It turned it off, and I noticed that the Scope was somewhat warm. It sits on the top of my table, nothing restricts the "airflow". It wasn't hot, but somewhat warm. After the Scope cooled down it worked like a charm again. 
Had a #1 on a 2236. Did a recap of the power supply, it was gone...
The beam finder on my 22xx oscilloscopes displays as a bright unfocused dot if there is no signal, no sweep, and no readout to be displayed.
It is probably a 2236 who has worked for a while with defective fan.
The -8.6V rail is actually -8.55V The Service manual show me, that it should be between -8.56V and -8.64V, so it missed the target by 10mV.A high pitched noise varying with brightness ajust and overheating is coming from a defective (internal short circuit between turns caused by isolation failure) inductor T906 (120-1439-00 or 01) in the power supply.
For this reason, this defective inductor is no more able to store energy during on state of the mosfet, as it should.
Current in the Mosfet Q9070 became then abnormal with very short on state and high peak value and dissipated power increase. (reason of overheating of the Mosfet)
That's a commun failure of 2235.
It's not occurring with 2236 because it has a fan and the inductor works at lower temperature (better cooling of the power supply)
I installed a fan in my 2235 for this reason.
I already had three 2235 oscilloscopes with this problem.
I have not disassembled and uncoiled the defective T906 inductors to find which winding presented the default because I did not see the usefulness of this information.
The inductance is not repairable anyway and it seems quite difficult to make one yourself because it is very compact.
NB; The ferrite core must have a gap.
The best way to verify if T906 is defective or not is to feed the scope through an isolating transformer and to look with another scope if the waveform of current on R907 (0.2R) is correct or not. (see waveform 39 - 150mV peak)
I'm back again. I've finished with my job, so now I and my 2235 have some "free" time for each other. I disassembled her and she's currently sitting on my workbench while completely exposed. Now I'm waiting for her to warm up. While she's cool I mad some measurements and all of her power rails are spot on except one.The -8.6V rail is actually -8.55V The Service manual show me, that it should be between -8.56V and -8.64V, so it missed the target by 10mV.
I looked at the caps too and they looks OK, but they are quite old, so I'm going to change them. When I'm finished with the warm power measurements I'm going to measure the sizes of the caps. They are no good, if they can't fit in.
Just in case I want to ask something. Do you think, that the +8.6V or the +5V rail can handle the power requirement of an extra small (~8cm) PC fan? I'm pretty sure they can, but just in case I want to know your opinion.
it is difficult to distinguish measurement mistakes from reality and having someone work on the primary side of an off-line circuit is challenging enough


Quoteit is difficult to distinguish measurement mistakes from reality and having someone work on the primary side of an off-line circuit is challenging enoughYou are kidding ?
To work on the primary side of a SMPS using an isolating transformer is challenging ?
I do this everyday....
Measurement mistakes ?
If someone is not able to use a scope and compare with pictures of the service manual, better not to try repair a scope.
When inductor T906 (120-1439-00 or 01) is faulty, waveform of current on R907 (0.2R) is very different from the correct value.
It is not challenging at all, there is no way to make a mistake.
When inductor T906 (120-1439-00 or 01) is faulty, waveform of current on R907 (0.2R) is very different from the correct value.
It is not challenging at all, there is no way to make a mistake.
You may want to check the 2236 schematic
QuoteYou may want to check the 2236 schematic
Please use 2236A (improvement in this respect over 2236) schematic as an example and use a common low power fan.