Sorry about the quality of the pictures, it is my first public teardown and I don't have a proper lightbox.
Like I said, it's not that ugly, but the soldering quality is mediocre, through-hole resistors hand-soldered on the surface, a lot of flux residue, dubious components like those "Chang" capacitors, "Golden Power" battery.
About the design itself, even if it's UL listed, I don't see a lot of input protection, there is one MOV on the small board, a PTC in the corner and that's it.

QuoteLike I said, it's not that ugly, but the soldering quality is mediocre, through-hole resistors hand-soldered on the surface, a lot of flux residue, dubious components like those "Chang" capacitors, "Golden Power" battery.Batteries are easily exchangeable, but capacitors are another story and most won't go through the trouble to put in something better. Having said that, the capacitors are probably not going to be stressed so they should last a while.
QuoteAbout the design itself, even if it's UL listed, I don't see a lot of input protection, there is one MOV on the small board, a PTC in the corner and that's it.For input protection on the volts/ohm jack, most of the good ones have a PTC, MOVs (plural) and some fusible or power resistor. The better ones also have components with higher rated voltages like 1000V. Better meters also have 1000V rated fuses.
As you can see in the EX430 (see snapshot), the maximum input protection for resistance, frequency, capacitance, duty cycle, diode test and continuity is 250Vrms for 15 seconds max. The EX430 fuses are rated for 250V.
In addition to quality components, better meters have isolation slots, molded case plastic to contain any possible arc over and good creepage distance between traces/components.
This is my experience with the EX430.
https://www.eevblog.com/forum/testgear/extech-ex430-multimeter-teardown-photos/
They have a separate board for the input. Why not implement a better input protection to that board?![]()
![]()
I thought that CAT II and III rating implied the meter being able to sustain a several kV transient. How will they be able to cut 8 kV here ?
Yeah, I have a Fluke 87 III and it's really impressive compared to this meter in terms of input protection.
That's weird, I have already searched for a teardown of the EX430 in order to compare with mine, but haven't found this topic. So my post is pretty much a duplicate. It may still be interesting to compare a few different units.
I guess more input protection components would increase the cost of the device.
I thought that CAT II and III rating implied the meter being able to sustain a several kV transient. How will they be able to cut 8 kV here ?Dave explains it in eevblog #373 (start at around 15:56m for the v/ohm input protection).
In addition, I included the snippet from the Fluke 87 series II service manual that is CAT III 1000V rated. As well, I included a Littlefuse MOV varistor datasheet showing the maximum clamping voltage (1500) for a 575V rated varistor.
So with the power/fusible resistors, MOVs (plural) and PTC working together, I suspect that is how the 8kV figure is achieved.
Actually my question was how does the Extech 430 achieve a high voltage isolation capacity ?
Is the big blue round guy on the daughter board a MOV ?
Do you think this Extech 430 really meets its CAT rating ?
QuoteIs the big blue round guy on the daughter board a MOV ?Yes. Are there markings on it?
I would have no doubt that the EX430 design actually meets its CAT ratings. The proof is the UL listing.
Yep, "VDR 14D220K". There a few results on alibaba that say it's a varistor.

Yep, "VDR 14D220K". There a few results on alibaba that say it's a varistor.From two different searches, the part number suggest it is a 22V rated???![]()
And what the hell is this ?!
And what the hell is this ?!
I tried to follow all the tracks of the daughter board and it looks like this varistor is placed between the mA input (after the fuse) and common. So it doesn't protect the Volts input. Its purpose is to short "mA" and "common" inputs and let the fuse blow. Do you agree with this ?
I don't have the EX420 anymore, but on the pictures I took of it you can see that all 4 resistors are on it. Doesn't mean anything of course, but on the 430 pictures it really looks like that resistor did fall off during construction...
Attached is the relevant EX420 picture.
EDIT: I just noticed the EX420 input pcb says 430, they probably use the same pcb for both, so it really looks like OP's 430 is missing the resistor...