Author Topic: Multimeter spreadsheet (80 manufacturers and over 400 meters listed)  (Read 1070104 times)

0 Members and 22 Guests are viewing this topic.

Offline WytnuclsTopic starter

  • Super Contributor
  • ***
  • Posts: 3045
  • Country: be
Re: Multimeter spreadsheet
« Reply #175 on: December 08, 2014, 11:11:25 am »
I overlooked the UNI-T UT805A, which has a respectable 200,000 count.
Here is the updated spreadsheet:
« Last Edit: May 09, 2017, 10:18:39 pm by Wytnucls »
 

Offline WytnuclsTopic starter

  • Super Contributor
  • ***
  • Posts: 3045
  • Country: be
Re: Multimeter spreadsheet
« Reply #176 on: December 08, 2014, 11:28:03 am »
This is a fantastic resource!  I really appreciate all that it takes to compile and maintain a list like this.  As I zero in on my decision, I want to be sure that I truly have this correct...
I get the difference between Accuracy and Precision, so we don't need to go there.  But I notice on some of the higher end units (50,000 counts at .02% + 2 digits), it seems like the precision exceeds the accuracy.
So if I measure 5 volts, it should show as 5.0000 (that would be th 50,000 counts of precision - correct?)  .02% would then be +/- .001 V (ignoring the +2 digits for now).  If I am correct to this point, then the last digit exceeds the accuracy of the device.  Am I thinking of this correctly?
I'm not sure I understand your post. Accuracy is never better than resolution. In your 50,000 resolution example, the accuracy of 0.02%+2, means that your 5V reading could be anywhere between 5.0012V and 4.9988V and still be within specs.
 

Offline KD0RC

  • Regular Contributor
  • *
  • Posts: 114
  • Country: us
Re: Multimeter spreadsheet
« Reply #177 on: December 09, 2014, 03:32:15 am »

I'm not sure I understand your post. Accuracy is never better than resolution. In your 50,000 resolution example, the accuracy of 0.02%+2, means that your 5V reading could be anywhere between 5.0012V and 4.9988V and still be within specs.
[/quote]
Thanks for that.  I expected (naively...) that the last digit would be where the uncertainty would lie.  I am not an engineer, so I often make assumptions that are not correct.  Your answer confirms my statement that the last digit of precision exceeds the accuracy.  That is probably not the right way to say it, but now I understand what I am looking at when given the resolution and accuracy specs.
I don't really need .02% for the hobbyist things that I do, but it would be orders of magnitude better than anything I have ever owned, and for my purposes, would act like a standard against which to check my other gear.
 

Offline johansen

  • Super Contributor
  • ***
  • Posts: 1436
Re: Multimeter spreadsheet
« Reply #178 on: December 09, 2014, 04:08:24 am »
Accuracy is never better than resolution.

resolution is cheaper than accuracy.

I had an old fluke kelvin varley vacuum tube volt meter with its 49,999 count resistor divider..
interesting.. for certain. but i can measure microvolts with a 4$ digital scale today.

but can i do better than a weston cell voltage standard?
really good question. i was able to measure my weston cell (my version of the fluke differential vacuum tube volt meter came with it internally) back in 2010 to a 1.0186mV with +/- .1mV but that's not really a sufficiently solid datapoint is it...
 

Offline HighResolution

  • Contributor
  • Posts: 12
  • Country: us
Re: Multimeter spreadsheet
« Reply #179 on: December 25, 2014, 07:41:36 am »
The GW Instek GDM-461 look very similar and has very similar (identical?) specs to the UNI-T UT61E but has 1000V fuses according to the manual.

Costs ~$100.  Might be a good buy to get those specs with better input protection?
 

Offline cncjerry

  • Supporter
  • ****
  • Posts: 1475
Re: Multimeter spreadsheet
« Reply #180 on: December 26, 2014, 07:09:25 am »
Nice work and this is a nit: HP 3478a has GPIB, I have two; HP 3468a doesn't. 
 

Offline WytnuclsTopic starter

  • Super Contributor
  • ***
  • Posts: 3045
  • Country: be
Re: Multimeter spreadsheet
« Reply #181 on: December 26, 2014, 09:25:03 am »
Correct, the HP3468A/B has a HP-IL interface instead of HP-IB. Database will be corrected accordingly.
 

Offline WytnuclsTopic starter

  • Super Contributor
  • ***
  • Posts: 3045
  • Country: be
Re: Multimeter spreadsheet
« Reply #182 on: January 02, 2015, 02:55:29 am »
Yes, you're right. Probably confused with the 5 volt range while reading in a hurry. Will be corrected in the next edition. Thanks.
 

Offline lpc32

  • Frequent Contributor
  • **
  • Posts: 461
Re: Multimeter spreadsheet
« Reply #183 on: January 06, 2015, 04:14:53 pm »
The table shows UT139C voltage bandwidth as 10kHz, the manuals say 1kHz. Did someone confirm it to work higher?
 

Offline WytnuclsTopic starter

  • Super Contributor
  • ***
  • Posts: 3045
  • Country: be
Re: Multimeter spreadsheet
« Reply #184 on: January 06, 2015, 04:19:40 pm »
Haven't updated the spreadsheet on page 1 for a while.
Check the latest one:
« Last Edit: May 09, 2017, 10:19:06 pm by Wytnucls »
 

Offline lpc32

  • Frequent Contributor
  • **
  • Posts: 461
Re: Multimeter spreadsheet
« Reply #185 on: January 06, 2015, 08:25:03 pm »
Thanks.
 

Offline WytnuclsTopic starter

  • Super Contributor
  • ***
  • Posts: 3045
  • Country: be
Re: Multimeter spreadsheet
« Reply #186 on: January 12, 2015, 10:03:22 am »
Some Fluke app notes to help you choose a multimeter:
1. ABCs of DMMs - Multimeter features and functions explained http://media.fluke.com/documents/2100079_6003_ENG_C_W.pdf
2. ABCs of multimeter safety http://media.fluke.com/documents/1263690_6003_ENG_J_W.PDF
3. Dual impedance digital multimeters http://media.fluke.com/documents/2718074_0000_ENG_B_W.PDF
4. Why true RMS http://media.fluke.com/documents/1260729_0000_ENG_E_W.PDF
5. How robust is your handheld multimeter? http://media.fluke.com/documents/4100693_0000_ENG_A_W.PDF
6. Multimeter measurements on variable frequency drives http://media.fluke.com/documents/3291974_0000_ENG_A_W.PDF
7. Ten dumb things smart people do when testing electricity http://media.fluke.com/documents/2071940_6003_ENG_B_W.pdf
8. Testing your test leads http://media.fluke.com/documents/3185695_6003_ENG_A_W.pdf
9. How to use Fast/Peak Min/Max on your DMM http://media.fluke.com/documents/2722845_6116_ENG_A_W.PDF
10. Understanding specifications for precision multimeters http://media.fluke.com/documents/2547797_6203_ENG_B_W.PDF
11. Who sets the rules for electrical testing and safety? http://media.fluke.com/documents/2173075_6003_ENG_B_W.pdf
 

Online tautech

  • Super Contributor
  • ***
  • Posts: 31994
  • Country: nz
  • Taupaki Technologies Ltd. Siglent Distributor NZ.
    • Taupaki Technologies Ltd.
Re: Multimeter spreadsheet
« Reply #187 on: January 12, 2015, 07:45:00 pm »
Some Fluke app notes to help you choose a multimeter:
1. ABCs of DMMs - Multimeter features and functions explained http://media.fluke.com/documents/2100079_6003_ENG_C_W.pdf
2. ABCs of multimeter safety http://media.fluke.com/documents/1263690_6003_ENG_J_W.PDF
3. Dual impedance digital multimeters http://media.fluke.com/documents/2718074_0000_ENG_B_W.PDF
4. Why true RMS http://media.fluke.com/documents/1260729_0000_ENG_E_W.PDF
5. How robust is your handheld multimeter? http://media.fluke.com/documents/4100693_0000_ENG_A_W.PDF
6. Multimeter measurements on variable frequency drives http://media.fluke.com/documents/3291974_0000_ENG_A_W.PDF
7. Ten dumb things smart people do when testing electricity http://media.fluke.com/documents/2071940_6003_ENG_B_W.pdf
8. Testing your test leads http://media.fluke.com/documents/3185695_6003_ENG_A_W.pdf
9. How to use Fast/Peak Min/Max on your DMM http://media.fluke.com/documents/2722845_6116_ENG_A_W.PDF
10. Understanding specifications for precision multimeters http://media.fluke.com/documents/2547797_6203_ENG_B_W.PDF
11. Who sets the rules for electrical testing and safety? http://media.fluke.com/documents/2173075_6003_ENG_B_W.pdf
Would it be better to embed these links in the first post?
Easy to find then.  ;)
Avid Rabid Hobbyist
 

Offline WytnuclsTopic starter

  • Super Contributor
  • ***
  • Posts: 3045
  • Country: be
Re: Multimeter spreadsheet
« Reply #188 on: January 13, 2015, 08:46:00 am »
Yeah, was meant to do that. Anyway, it's done now. Thanks for the reminder. Still need to update the front tables (Done).
« Last Edit: January 13, 2015, 10:13:14 am by Wytnucls »
 

Offline _Wim_

  • Super Contributor
  • ***
  • Posts: 1652
  • Country: be
Re: Multimeter spreadsheet
« Reply #189 on: January 13, 2015, 05:36:50 pm »
Some Fluke app notes to help you choose a multimeter:
1. ABCs of DMMs - Multimeter features and functions explained http://media.fluke.com/documents/2100079_6003_ENG_C_W.pdf
2. ABCs of multimeter safety http://media.fluke.com/documents/1263690_6003_ENG_J_W.PDF
3. Dual impedance digital multimeters http://media.fluke.com/documents/2718074_0000_ENG_B_W.PDF
4. Why true RMS http://media.fluke.com/documents/1260729_0000_ENG_E_W.PDF
5. How robust is your handheld multimeter? http://media.fluke.com/documents/4100693_0000_ENG_A_W.PDF
6. Multimeter measurements on variable frequency drives http://media.fluke.com/documents/3291974_0000_ENG_A_W.PDF
7. Ten dumb things smart people do when testing electricity http://media.fluke.com/documents/2071940_6003_ENG_B_W.pdf
8. Testing your test leads http://media.fluke.com/documents/3185695_6003_ENG_A_W.pdf
9. How to use Fast/Peak Min/Max on your DMM http://media.fluke.com/documents/2722845_6116_ENG_A_W.PDF
10. Understanding specifications for precision multimeters http://media.fluke.com/documents/2547797_6203_ENG_B_W.PDF
11. Who sets the rules for electrical testing and safety? http://media.fluke.com/documents/2173075_6003_ENG_B_W.pdf

Thanks for these and your excellent excel file. Much appreciated!
 

Offline TiN

  • Super Contributor
  • ***
  • Posts: 4543
  • Country: ua
    • xDevs.com
Re: Multimeter spreadsheet (38 brands and 210 meters listed)
« Reply #190 on: January 19, 2015, 04:49:10 am »
Few small corrections: Keithley 2001 and 2002 do not have RS232. DB9 is digital portm not serial
YouTube | Metrology IRC Chat room | Let's share T&M documentation? Upload! No upload limits for firmwares, photos, files.
 

Offline WytnuclsTopic starter

  • Super Contributor
  • ***
  • Posts: 3045
  • Country: be
Re: Multimeter spreadsheet (38 brands and 210 meters listed)
« Reply #191 on: January 19, 2015, 06:36:03 am »
Indeed. Will correct database in next edition. Thanks.
 

Offline radiomog

  • Regular Contributor
  • *
  • Posts: 58
  • Country: mc
Re: Multimeter spreadsheet (38 brands and 210 meters listed)
« Reply #192 on: January 27, 2015, 03:08:24 am »
gotta tag this thread to view in more detail later.. looks very interesting. thanks for sharing.
My job is so secret, even I don't know what I'm doing!
 

Offline masster

  • Regular Contributor
  • *
  • Posts: 83
  • Country: ro
Re: Multimeter spreadsheet (38 brands and 210 meters listed)
« Reply #193 on: February 01, 2015, 12:11:20 am »
@Wytnucls
The spreadsheet is far from being easy to read. The 'legend' on top is very ambiguous. For instance, what is the difference between "Important missing feature" and "Essential missing feature" ?? Just semantics. What objective criteria used to separate them? 10000 counts is an " Essential missing feature"? According to who? According to what?

A beginner trying to choose a multimeter will have a hard time deciphering the abbreviations in the table header (Count AC+ BW DUT Volt Accuracy uA A bw m? nS/M? pF mF Dio ?Hz % W T° PC Kit Int Log Clock Disp Light 'I'''I' M/m Peak Hld dB LoZ VFD Evt Batt Life F G E Pwr Jack Fuse P/F 4-20 NCV Price). You should insert comments to all of them on every sheet.

The colors: What objective criteria is used to differentiate it from a white cell?

The stars: Very generic classification, based only on brand.

The spreadsheet might have a meaning to you, but you don't seem to know how to make it easy to understand at a glance and useful to somebody other than you.



« Last Edit: February 01, 2015, 12:41:45 am by masster »
 

Offline WytnuclsTopic starter

  • Super Contributor
  • ***
  • Posts: 3045
  • Country: be
Re: Multimeter spreadsheet (38 brands and 210 meters listed)
« Reply #194 on: February 01, 2015, 07:12:00 am »
Like Dave says often, it is hard to please everybody. I spent a lot of time putting all this together. Initially, it was for my own use. I decided to publish it when I realized it could be useful to someone else too.
Of course, the listing is biased to reflect my preferences for an ideal multimeter, based on my experience. It may not always coincide with your desires. Just ignore all the cell coloring in that case.
Unfortunately, there is no other free database that I'm aware of, unless you decide to make a better one, with your own biases.
« Last Edit: February 01, 2015, 07:59:28 am by Wytnucls »
 

Offline WytnuclsTopic starter

  • Super Contributor
  • ***
  • Posts: 3045
  • Country: be
Re: Multimeter spreadsheet (38 brands and 210 meters listed)
« Reply #195 on: February 01, 2015, 05:09:11 pm »
I'll answer some of your criticisms for the benefit of all users:

'The spreadsheet is far from being easy to read. The 'legend' on top is very ambiguous. For instance, what is the difference between "Important missing feature" and "Essential missing feature" ?? Just semantics. What objective criteria used to separate them? 10000 counts is an " Essential missing feature"? According to who? According to what?'
[/i]
There is a difference between important (really nice to have) and essential (show stopper).
The database is slanted towards electronic work, not electrical chores. A high count and accuracy is desirable for at least one meter, if you have several, like you should have. For the simple reason, that it can serve as a good reference for the others, since calibration is so expensive and often out of reach for hobbyists. Those meters are usually better made, have larger AC bandwidth, are built with proper external voltage references and have little drift.
10,000 count is not highlighted amber per se, unless the count falls below 7,000 for other functions. It doesn't mean that those meters are useless, just that they are not ideal for electronics, but could be used as secondary meters.

A beginner trying to choose a multimeter will have a hard time deciphering the abbreviations in the table header
Every one of them is well explained in the first row, with annotations. These annotations change regularly and are not repeated for each brand, to limit database maintenance time.

The stars: Very generic classification, based only on brand.
The stars are based on brands only, for a general guideline. I don't own enough of the meters in the database for objective individual assessments. Look for independent reviews on this site for any particular meter.
« Last Edit: February 06, 2015, 02:00:53 pm by Wytnucls »
 

Online tautech

  • Super Contributor
  • ***
  • Posts: 31994
  • Country: nz
  • Taupaki Technologies Ltd. Siglent Distributor NZ.
    • Taupaki Technologies Ltd.
Re: Multimeter spreadsheet (38 brands and 210 meters listed)
« Reply #196 on: February 01, 2015, 09:00:46 pm »
Like Dave says often, it is hard to please everybody.

Unfortunately, there is no other free database that I'm aware of, unless you decide to make a better one, with your own biases.
+1
There's more than enough info and what is missing....fill it in for oneself from the available info elsewhere.
As Wytnucls says, it is a living document and change is never ending.

The biggest worry is WHO will pick it up and continue when Wytnucls can no more.  :-//
Hats off to you, it's a great piece of work.  :clap:
Avid Rabid Hobbyist
 

Offline Esthan

  • Contributor
  • Posts: 11
  • Country: pl
Re: Multimeter spreadsheet (38 brands and 210 meters listed)
« Reply #197 on: February 02, 2015, 06:59:46 am »
Hi,

first of all a warm welcome from Poland to all EEVblog forum users.

I have a question regarding the Kyoritsu products.

In the filter section of this magnificent xls document, all of the Kyoritsu product get 5 Stars just like Fluke, Gossen etc - do I get it right that their products are, on average top quality or safe in their price range just like Gossen or Fluke?

Tried to find some in depth reviews or teardowns of the KEW1012 (or any Kyoritsu) that sparked my interest and there is little to no info...
 

Offline WytnuclsTopic starter

  • Super Contributor
  • ***
  • Posts: 3045
  • Country: be
Re: Multimeter spreadsheet (38 brands and 210 meters listed)
« Reply #198 on: February 02, 2015, 07:21:37 am »
Their cheaper meters may not be made in Japan, so quality may vary. Check also Yokogawa for similar meters:
http://tmi.yokogawa.com/products/portable-and-bench-instruments/digital-multimeters/

« Last Edit: February 02, 2015, 07:30:13 am by Wytnucls »
 

Offline Esthan

  • Contributor
  • Posts: 11
  • Country: pl
Re: Multimeter spreadsheet (38 brands and 210 meters listed)
« Reply #199 on: February 02, 2015, 07:23:57 am »
Thanks.

So how am I to understand those 5 Stars in the filter section?
 


Share me

Digg  Facebook  SlashDot  Delicious  Technorati  Twitter  Google  Yahoo