At the rated voltage (1KV) it certainly is possible to draw an arc large enough to damage the contacts. I had made a video to demonstrate this after anther member had damaged a 121GW. While you should never rotate the function switch with the meter connected to an active circuit, a few of us have bad habits. Mix that with a little HV.....
Fungus finds the fungus...

)At the rated voltage (1KV) it certainly is possible to draw an arc large enough to damage the contacts. I had made a video to demonstrate this after anther member had damaged a 121GW. While you should never rotate the function switch with the meter connected to an active circuit, a few of us have bad habits. Mix that with a little HV.....
Except that's not what happened here.
The fault was there out of the box, this is not a used meter, and there is no charring.
It's a physical damage issue, not electrical.
The energy to do that couldn't have come via the V jack, there is at least a PTC + 1K in series with every path. Even if the contacts shorted the VA range contacts and the ohms contacts, worst case path is V jack to GND via a PTC + 1K.
Why power switch has to be on rotary switch not separate switch. My Metex has separate power ON OFF switch, it is very usefull.
Um, what is the omission you are talking about?
I just see a gorilla used 180 grit sandpaper on the rotary switch portion of the PCB. They should use an angle grinder instead. WTF are they doing. Those scratches are terrible on thin copper.
The rotary switch tracks don't look gold plated? Maybe they are sanding off the ENIG and plating something else there. edit: nope, the via's still are gold. So they have some selective plating going on here.
I don't see pogo-pin indentations anymore.
P.S. - If you clean the PCB with IPA, don't forget to lubricate the switch portion.
Um, what is the omission you are talking about?
I just see a gorilla used 180 grit sandpaper on the rotary switch portion of the PCB. They should use an angle grinder instead. WTF are they doing. Those scratches are terrible on thin copper.
The rotary switch tracks don't look gold plated? Maybe they are sanding off the ENIG and plating something else there. edit: nope, the via's still are gold. So they have some selective plating going on here.
I don't see pogo-pin indentations anymore.
P.S. - If you clean the PCB with IPA, don't forget to lubricate the switch portion.
which lubricant do you recommend
this is not burned but the layer is separated ... i belive bad procesing
Is that PCB pothole on your multimeter? I don't see it in your first pics so I'm a bit confused.
I believe the picture Kartika linked in their last post is simply an enlargement of the original one that Dave linked, with their (Kartika's) opinion as to the cause.
It is not good to see those scrubbing marks though, very disappointing from a manufacturer of Brymen's reputation.