Author Topic: Fluke 8060A Somewhat Functional  (Read 493 times)

0 Members and 2 Guests are viewing this topic.

Offline Christine1953Topic starter

  • Newbie
  • Posts: 6
  • Country: us
Fluke 8060A Somewhat Functional
« on: May 08, 2024, 06:35:08 pm »
Hello Everyone,
This is my first post here!

I have an old 8060A that was given to me many moons ago because it appeared to be DOA, and it's been sitting in my junk box for at least 15 years untouched. When I received it, it probably had less than 20 hours of use, and since we moved on to something else in the shop, it was discarded.

It turns out that if you power it with the cable, it fires right up, although there are some issues. It looks like a nice meter. My question is should I try to repair it, or toss it.

I've worked on lots of equipment, so getting in there isn't a problem, and I have good tools, but I'm not all that familiar with working on test equipment. I was in broadcast, and when the test equipment broke down, we just sent it out for repair

I have the users manual with the schematics and such, and have read some of the blogs/Youtube repair videos on repairing these units, but I could use some help navigating the lay of the land so to speak when it comes to repairing it.

Here's what I know, and what I've done so far. I've got it to fire up. Display is good, passes the ratio test, and the switch logic test. Resistance somewhat works. Accuracy diminishes as the resistance increases. In the M ohm area it's out of whack by at least 20%. The continuity and buzzer button works. DC/AC voltage functions do not work at all, just gibberish, and the meter does not zero out. HiZ DC volts function operates/tracks normally following a variable supply on the 200mv and 2V scale as it should. I have not yet checked the ammeter section. I took the unit apart and it appears very clean on the inside and no capacitors are leaking, but I will do a more through investigation later to confirm. Leaking capacitors seem to be an issue with these units so I'll probably just shotgun them all when I do the repairs.

So to summarize:
Battery has a bad connection to the board.
AC/DC voltage function not working.
Amps, Hz, dB, REL functions are untested.
Voltages at the test points have not been checked.
Unit probably needs a total calibration after the repairs.

If anyone has seen these issues and can point me in the direction of some discussions that address the problems I would most appreciate the help! Also, I see the owners manual mentions several times about being careful about handling the boards because of sensitive components and static discharge, but I regularly see in the videos people with little or no protection handling them. Are they touchy? I always use a wrist strap, work on a wood benchtop, and have a good soldering station.

Well that's it for now, and thanks for your help!
Christine
 

Online J-R

  • Frequent Contributor
  • **
  • Posts: 988
  • Country: us
Re: Fluke 8060A Somewhat Functional
« Reply #1 on: Yesterday at 01:25:42 am »
My first thought is check the 1K fusible resistor R2.

Beyond that, I would suggest checking every function and range since then you will have a really good idea of which part of the schematic is the likely culprit.
 

Online lowimpedance

  • Super Contributor
  • ***
  • Posts: 1248
  • Country: au
  • Watts in an ohm?
Re: Fluke 8060A Somewhat Functional
« Reply #2 on: Yesterday at 02:36:57 am »
As it has been sitting for years without use the switches could have oxide build up , so a good 'exercise' of each button is in order.
Also definitely replace those electrolytic caps.
Odd the battery connection was bad after only 20h operation , check the continuity from the battery tabs to the pcb. perhaps the power switch or the ext power in jack are iffy too.
The odd multimeter or 2 or 3 or 4...or........can't remember !.
 

Offline AVGresponding

  • Super Contributor
  • ***
  • Posts: 4677
  • Country: england
  • Exploring Rabbit Holes Since The 1970s
Re: Fluke 8060A Somewhat Functional
« Reply #3 on: Yesterday at 11:01:53 am »
Getting some deoxit or similar in the switches is for sure a good start. When you compile a BOM for recapping, bear in mind later versions have a separate TRMS daughter-board, and this means 1 less crapacitor to replace.
nuqDaq yuch Dapol?
Addiction count: Agilent-AVO-BlackStar-Brymen-Chauvin Arnoux-Fluke-GenRad-Hameg-HP-Keithley-IsoTech-Mastech-Megger-Metrix-Micronta-Racal-RFL-Siglent-Solartron-Tektronix-Thurlby-Time Electronics-TTi-UniT
 

Offline rsjsouza

  • Super Contributor
  • ***
  • Posts: 5990
  • Country: us
  • Eternally curious
    • Vbe - vídeo blog eletrônico
Re: Fluke 8060A Somewhat Functional
« Reply #4 on: Yesterday at 11:43:38 am »
Welcome to the forum.

Fluke 8060A meters are still quite useful in the field, so definitely don't toss it.

If possible, could you post hi-res photographs of its interior? This could help further with any diagnosis.

As for the symptoms, as others have said switch contacts might be a problem.


Resistance somewhat works. Accuracy diminishes as the resistance increases. In the M ohm area it's out of whack by at least 20%.
I don't know your level of expertise, so that might be useless but it is a common error. When doing hundreds of kΩ or MΩ measurements, please do not touch both leads, so you don't add your body resistance to the measurements.

Vbe - vídeo blog eletrônico http://videos.vbeletronico.com

Oh, the "whys" of the datasheets... The information is there not to be an axiomatic truth, but instead each speck of data must be slowly inhaled while carefully performing a deep search inside oneself to find the true metaphysical sense...
 

Offline rodcastler

  • Regular Contributor
  • *
  • Posts: 99
  • Country: cl
Re: Fluke 8060A Somewhat Functional
« Reply #5 on: Yesterday at 11:57:24 am »
There's this thread that may help as well:
https://www.eevblog.com/forum/testgear/old-fluke-multimeters/
 

Offline Christine1953Topic starter

  • Newbie
  • Posts: 6
  • Country: us
Re: Fluke 8060A Somewhat Functional
« Reply #6 on: Yesterday at 01:18:18 pm »
Thanks everyone!
 I'd say my expertise level is OK. I was a pretty good trouble shooter back in the day, but I've been away from it for a long time, so let's call it "a diamond in the rough". I think the issue with the battery clip is nothing more than a bad connection somewhere along the line. I'm suspecting the switch in the power jack. When I first started playing with it, I put a battery in and nothing. Then I just hit the side of the case with my knuckles, (impact maintenance), and it came to life. I can trace that out, but for now I'll just use the jack for power. I have some deoxit so I'll clean the switches. I can open it up separate the boards and take some pictures to post, and then we can get a cap list together. I probably have most of what I need on hand. And, I'll go ahead and test the rest of the functions, and check R 1 and 2. Give me a day or two on the results.

The meter is in nice shape and I'd like to get it up and running if possible. If I remember correctly, it was quite pricey for the time. Thanks! C-
 

Offline Christine1953Topic starter

  • Newbie
  • Posts: 6
  • Country: us
Re: Fluke 8060A Somewhat Functional
« Reply #7 on: Yesterday at 10:01:27 pm »
There's this thread that may help as well:
https://www.eevblog.com/forum/testgear/old-fluke-multimeters/
Thanks! It's a long thread with lots of history, but I'll work through it. Maybe I need to ping the designer personally! LOL
 

Offline Christine1953Topic starter

  • Newbie
  • Posts: 6
  • Country: us
Re: Fluke 8060A Somewhat Functional
« Reply #8 on: Yesterday at 10:05:59 pm »
First off thanks for the help. And, I'm not used to this style of forum so I probably need some help in posting my progress. ( I'm such a Luddite)! I'll just reply to someones comment in the meantime until I get things under control.
« Last Edit: Yesterday at 10:46:55 pm by Christine1953 »
 

Offline Christine1953Topic starter

  • Newbie
  • Posts: 6
  • Country: us
Re: Fluke 8060A Somewhat Functional
« Reply #9 on: Yesterday at 10:14:40 pm »
I took a few shots of the board and display. The board and the display look great! I see no leaking capacitors (at least from the top view) I inspected the board thoroughly and there is zero corrosion anywhere. It looks like this is rev. H S/N is 3620934. The display is very clean! If we need any pics of a certain area, just let me know.

 

Offline Christine1953Topic starter

  • Newbie
  • Posts: 6
  • Country: us
Re: Fluke 8060A Somewhat Functional
« Reply #10 on: Yesterday at 10:45:24 pm »
I made some progress. Battery clip was bad. Loose rivet. I replaced it thinking that was the problem, but no dice. I'm still suspecting the switch in the side power jack. I cleaned all the switches and that helped. Ohms (all settings) now working and I'm getting reasonable readings. DC readings are now working on the 200mv and 2v positions. 20/200/1000 not working see the picture. I seem to remember seeing something about that in the troubleshooting section. It sounds like this is a common problem. If anyone has some suggestions please let me know. As I said, I'm unfamiliar with the "design language" used in test equipment, so when I look at the schematic it can be a bit puzzling. I'm guessing by the time I'm done I'll have a much better idea of what it is about. And, it doesn't help that I'm somewhat dyslexic LOL! Whenever there are a lot of switches and parallel wires on a schematic it's somewhat problematic for me. I can work through it, but it takes some time.

Also I have some broken buttons. two of the tan ones. The plastic must be brittle and I'll have to glue them in place or something.

I'm not sure how to test the AC functions. All I can think of is to use a variac with a current limiting resistor, and maybe making a voltage divider with some resistors??? I'm pretty unsure there and I really don't like playing around with line voltage. So if anyone has an idea that would be great.

I'm pretty encouraged at this point. After doing a close inspection, the meter is in excellent condition considering it's age. The caps all look good and there's no leakage or corrosion that I can see. I'm planning on replacing all of them as soon as I get things a little more under control.

I'll see if I can the readings from the test points tomorrow, for some more clues. I just ran out of time today. Thanks again for the help!
C-
 

Online tautech

  • Super Contributor
  • ***
  • Posts: 28459
  • Country: nz
  • Taupaki Technologies Ltd. Siglent Distributor NZ.
    • Taupaki Technologies Ltd.
Re: Fluke 8060A Somewhat Functional
« Reply #11 on: Today at 12:47:08 am »
There's this thread that may help as well:
https://www.eevblog.com/forum/testgear/old-fluke-multimeters/
Thanks! It's a long thread with lots of history, but I'll work through it. Maybe I need to ping the designer personally! LOL
You might like to link this thread into the one member rodcastler brought your attention to.
Scroll to the top of page and copy the URL and paste it in a new post there, maybe with a wee mention about this thread. Don't use the first post link but the one higher up:
https://www.eevblog.com/forum/testgear/fluke-8060a-somewhat-functional/

In years to come as others are researching these DMM's it's helpful to have links to other threads.
Avid Rabid Hobbyist
Siglent Youtube channel: https://www.youtube.com/@SiglentVideo/videos
 
The following users thanked this post: Christine1953

Online J-R

  • Frequent Contributor
  • **
  • Posts: 988
  • Country: us
Re: Fluke 8060A Somewhat Functional
« Reply #12 on: Today at 01:07:31 am »
Just to double-check, you're removing the barrel connector when trying it on battery?

A variable transformer would be great for running some AC tests.  For voltage, of course the setup should be self-explanatory, but for current I would use an incandescent light bulb in series with the Fluke as well as another known-good DMM.  You should be able to check all the ranges with that setup.

Some high voltage safety tips are make connections with the power switched off, only use one hand at a time, wear rubber-soled shoes....
 

Offline blue_lateral

  • Contributor
  • Posts: 15
  • Country: us
Re: Fluke 8060A Somewhat Functional
« Reply #13 on: Today at 02:00:54 am »
DONT put deoxit in the switches as someone suggested earlier. That is a mistake. I did that once many years ago. It's apparently too conductive for the high-Z circuitry. It is very hard to get rid of. Several washings with isopropyl did eventually make the meter functional again.

The "old fluke multimeters" thread referenced in rodcastler's post is the one you want.
« Last Edit: Today at 02:14:30 am by blue_lateral »
 