TiN had asked me about the 40K reference (R207) on the A1 board installed in my early unit.
It is marked "VISHAY B 8822 40K000" and "0.1% 0699 -1125" - pics attached.
On my original unit which is 1990-1991 time period the resistor is marked "VISHAY 89012 301031" and "40K000 0.1%".
While I had it open installed I performed Service Note 3458A-01C which is to replace capacitors C142 and C143 on the A3 board with 2.2 uF units which improve linearity at very low levels of input signal.
I replaced the fans in both units with a Papst model # 622L.
I also came across a resistor (R403) on the A3 board that had never been properly installed. One leg was stuck under another resistor instead of being soldered in its proper place. They both go to the same place however so it is fully functional.
They are both warming up again.
TheSteve, at the price you can find them, three 3458 seems essential, otherwise you have the 'two clocks ' problem! HiHi.
Your latest one came out before mine 2823A050xx. (mine also NVRAM failure but 15yo Dallas, I was able to scavenge cal data using TL866) but no where near the bargain price!
Robert
The price was just luck - the first one I came across, the second a friend found and messaged me about. Both were bought with best offers, it is amazing nobody else clicked buy-it-now at a higher price - especially after seeing the price two broken/damaged ones closed for today.
The odds of two were slim, a third doesn't seem likely.
The new fans are working much better then the originals, much more air flow - even though both 3458A's are fully warmed up the CAL 72 values have changed. I'll log both of them to see how the drift looks. I need to find a darned GPIB cable - tough to find locally.
resistor (R403) on the A3 board that had never been properly installed.
Lucky bodge with resistor
I had cap on A6 with same problem in my 3245A :
Good old days, when you can see assembly problems with naked eye. Those days are gonsky, with modern 01005 SMDs and BGAs
Congratulations
Does anyone know if it is possible to get a U180 without buying a A3 Board?
...
The new fans are working much better then the originals, much more air flow - even though both 3458A's are fully warmed up the CAL 72 values have changed. I'll log both of them to see how the drift looks. I need to find a darned GPIB cable - tough to find locally.
With a considerably stronger fan one might consider an extra shield/cover over the reference module, to keep the temperature there at a similar slightly warmer level and avoid to much turbulence.
The odds of two were slim, a third doesn't seem likely.
The rule seems to be:
The first one you have to find and it can be hard.
Then they start coming to you very easily.
Congratulations on your second one.
Very impressive, of what is possible.
With a considerably stronger fan one might consider an extra shield/cover over the reference module, to keep the temperature there at a similar slightly warmer level and avoid to much turbulence.
I don't think the new fan has a higher CFM then the original, they are just ~25 years newer and as such blow a little better. Looking at the Keysight website the fan they currently spec/sell is the exact fan I used, the Papst 622L. With that in mind I don't think it would be a good idea to mess with shields/covers.
With the new fans installed in both units I am logging the CAL 72 values daily.
Date/time 3458A serial 5XXX 3458A serial 4XX
Day 0, Dec 30, 19:00 .989773311 32.1C .988870057 31.4C
Day 1, Dec 31, 19:00 .989773289 32.2C .988870101 31.5C
Day 2, Jan 1, 19:00 .989773232 32.2C .988869970 31.6C
Day 3, Jan 2, 19:00 .989773424 32.2C .988870044 31.7C
Day 4, Jan 3, 19:00 .989773419 32.5C .988869978 32.1C
Day 5, Jan 4, 19:00 .989773351 32.2C .988869669 31.8C
Day 6, Jan 5, 19:00 .989773242 31.5C .988869148 31.0C
Day 7, Jan 6, 19:00 .989773382 31.9C .988869253 31.5C
Total PPM shift 0.010 PPM/day 0.116 PPM/day