Is it any more dangerous to float my Rigol 1052 and take sensible precautions such as:
Using a shrouded probe with no ground clip attached (I even have probes with shrouded BNC connector at the scope end)
Use a scope probe 'clip' so you don't need to hold the probe on the test point when powering the DUT - even solder a short wire to the test point and clip your probe to that if needs be.
Use a separate lead from floating scope ground to croc clip on floating hot ground
Than it is to use my Fluke 79 with the negative terminal clipped to hot ground while I poke around measuring voltages on the SMPS primary side with the positive probe?
Both meter and scope have no, or very few, exposed metal parts at floating potential that could be accidentally touched.
I honestly don't think the way we were taught 'back in the day' was irresponsible or dangerous - I think we were just properly taught to understand exactly what we were doing, the risks involved and how to minimise them. This included using shrouded probes, a separate ground connection to the chassis (not on the scope probe) and clipping the scope probe to component lead you wanted to test, then powering up the DUT. Oh and there was nothing grounded on the wooden bench, plus it had a rubber surface mat and you stood on a rubber mat.So you already know there is a lot more to it than just cutting the earth wire! Good, because many people think an isolation transformer is enough and happily ground the DUT again using their oscilloscope's ground lead.
The modern day equivalent to setting up an entire isolated work environment is a CAT rated differential probe. Much cheaper and much easier..... And the differential probe is the cheapest chinese found from ebay where the isolation can be anything a common semiconductor junction included.
I'm just in the middle of a review and teardown of an isolated Siglent portable scope.


Ground The Instrument.
The instrument is grounded through the Protective Earth lead of the power cord. To avoid electric shock, it is essential to connect the earth terminal of power cord to the Protective Earth terminal before any inputs or outputs.
DS1000Z series digital oscilloscopes can make measurements in Measurement Category I.
WARNING This oscilloscope can only be used for measurements within its specified measurement categories.
Measurement category I is for measurements performed on circuits not directly connected to MAINS. Examples are measurements on circuits not derived from MAINS, and specially protected (internal) MAINS derived circuits. In the latter case,
transient stresses are variable; for that reason, the transient withstand capability of
the equipment is made known to the user.
Fortunately, there are solutions available which are not only safe but much more accurate than the extremely dangerous and unacceptable procedure of floating your oscilloscope. These fall into general categories: differential probes, isolated input oscilloscopes, isolation transformers and isolation amplifiers.
“Floating” a ground-referenced oscilloscope is the technique of defeating the oscilloscope’s protective grounding system disconnecting “signal common” from earth, by either defeating the grounding system or using an isolation transformer. This technique allows accessible parts of the instrument such as chassis, cabinet, and connectors to assume the potential of the probe ground lead connection point. This technique is dangerous, not only from the standpoint of elevated voltages present on the oscilloscope (a shock hazard to the operator), but also due to cumulative stresses on the oscilloscope’s power transformer insulation. This stress may not cause immediate failure, but may lead to future dangerous failures (a shock and fire hazard), even after returning the oscilloscope to properly grounded operation. Not only is floating a ground-referenced oscilloscope dangerous, but the measurements are often inaccurate. This potential inaccuracy results from the total capacitance of the oscilloscope chassis being directly connected to the circuit-under-test at the point where the ground lead is connected

If you read documentation from scope manufacturers back then, they also recommended against it.
Unlike in a formal course where participants have some understanding of the circuitry and the risks involved, on a public forum where all and any can read and deduce rightly or wrongly the procedures and risks involved it is nothing but irresponsible to discuss the merits of floating a scope. Sure some do it but it should never be promoted in a venue such as this as we have no idea of the skill level and understanding of those reading.
IMNSHO the only way that statement could have been improved is if it had been posted in a larger font.
If your neutral/earth are tied together you need to revise your wiring.
Unlike in a formal course where participants have some understanding of the circuitry and the risks involved, on a public forum where all and any can read and deduce rightly or wrongly the procedures and risks involved it is nothing but irresponsible to discuss the merits of floating a scope. Sure some do it but it should never be promoted in a venue such as this as we have no idea of the skill level and understanding of those reading.
IMNSHO the only way that statement could have been improved is if it had been posted in a larger font.
You can both bite me.
Do you really think it better to not discuss the merits and hazards and instead rely on an argument from authority or silence? How do you expect people to gain understanding without discussion? Nancy Reagan's "just say no" campaign was as much a success as abstinence only education.
My feeling exactly! Stifling discussion achieves nothing. An informed discussion that lays out the risks (and any merits) can only be beneficial. This is an electronics discussion forum with many experienced, informed and articulate engineers and hobbyists. This is exactly where this discussion should occur. There are countless places on the web where incomplete or wrong information about this and related topics is available. Refusing to discuss it here and provide correct information does not make anyone more safe.Bravo to those who have provided good information here. I always learn more by reading these threads.

Do you really think it better to not discuss the merits and hazards and instead rely on an argument from authority or silence? How do you expect people to gain understanding without discussion?
The transformer issue is a tough one. Off-line powered instruments rely on the 50/60 Hz or switching power transformer for isolation from neutral and hot (1) and floating the instrument by breaking the ground connection increases the potential across the transformer. Some old instruments specified a peak float voltage and incorporated a special transformer to handle floating operation.
During floating operation, any exposed ground connection including the BNC connectors can become a hazard. Instruments designed with this in mind have special construction which tries to prevent exposure and include special BNC connectors and probes. Fluke uses or used plastic BNC connectors much like the shielded banana jacks now common on multimeters with probes to match; none of the BNC coaxial ground connection is exposed.
Even if you do float the oscilloscope safely using an external isolation transformer, considerable common mode capacitance still exists and all of the probe grounds are still tied together. Oscilloscopes designed with isolated inputs largely avoid these problems although their inputs are still unbalanced by common mode capacitance through the environment and the isolated power supply.
If you read documentation from scope manufacturers back then, they also recommended against it.
It depends on the time period and manufacturer. Some recommended floating the oscilloscope by disconnecting the ground and even specified the maximum floating voltage with little or no warning about the dangers. Suitable high voltage differential probes were very rare (2) and isolated probes and inputs did not exist leaving A-B measurements with their limitations, floating the device under test which is not always practical, and floating the oscilloscope.
Unlike in a formal course where participants have some understanding of the circuitry and the risks involved, on a public forum where all and any can read and deduce rightly or wrongly the procedures and risks involved it is nothing but irresponsible to discuss the merits of floating a scope. Sure some do it but it should never be promoted in a venue such as this as we have no idea of the skill level and understanding of those reading.
IMNSHO the only way that statement could have been improved is if it had been posted in a larger font.
You can both bite me.
Do you really think it better to not discuss the merits and hazards and instead rely on an argument from authority or silence? How do you expect people to gain understanding without discussion?

An informed discussion that lays out the risks (and any merits) can only be beneficial.
If you read documentation from scope manufacturers back then, they also recommended against it.
It depends on the time period and manufacturer. Some recommended floating the oscilloscope by disconnecting the ground and even specified the maximum floating voltage with little or no warning about the dangers. Suitable high voltage differential probes were very rare (2) and isolated probes and inputs did not exist leaving A-B measurements with their limitations, floating the device under test which is not always practical, and floating the oscilloscope.
[,,,]
(2) Tektronix had some suitable differential amplifiers starting in 1960 but the first suitable x10 probes which could be used with them in a high voltage application were introduced in 1962. They introduced a good differential amplifier in 1967.
and a BIG thanks for nothing to the "JUST DON'T DO IT" brigade![]()
when all that a responsible technician, advanced hobbyist or EE requires is some simple 'how to and how not to' information,
so they can decide if their equipment and skill set is up to the task (or not) for a particular probe and measurement scenario,
taking ALL possible safety precautions,
including 'think before you probe'![]()
![]()
![]()
Otherwise most uninformed/misinformed/Darwin Dumbass wannabees will try something NQR anyway, and BANG !![]()
![]()
I've seen very few people on this forum, in this or the multiple past threads on this topic, say that floating a scope is ever ok. The vast majority of posters always say "don't do it". That is as it should be. The more important thing IMO, is the explanation and discussion about why it is not ok to do it. There are many people who will not heed safety advice unless they can be made to understand *WHY* a certain practice is unsafe.