phew:
Avid Rabid Hobbyist
Siglent Youtube channel: https://www.youtube.com/@SiglentVideo/videos
 

Offline markone

  • Frequent Contributor
  • **
  • Posts: 728
  • Country: it
Re: Siglent SDS2000X Plus
« Reply #3877 on: November 23, 2022, 11:45:05 pm »
Ok, I will check further  downloading manuals from official web site.

How about the PC application to manage LA operations, does it exist  ?
 

Online Martin72

  • Super Contributor
  • ***
  • Posts: 5893
  • Country: de
  • Testfield Technician
Re: Siglent SDS2000X Plus
« Reply #3878 on: November 23, 2022, 11:47:29 pm »
Quote
I don't think HD protocol triggering is any different to X Plus or for that matter 5000X or 6000A models.

Me too.
The decoding menus on my HD are 100% identical to the SDS2k+.
And I bet it is identical to 5000X/6000A also.
The decoders are in all cases simply outfitted, to say it in this way.

Offline tautechTopic starter

  • Super Contributor
  • ***
  • Posts: 28471
  • Country: nz
  • Taupaki Technologies Ltd. Siglent Distributor NZ.
    • Taupaki Technologies Ltd.
Re: Siglent SDS2000X Plus
« Reply #3879 on: November 23, 2022, 11:53:15 pm »
How about the PC application to manage LA operations, does it exist  ?
Do you mean remotely ?
Yes the webserver can do that and probably the SCPI command set too, I need check or you can in the programming manual:
https://int.siglent.com/u_file/download/22_07_25/SDS_ProgrammingGuide_EN11C.pdf
Avid Rabid Hobbyist
Siglent Youtube channel: https://www.youtube.com/@SiglentVideo/videos
 

Offline markone

  • Frequent Contributor
  • **
  • Posts: 728
  • Country: it
Re: Siglent SDS2000X Plus
« Reply #3880 on: November 24, 2022, 12:26:08 am »
How about the PC application to manage LA operations, does it exist  ?
Do you mean remotely ?
Yes the webserver can do that and probably the SCPI command set too, I need check or you can in the programming manual:
https://int.siglent.com/u_file/download/22_07_25/SDS_ProgrammingGuide_EN11C.pdf

Thank you very much, I just checked the SCPI manual and for what concern LA commands I can only find something related to pattern trigger, it does not seems that you can operate the LA acquisition remotely.

My main need would be to transfer LA acquisition memory to a windows PC in order to visualize, analyze and store it to disk by an existing Windows application. 
 

Offline mwb1100

  • Frequent Contributor
  • **
  • Posts: 529
  • Country: us
Re: Siglent SDS2000X Plus
« Reply #3881 on: November 24, 2022, 12:39:39 am »
My main need would be to transfer LA acquisition memory to a windows PC in order to visualize, analyze and store it to disk by an existing Windows application.

I think you want to look at the ":WAVeform:DATA" query.  There is an example for downloading the data for a digital channel starting on page 562 of the PDF.
 

Offline tautechTopic starter

  • Super Contributor
  • ***
  • Posts: 28471
  • Country: nz
  • Taupaki Technologies Ltd. Siglent Distributor NZ.
    • Taupaki Technologies Ltd.
Re: Siglent SDS2000X Plus
« Reply #3882 on: November 24, 2022, 12:45:20 am »
My main need would be to transfer LA acquisition memory to a windows PC in order to visualize, analyze and store it to disk by an existing Windows application.
I think you can do that with a CSV or BIN file save.
With the front panel blue Print button we get PNG screenshots to internal or external storage but within the Save/Recall menu we have more filetype options which we can operate locally or remotely and even save to specified addresses.
The File Manager feature in the Save/Recall menu is quite powerful.
Avid Rabid Hobbyist
Siglent Youtube channel: https://www.youtube.com/@SiglentVideo/videos
 

Offline mwb1100

  • Frequent Contributor
  • **
  • Posts: 529
  • Country: us
Re: Siglent SDS2000X Plus
« Reply #3883 on: November 24, 2022, 12:59:40 am »
As far as getting trigger on a multiple byte sequence of data  (and assuming Saleae is not what you want to use), you might want to consider using a microcontroller with a simple firmware program that monitors the serial port and when it sees the data sequence you're interested in it toggles a GPIO that the oscilloscope or logic analyzer you're using can trigger on.  I'd think this would be fairly straightforward to code up since sample code that reads/writes a serial port is generally readily available for most any microcontroller.

With a setup like that, you'd be able to use whatever scope or LA you want.

I don't know how easy that'd be compared to whatever capability Saleae's tools have (do you need to write an "HLA" to be able to trigger on an arbitrary byte sequence?).
 

Offline markone

  • Frequent Contributor
  • **
  • Posts: 728
  • Country: it
Re: Siglent SDS2000X Plus
« Reply #3884 on: November 24, 2022, 01:02:46 am »
My main need would be to transfer LA acquisition memory to a windows PC in order to visualize, analyze and store it to disk by an existing Windows application.

I think you want to look at the ":WAVeform:DATA" query.  There is an example for downloading the data for a digital channel starting on page 562 of the PDF.

Yes, you are right, it's possible to select digital channel as source data :

:WAVeform:SOURce D0
:WAVeform:DATA?

For some reason i was convinced that :WAV command group was related only to analog wavforms, but i was wrong    :)

Ok, time to study in detail SCPI manual ...
 

Offline markone

  • Frequent Contributor
  • **
  • Posts: 728
  • Country: it
Re: Siglent SDS2000X Plus
« Reply #3885 on: November 24, 2022, 01:18:15 am »
As far as getting trigger on a multiple byte sequence of data  (and assuming Saleae is not what you want to use), you might want to consider using a microcontroller with a simple firmware program that monitors the serial port and when it sees the data sequence you're interested in it toggles a GPIO that the oscilloscope or logic analyzer you're using can trigger on.  I'd think this would be fairly straightforward to code up since sample code that reads/writes a serial port is generally readily available for most any microcontroller.

With a setup like that, you'd be able to use whatever scope or LA you want.

I don't know how easy that'd be compared to whatever capability Saleae's tools have (do you need to write an "HLA" to be able to trigger on an arbitrary byte sequence?).

As I explained in another thread, MCU board programmed on the purpose is the system that I used till know, but i would like to change approach to be more flexible on the field and because now i'm developing many application on half duplex mono wire serial buses and I need to get actual sampling data of the line to spot bus collisions.

Of course i have other user cases for the DSO iteself, otherwise the purchase would not be justified to sniff serial line.

If LA command set includes what I need, i will add the LA kit on sale.
 

Offline BillyO

  • Super Contributor
  • ***
  • Posts: 1434
  • Country: ca
Siglent SDS2000X Plus - features help
« Reply #3886 on: November 28, 2022, 03:32:04 am »
I'm reaching out to more experienced (than me) SDS2000XP users to get some help.

I'm trying to set the time zone and I'm also trying to set up network storage, however I can't seem to get either to work and neither one is in the manual.  Has anyone been successful in getting these features to work?
Bill  (Currently a Siglent fanboy)
--------------------------------------------------
Want to see an old guy fumble around re-learning a career left 40 years ago?  Well, look no further .. https://www.youtube.com/@uni-byte
 

Offline tubularnut

  • Regular Contributor
  • *
  • Posts: 225
  • Country: gb
Re: Siglent SDS2000X Plus - features help
« Reply #3887 on: November 28, 2022, 04:02:21 am »
I'm reaching out to more experienced (than me) SDS2000XP users to get some help.

I'm trying to set the time zone and I'm also trying to set up network storage, however I can't seem to get either to work and neither one is in the manual.  Has anyone been successful in getting these features to work?

Hi,

Actually there is a "Timezone bug" present, I guess it´s "valid" for all touchscreen-models (2k+, 2k HD, 5k+, 6k A) because of more or less the same software.
 
The following users thanked this post: BillyO

Offline BillyO

  • Super Contributor
  • ***
  • Posts: 1434
  • Country: ca
Re: Siglent SDS2000X Plus
« Reply #3888 on: November 28, 2022, 04:12:07 am »
Well, I was certainly not going to pull out the "B" word again.  The last time I did it created an international kerfuffel.

So that explains the undocumented time zone feature .. it can't be set.

How about the network storage? 
Bill  (Currently a Siglent fanboy)
--------------------------------------------------
Want to see an old guy fumble around re-learning a career left 40 years ago?  Well, look no further .. https://www.youtube.com/@uni-byte
 

Offline bingo600

  • Super Contributor
  • ***
  • Posts: 1989
  • Country: dk
Re: Siglent SDS2000X Plus
« Reply #3889 on: November 28, 2022, 02:35:34 pm »
How about the network storage?

From the 1.5.2R1 - RelNotes

Code: [Select]
Fixed
 Network storage only works for SMB1.0

What vers are you running ?

« Last Edit: November 28, 2022, 02:37:47 pm by bingo600 »
 
The following users thanked this post: BillyO

Offline BillyO

  • Super Contributor
  • ***
  • Posts: 1434
  • Country: ca
Re: Siglent SDS2000X Plus
« Reply #3890 on: November 28, 2022, 03:15:32 pm »
SMB 1.0.  My NAS is rather old and does not support anything better.

Is there any documentation about setting it up?  It is, or should be, fairly simple unless they have done something weird.  I'm using:

Drive letter: Z:
Directory: //[server_name]/share
User: [server_name]/[user_name]
Password: [password]


Result:  Mount directory error: please check your parameters!

I have tried slashes the other way in various combination.  I've tried not including the [server_name] in the user name.  I've tried different drive letters.  The NAS is a Thecus N4100PRO and it's running a form of Linux.


Update:  Never mind.  Even though prompts in the setting page suggest using the server name, the scope seems to have a problem with name resolution.  I tried the IP address of the server and it worked.  Thanks for helping out.

Update #2:

So this is what worked for me:

Drive letter: I:
Directory: //[server_IP]/[share_name]
User: [server_name]/[user_name]
Password: [password]

When is saves something, the scope creates a directory called "SIGLENT" on the specified share and stores files there.

Just an opinion, but if Siglent want to considered a top tier brand, they will have to up their game on documentation.  Any time a feature is added or amended they should create and publish an addendum to the documentation and/or update the documentation.

Also in the release note for 1.5.2R1 - "Supported time zone setting"   :(
« Last Edit: November 28, 2022, 03:34:50 pm by BillyO »
Bill  (Currently a Siglent fanboy)
--------------------------------------------------
Want to see an old guy fumble around re-learning a career left 40 years ago?  Well, look no further .. https://www.youtube.com/@uni-byte
 

Offline Peter_O

  • Frequent Contributor
  • **
  • Posts: 421
  • Country: de
Re: Siglent SDS2000X Plus
« Reply #3891 on: November 28, 2022, 05:36:21 pm »
snip: Just an opinion, but if Siglent want to considered a top tier brand, they will have to up their game on documentation. 

Such approach would end up in top tier brand prices too. Not what I want.   ;)
 
The following users thanked this post: Martin72

Offline BillyO

  • Super Contributor
  • ***
  • Posts: 1434
  • Country: ca
Re: Siglent SDS2000X Plus
« Reply #3892 on: November 28, 2022, 08:04:53 pm »
Such approach would end up in top tier brand prices too. Not what I want.   ;)

Not really.  Not in this day and age.  Modern publication and content management tools would make it a snap and they have been around for a couple of decades now so there is a lot of expertise surrounding them and proper use practices.
Bill  (Currently a Siglent fanboy)
--------------------------------------------------
Want to see an old guy fumble around re-learning a career left 40 years ago?  Well, look no further .. https://www.youtube.com/@uni-byte
 

Online Martin72

  • Super Contributor
  • ***
  • Posts: 5893
  • Country: de
  • Testfield Technician
Re: Siglent SDS2000X Plus
« Reply #3893 on: November 28, 2022, 08:39:50 pm »
Quote
Not really.  Not in this day and age. 

I could go and tell it our sales department, maybe our customers would get the documentations for free then instead of paying extra. ;)

BTW, in lecroy manual you won´t find further explanations to the described functions.


Offline BillyO

  • Super Contributor
  • ***
  • Posts: 1434
  • Country: ca
Re: Siglent SDS2000X Plus
« Reply #3894 on: November 28, 2022, 11:25:22 pm »
BTW, in lecroy manual you won´t find further explanations to the described functions.

In Siglent manuals you won't find the functions described at all.


I guess everyone is okay with poor documentation? :-//
Bill  (Currently a Siglent fanboy)
--------------------------------------------------
Want to see an old guy fumble around re-learning a career left 40 years ago?  Well, look no further .. https://www.youtube.com/@uni-byte
 

Online 2N3055

  • Super Contributor
  • ***
  • Posts: 6738
  • Country: hr
Re: Siglent SDS2000X Plus
« Reply #3895 on: November 28, 2022, 11:37:22 pm »
BTW, in lecroy manual you won´t find further explanations to the described functions.

In Siglent manuals you won't find the functions described at all.


I guess everyone is okay with poor documentation? :-//

Programers manual (for SCPI) is reference manual. And Siglent manual is not less detailed than Keysight one. On my 3000T I also use SCPI manual for details..

In perfect world we would still have documentation like old Hp had. But today they don't have it either..
 

Online Martin72

  • Super Contributor
  • ***
  • Posts: 5893
  • Country: de
  • Testfield Technician
Re: Siglent SDS2000X Plus
« Reply #3896 on: November 28, 2022, 11:40:41 pm »
Quote
I guess everyone is okay with poor documentation?

My guess is, the manufacturers assume that you already know for what the functions are good for.
It´s not their task to educate you, this must be happen before.

Offline mwb1100

  • Frequent Contributor
  • **
  • Posts: 529
  • Country: us
Re: Siglent SDS2000X Plus
« Reply #3897 on: November 29, 2022, 12:09:15 am »
I'm sure that plenty of people here remember the days when you'd get schematics and possibly even source code for the software (remember that IBM published the BIOS source for the PC).

Sometimes you can still get that info, but almost never in the manual anymore.  You've gotta dig it out of some support site or github repository - if it's made available at all.

I feel like I need a custom rant emoji.
 
The following users thanked this post: balnazzar

Offline BillyO

  • Super Contributor
  • ***
  • Posts: 1434
  • Country: ca
Re: Siglent SDS2000X Plus
« Reply #3898 on: November 29, 2022, 03:09:24 am »
Programers manual (for SCPI) is reference manual.
Not sure about that.  There are many things about the scope not included in the programmers manual.
Bill  (Currently a Siglent fanboy)
--------------------------------------------------
Want to see an old guy fumble around re-learning a career left 40 years ago?  Well, look no further .. https://www.youtube.com/@uni-byte
 

Offline BillyO

  • Super Contributor
  • ***
  • Posts: 1434
  • Country: ca
Re: Siglent SDS2000X Plus
« Reply #3899 on: November 29, 2022, 03:11:31 am »
Quote
I guess everyone is okay with poor documentation?

My guess is, the manufacturers assume that you already know for what the functions are good for.
It´s not their task to educate you, this must be happen before.
Just how are you to be pre-educated on a new (undocumented) feature to a scope?
Bill  (Currently a Siglent fanboy)
--------------------------------------------------
Want to see an old guy fumble around re-learning a career left 40 years ago?  Well, look no further .. https://www.youtube.com/@uni-byte
 