Reparation Hameg HM604-2
Posted by
skopg
on 29 Jul, 2023 07:38
-
Good evening everyone, I
I am new and new to electronics...
I come to you because I have 1 Hameg HM 604-2, and it does not work anymore... I explain I turn on there is a signal on the screen (horizontal line that I can only move in X), if I switch Off Search Delay on Del TRIG there too the lamp lights up (photo1).
If I test a component, there is the L-shaped signal on the screen, juntion BC or even in the M17 604-2 manual (photo2).
But if I turn or press any button NOTHING happens...
Can you help me and tell me what component(s) need to be changed?
Thank you for your feedback.
-
#1 Reply
Posted by
RAPo
on 29 Jul, 2023 08:19
-
Set knobs to default position as described in the manual:
-all pushbuttons in out position
-the three knobs for variable control with the white arrow to calibrated/dent position
-all knobs with markers to mid position (vertical up)
-level knob on left stop AT
-the 3 lever switches to upper most position
-coupling switches for chI & Ch.II to GND
Adjust Y-pos.i an d x.pos to center the trace.
Report back.
set channel I to AC coupling
apply the test signal to channel 1
Please report back what you'll get on the crt.
-
#2 Reply
Posted by
skopg
on 29 Jul, 2023 16:18
-
Hello,
I come back to you after doing what you asked me to do.
On photo1 start-up, on photo2 after 10 sec; Pressed the button on channel 1, set pos Y and X.
Then switched to AC (the signal does not move), but if I restart the button of Channel1 ...
There the signal goes up a box (photo3).
Thank you for telling me what to do next.
ElectroniquementVĂ´tre
Skopg
-
#3 Reply
Posted by
skopg
on 29 Jul, 2023 16:45
-
Hello, I come back to you after doing what you asked me to do.
On photo1 start-up,
on photo2 after 10 sec; Pressed the button on channel 1, set pos Y and X.
Then switched to AC (the signal does not move), but if I restart the button of Channel1 ...
There the signal goes up a box (photo3).
Thank you for telling me what to do next.
electroniquementvĂ´tre
-
#4 Reply
Posted by
RAPo
on 29 Jul, 2023 16:59
-
We are making progress.
Could you explain what "but if I restart the button of Channel1" is? Which button did you press?
Is the trace in AC coupling with or without a signal applied? In photo 3 it looks like there is no signal applied to channel 1.
Could you please set the red buttons with the white triangle to a calib/dent setting? The arrow has to point to the right.
Also please turn the hold button completely clockwise.
Set the channel I to .1V/div, time base to 0.2ms and AC coupling.
Set your probe on x10 and pick up the signal from the cal. 2V port.
Do you see a square wave with the upper part at the first upper division (and the lower part at the first lower division) with the square wave in the horizontal middle?
-
#5 Reply
Posted by
skopg
on 29 Jul, 2023 18:49
-
Re,
when I say "I restart button 1", it means that I took it out.
The trace in the AC coupling, a line remains in the center, when I take out the button of channel1 the signal goes up a square.
According to what you asked it appeared two lines, and after a few seconds a square signal came going from left to right and then, again the 2 lines.
-
#6 Reply
Posted by
RAPo
on 30 Jul, 2023 06:57
-
Thanks, this looks good.
Please keep the AC coupling and only show+trigger on channel I settings.
I expected more distance on the lines, but I can see where it comes from: you have 1V/div, I asked for 0.1V/div. No problem.
Could you please use the 0.1V setting?
With this setting, can you shift the trace up/down with Y.Pos I knob?
can you shift left/right with the X.pos knob?
What worries me is that there is no trigger lock. That will be the next point of investigation.
-
#7 Reply
Posted by
skopg
on 04 Aug, 2023 06:59
-
Hello,
fit after this little break... That's what I did.
-
#8 Reply
Posted by
skopg
on 04 Aug, 2023 07:03
-
Hello, fit after this little break...
That's what I did.
Yes I can move up and down and also from left to right
-
#9 Reply
Posted by
RAPo
on 04 Aug, 2023 07:19
-
and it seems that the scope is also triggering!
So what is the problem now?
-
#10 Reply
Posted by
skopg
on 04 Aug, 2023 07:36
-
You talked to me: What worries me is that there is no trigger lock.
This will be the next point of inquiry.
What should I do now.
-
#11 Reply
Posted by
RAPo
on 04 Aug, 2023 10:22
-
Well, I see that the scope is triggering: the led next to 'TRIG' is lit.
So you now have a scope that is working on channel 1 (it shows a trace that can be moved vertically and horizontally and the trace is triggered) that is quite a better situation than we started with ("I explain I turn on there is a signal on the screen (horizontal line that I can only move in X), if I switch Off Search Delay on Del TRIG there too the lamp lights up").
You could test if channel 2 is working (do the same as above, but then for channel 2), but I guess it will be oke. Do you have any more concerns?
-
#12 Reply
Posted by
skopg
on 04 Aug, 2023 12:23
-
RE,
I'm doing this for channel 2: (Set channel 2 to .1V/div, the timebase to 0.2ms and AC pairing.
Set your probe to x10 and pick up the signal from the cal port. 2V).
And I will keep you informed of the situation.
-
#13 Reply
Posted by
RAPo
on 04 Aug, 2023 12:43
-
and be sure to press the ChI/ChII button this time ;-)
-
#14 Reply
Posted by
skopg
on 04 Aug, 2023 15:35
-
I forgot on T344 I have 0.760mv (not easy to access).
Redo on channel2 (I pushed on CH I/II), ok but the Trigger lamp does not light up.
I put back on channel 1 and the Trigger also does not light up???
-
#15 Reply
Posted by
skopg
on 04 Aug, 2023 15:39
-
But I can move down/up and left/right
-
#16 Reply
Posted by
RAPo
on 05 Aug, 2023 07:45
-
That's a bummer. Could you post a photo with the settings where you cannot trigger?
If you set the test signal back to channel I does it show the two lines?
If so press the dual button, set coupling of channel II to gnd do you see three lines (maybe after positioning chII up/down)?
I meant not only the voltage, but is the transistor working? (0.76 is not a bad value).
-
#17 Reply
Posted by
skopg
on 05 Aug, 2023 08:21
-
Hello, I'm looking, for the transistor I just know how to measure but I don't know how to trigger it (what you asked me).
You ask me to press the double button...which...
-
#18 Reply
Posted by
RAPo
on 05 Aug, 2023 08:30
-
For now, please don't focus on the transistor. We are in the process of discovering what works and what does not. After that, we get to the stage of looking at components for the culprit.
We had significant progress by just fiddling with the knobs. It would be great if we could return to a state where channel I displayed the correct info.
-
#19 Reply
Posted by
skopg
on 05 Aug, 2023 08:39
-
-
#20 Reply
Posted by
RAPo
on 05 Aug, 2023 09:05
-
Thanks. I see that some knobs have changed.
Please turn hold off completely clockwise to a minimum.
Set the delay knob to 0.1us
Set the knobs with the white arrow completely to the right in the detent position.
Set time base to 0.2/div
Move the channel II trace up so it is in the middle of the CRT.
Next, set coupling ch I to AC, set coupling Ch II to GND
Set both CHI and II to 0.1V/div
Set the CHI/II button into a depressed state.
Press the dual button in and apply the test signal 2V to channel I, there should be 3 lines on the screen and the Trig LED should be lit.
The middle line should be moveable by the Y-pos II knob.
Do you see them?
If so, reverse the role of channel I and II
Keep the dual button pressed in.
Set Ch I to GND coupling, CH II to AC coupling.
Press the CHI/II button in (this ensures triggering will be on channel II)
Apply the test signal to CH II
Again the three lines should appear on the CRT.
In this case the middle line should be moveable by the Y-pos CHI knob.
-
#21 Reply
Posted by
skopg
on 05 Aug, 2023 10:25
-
-
#22 Reply
Posted by
skopg
on 05 Aug, 2023 10:27
-
-
#23 Reply
Posted by
RAPo
on 05 Aug, 2023 10:41
-
It is not very easy to see, but did you press the chanI/II Trig I/II button in the last photo (canal2sonde.jpg)?
If you do, I'll bet the triggering led comes on and as far as I can see your scope is working fine.
-
#24 Reply
Posted by
skopg
on 05 Aug, 2023 11:54
-
yes I did it