Hello.
I have just purchased a BM869s(elma rebrand in scandinavia), and i like it, but i am a bit unsure if something is wrong.
If i simply take the leads put into COM and ohm/v/capacitance and put on table, spaced around 10 centimeters apart, and toggle through the different modes, it does not go to zero, and isnt fully "stable".
I have gone through all settings and written down what happens, so perhaps someone can tell me if something is wrong or not. the probes are not touching eachother, or anything for that matter.
here goes:
first position:
AC voltage mode:
voltage: 5.xx to 0.1v over 10sec
Hz mode:
voltage: 3.xx to 000.1v over 5 seconds
second position:
AC voltage mode:
0.3xx - 0.08xxv over 10 sec
dBm mode:
starts at 4dbM, goes to 2.3x over 5 seconds
hz mode:
starts at 0.09xv goes to 0.083 in 5 seconds
third position:
DC voltage mode:
starts at 0.03xxx and very fast goes to -0.0000v, analog bargraph flipping the second "bar"
AC + DC dual display mode
AC voltage: 0.15x to 0.086v in 5 seconds
DC voltage takes ~5-10 sec to go down to 0.001v
AC+DC + AC dual display mode:
AC starts at ~0.1 and goes to 0.86 over 5 sec
DC+AC starts at 0.51xx and quickly goes to 0.086xv
fourth position:
DC mv mode:
starts at i think ~3mV quickly (1 sec) goes to 0.03mV, bargraph swings 3-4 bars a couple of times pr second
DC+AC dual mv mode:
AC starts at 95.x mV and stays there
DC goes fast to 0.04mV and stays there
DC+AC + AC dual mode display:
AC starts and stays at roughly 95-96mV
DC+AC starts high, and inside of 1 sec jumps to ~94-95.xx mV
Hz mode:
00.000Hz
D% mode:
00.00
fifth position:
AC mV mode:
HZ starts at 50.01Hz, stays there
AC starts at ~150.xxmV and within 1 sec goes to 92.xx mV
Hz/dbm mode:
49.99hz
starts at 4dBm and within 1sec goes to 3.4x
sixth position (temps):
shows nothing on all modes
seventh position:
diod: 0L
capacitance: 00.02nF, instant there, doesnt change
eighth position:
ohms: autoranges down to .0L Mohm normally. same with beeper
conductance: starts at 0.1x and swings around doing down to 0.01nS over a couple of seconds
ninth position:
DC mV:
starts at -00.001, jumps between -00.000. -23.99 to -25 % 4-20. seems proper
DC+AC mA
AC starts at 00.01 mA, goes to 00.00 after a second or so
DC starts at 00.000 and goes to 00.001 after a few seconds.
DC+AC + AC
AC starts at 00.00
DC+AC starts at 6.4xx mA, after 10 seconds settles to 0.004mA
AC:
0.000Hz instantly
AC amperage starts and stays at 00.001mA
tenth position:
starts and zeroes down in at most 1 second on all modes
Thanks
Why do you think something is wrong? It's a bit weird to just chuck a pile of numbers at us and expect us to guess what "looks wrong" about it.
Anyway, I'm guessing you're expecting 0.00000V to be displayed on the DC volts ranges? Again, you should ask yourself why you expect that - if you short the probes together, that's 0V. If you leave the probes disconnected, there's nothing in particular defining the voltage across the probes, so the readout is undefined. The megaohm impedance of the multimeter itself will bring it close to zero, but not precisely so.
In short: short your leads to test for 0V.
thanks.
the reason i was a bit unsure is that all the other (albeit much less precise and less featureful) meters i've had, fluctuated far far less with flying leads.
I just retested with leads shorted, and it is not nearly like the numbers i posted, but it still takes several seconds to settle even after it settles scale, and on some of the AC/DC volt settings it never fully zeroes
I did not say it was broken, I said I did not know.
Don't forget that the digits on the right hand side of your BM869 are down in the microvolts; it's not fair to compare them to the digits on the RHS of less precise meters.
It is certainly possible that the BM869 is genuinely more noisy/unstable with open probe leads; but this is not a problem or fault at all.
The readings you get on AC volts could well be genuine, a single turn of wire will pick up a certain amount of mains hum.
yes i was thinking that it might be because this is a much much more precise instrument than those $10 things i have been testing before, just wanted to make sure
Thank you
Random number showing up on unconnected leads is to be expected. If you get random wild readings when connected to something, then I would worry.
yes i was thinking that it might be because this is a much much more precise instrument than those $10 things i have been testing before, just wanted to make sure
Thank you
Don't forget, you have a high input impedance on your meter. When they are floating, stray fields can cause very minute currents to form. These minute currents across the high impedance give the fluctuating voltages that you see.