Author Topic: Filtering signal from transducer  (Read 846 times)

0 Members and 1 Guest are viewing this topic.

Offline spintinasTopic starter

  • Contributor
  • Posts: 13
  • Country: at
Filtering signal from transducer
« on: February 25, 2018, 03:59:16 pm »
Hello dear colleagues, i have some questions for you about signal filtering. I want to project daq for my pressure transducer. Firstly, I have a signal from transducer (voltage about 30-60mV), then i design differential amplifier with gain of 20, and the output voltages transducer now are 500mV-1,1V, but when get a signal in osciloscope, there are a lot of noise. So, the question is, what filter is best to use for this type of signal (pressure transducer), also I want to know when it is better to filter signal? Before amplification or after differential amplifier?. I read some information on the internet and I understood, that one of the option for this situation is passive low pass (RC) filter, but I didnt know what value of cutoff frequency should be, and also low pass filter will attenuate my useful signal. So, guys, I would be thankful if you suggest some solutions.

Transducer datasheet: http://technel.com/Files/AB-HP.pdf

Differential amp schematics is attached, i use LM324, instead LMC6484.
 

Offline danadak

  • Super Contributor
  • ***
  • Posts: 1875
  • Country: us
  • Reactor Operator SSN-583, Retired EE
Re: Filtering signal from transducer
« Reply #1 on: February 25, 2018, 04:09:40 pm »
1) You have a differential amplifier whose performance is highly
dependent on resistor accuracy (ratio accuracy) to reject common
mode noise. I would advise you use an IA vs roll your own. This
is an excellent treatise on the use of IAs.

http://www.analog.com/en/education/education-library/dh-designers-guide-to-instrumentation-amps.html

2) Generally a low pass filer is desired, you trade off its BW versus settling time
when you design. Use TI filter designer, or Analog Devices, to do the design.


http://www.ti.com/tool/filterpro&DCMP=hpa_amp_general&HQS=NotApplicable+OT+filterpro


http://www.analog.com/designtools/en/filterwizard/


What are you doing with the result ? Here is a digital filter and differential A/D approach to
a bridge sensor problem. Note its shown as single ended, differential is just a setting. A/D is
20 bits so can look at a bridge directly in most applications. Cypress PSOC 5LP family.





Here is the board you would use to proto -



http://www.cypress.com/documentation/development-kitsboards/cy8ckit-059-psoc-5lp-prototyping-kit-onboard-programmer-and



Regards, Dana.
« Last Edit: February 25, 2018, 04:19:26 pm by danadak »
Love Cypress PSOC, ATTiny, Bit Slice, OpAmps, Oscilloscopes, and Analog Gurus like Pease, Miller, Widlar, Dobkin, obsessed with being an engineer
 