Siglent SDM3055 - stuck on boot screen
Posted by
darryl.lawler
on 02 Jul, 2019 15:28
-
Hello and thank you for providing such a great forum. Lots of useful information throughout the whole site.
I have found two other threads addressing this issue, but one is from 2014 and the other from 2016. I wasn't sure if it was proper to resurrect those threads, so here I am.
I have an SDM3055 stuck on the boot screen that is black with white writing "Siglent-Siglent Technologies."
The power turns on the fan and the DMM attempts to boot, but hangs on the boot screen.
I have downloaded the updated firmware, but I am unable to find instructions for updating without access to the display.
I have contacted support@siglent.com, per the instructions on their "After Sales" webpage.
While I was waiting for a response, I was hoping perhaps someone here may have a suggestion.
TIA for any replies and have a great day!
-
#1 Reply
Posted by
BillB
on 02 Jul, 2019 16:57
-
I had an SDM3045X that had done the same thing, and I was able to recover it using a recovery package provided by tech support.
-
#2 Reply
Posted by
tv84
on 02 Jul, 2019 17:06
-
I think member tautech can help.
-
-
I will try contacting him via PM. Thank you for the suggestion.
-
#4 Reply
Posted by
tautech
on 02 Jul, 2019 20:30
-
I think member tautech can help.
Yes possibly I can.
I will try contacting him via PM.
Got it.
There is a recovery package for the SDM3055 and it's restricted to service centers as we don't want it out in the public arena.
Depending on the FW version installed before the boot freeze the USB package may repair it otherwise the unit must be opened to access the internal SD card slot.
In the first instance contact your local guys in Ohio for help:
North American Headquarters
SIGLENT Technologies America, Inc
6557 Cochran Rd Solon, Ohio 44139
Tel: 440-398-5800
Toll Free:877-515-5551
Fax: 440-399-1211
info@siglent.com
www.siglentamerica.comCome back and tell us how you get on and if there's problems with getting onto good US tech support maybe I can grease the wheels.
-
-
Wanted to follow up on this thread and report that it was a very easy fix. Siglent's support was very responsive and their instructions were very clear.
Anyone who needs to perform this recovery . . . be sure you are using an 8GB or smaller USB stick and be sure to format it to FAT32. Those are really the only important details about the process.
Thank you to all of you who stepped in to assist. Greatly appreciated.
Take care and have a great day!
-
#6 Reply
Posted by
natrium
on 26 Apr, 2020 21:21
-
Hello,
I've got the same problem with my SDM3055 as darryl.lawler. It suddendly refused to boot and stucks at the bootscreen.
I already contacted my local sales office here in germany but didnt receive an answer. I read about the recovery package here in the forum. Is there any change to get this package?
Greetings from germany.
-
#7 Reply
Posted by
tautech
on 27 Apr, 2020 08:20
-
Hello,
I've got the same problem with my SDM3055 as darryl.lawler. It suddendly refused to boot and stucks at the bootscreen.
I already contacted my local sales office here in germany but didnt receive an answer. I read about the recovery package here in the forum. Is there any change to get this package?
Greetings from germany.
Welcome to the forum.
Oh dear that's no good but Siglent Hamburg should fix you up.
If not send me your email via a PM and I can share the tools we have to fix this.
For others reading......keep your firmware up to date !
-
#8 Reply
Posted by
iamsam
on 14 May, 2020 20:23
-
Hi everybody ! I'm new to the forum. I have the same siglent SDM3055 stuck in boot logo. I've only used it twice and boxed it for months. Now my workbench is finally finished , and here I am
I tried to contact Siglent America , but did not get any reply. Hoping somebody could help me here.. Thanks !
-
#9 Reply
Posted by
tautech
on 14 May, 2020 21:51
-
Hi everybody ! I'm new to the forum. I have the same siglent SDM3055 stuck in boot logo. I've only used it twice and boxed it for months. Now my workbench is finally finished , and here I am I tried to contact Siglent America , but did not get any reply. Hoping somebody could help me here.. Thanks !
Welcome aboard.
I have the recovery tools so send me a PM with your email and we'll sort your problems out.
I need pop out for a bit so it might be a few hours before I'm in touch.
-
#10 Reply
Posted by
iamsam
on 15 May, 2020 00:02
-
Yey ! Thank you ! tautech !! will pm u.
-
-
What causes this? How often does it occur? Seems like Siglent is aware of it - any chance for a more permanent firmware fix?
-
#12 Reply
Posted by
tautech
on 15 May, 2020 04:07
-
What causes this?
Old FW not updated.
How often does it occur?
At random.
Seems like Siglent is aware of it - any chance for a more permanent firmware fix?
Current FW doesn't have boot freeze issues AFAIK.
-
-
What causes this?
Old FW not updated.
How often does it occur?
At random.
Seems like Siglent is aware of it - any chance for a more permanent firmware fix?
Current FW doesn't have boot freeze issues AFAIK.
Cool - Thanks
-
#14 Reply
Posted by
iamsam
on 15 May, 2020 14:22
-
Hi Guys !
My : Siglent SDM3055 is up and running again
, thanks to tautech
-
#15 Reply
Posted by
tautech
on 15 May, 2020 14:31
-
Hi Guys !
My : Siglent SDM3055 is up and running again , thanks to tautech
No worries, happy to help when I can and glad you got it sorted with just USB recovery which indicates it's a newer unit or had the FW already updated to allow for USB recovery.
Enjoy.
-
#16 Reply
Posted by
tv84
on 15 May, 2020 14:40
-
tautech is the "SDM3055 Master Resurrector"!
-
#17 Reply
Posted by
tautech
on 15 May, 2020 14:45
-
tautech is the "SDM3055 Master Resurrector"!
I can do a few other ones too.....but the funny thing is never one I've sold, not ever !
Anyways, no way I can compete with your accomplishments !
-
#18 Reply
Posted by
tv84
on 15 May, 2020 14:51
-
Anyways, no way I can compete with your accomplishments !
No competition here, just us helping each other member.
-
#19 Reply
Posted by
advark
on 27 Jul, 2020 22:55
-
Hi everyone,
Just to add my 2 cents to the conversation. I just got my brand new SDM3055. Really happy. Opened the box, checked the manual, plugged the beast, switched the power on and... can't get pass the splash screen!
From what I found so far, it seems to be a known issue for the past... 4 years!!! Are you kidding me? Don't know when mine was "built" but "for those playing along at home" (pun intended!) S/N is SDM35FAC3R0271. According to Siglent's web site the lastest FW is from Nov 2019 and the previous one is from April 2018.
I'm so not impressed.... First Siglent gear I ever bought and I'm probably not gonna get another one. I ordered this gear along with a with a SDG1032X which I haven't received yet because it is BO for next 2-3 week and I'm wondering if I'm not gonna cancel that.
-
#20 Reply
Posted by
tautech
on 27 Jul, 2020 23:22
-
Hi everyone,
Just to add my 2 cents to the conversation. I just got my brand new SDM3055. Really happy. Opened the box, checked the manual, plugged the beast, switched the power on and... can't get pass the splash screen!
From what I found so far, it seems to be a known issue for the past... 4 years!!! Are you kidding me? Don't know when mine was "built" but "for those playing along at home" (pun intended!) S/N is SDM35FAC3R0271. According to Siglent's web site the lastest FW is from Nov 2019 and the previous one is from April 2018.
I'm so not impressed.... First Siglent gear I ever bought and I'm probably not gonna get another one. I ordered this gear along with a with a SDG1032X which I haven't received yet because it is BO for next 2-3 week and I'm wondering if I'm not gonna cancel that.
Drop me a PM with your email address and I can send you the recovery package.
That unit may have old firmware installed which was notorious for boot freezes and later versions seem to have eliminated this.
-
#21 Reply
Posted by
advark
on 29 Jul, 2020 20:14
-
Hey everyone,
My SDM3055 is ressurected
. Thanks to Tautec
-
#22 Reply
Posted by
hugo
on 30 Jul, 2020 21:17
-
tautech is the "SDM3055 Master Resurrector"!
I can do a few other ones too.....but the funny thing is never one I've sold, not ever !
Anyways, no way I can compete with your accomplishments !
What goes around comes around ... it's good karma ...
-
#23 Reply
Posted by
DSoldano
on 01 Aug, 2020 12:38
-
I now have the same problem. My SDM3055 unit is about 5 years old, worked like a charm up until now. Went to use it yesterday afternoon and it won't get past the boot screen. Anyway I can get the recovery package? I'd hate to wait until Monday to get in touch with the guys in Ohio.
Thanks
Dominick
-
#24 Reply
Posted by
klausES
on 01 Aug, 2020 14:33
-
Certainly immediately when Tautech is online again and sees this here.
It won't be long before he's online.
-
#25 Reply
Posted by
tautech
on 01 Aug, 2020 19:29
-
I now have the same problem. My SDM3055 unit is about 5 years old, worked like a charm up until now. Went to use it yesterday afternoon and it won't get past the boot screen. Anyway I can get the recovery package? I'd hate to wait until Monday to get in touch with the guys in Ohio.
Thanks
Dominick
Check your PM's and email.
-
#26 Reply
Posted by
DSoldano
on 02 Aug, 2020 13:23
-
Problem solved - a great round of applause for TauTech - he was extremely helpful (and over a weekend too). It is wonderful to see that such great service exists.
Thank you mate, I am one happy camper, unit is back up and running.
Dominick
-
-
Hi,
I have 3 years old oscilloscope Siglent SDS1052DL + and after starting the logo freezes and does not work. Can I ask for recovery packages?
-
#28 Reply
Posted by
tautech
on 06 Sep, 2020 21:17
-
Hi,
I have 3 years old oscilloscope Siglent SDS1052DL + and after starting the logo freezes and does not work. Can I ask for recovery packages?
PM received and returned. Check your inbox.
-
#29 Reply
Posted by
tautech
on 07 Sep, 2020 06:12
-
Hi,
I have 3 years old oscilloscope Siglent SDS1052DL + and after starting the logo freezes and does not work. Can I ask for recovery packages?
PM received and returned. Check your inbox.
Another fixed.
Keep your firmware up to date guys and gals.
-
#30 Reply
Posted by
wdecho
on 02 Nov, 2020 14:40
-
I have the same problem, Siglent 1052DL+ stuck on boot screen. The scope was out of warranty 2 months when this happened. Tried the math thing without any results.
-
#31 Reply
Posted by
tautech
on 02 Nov, 2020 18:37
-
I have the same problem, Siglent 1052DL+ stuck on boot screen. The scope was out of warranty 2 months when this happened. Tried the math thing without any results.
Welcome to the forum.
Drop the chaps in Ohio an email and they should supply you with a recovery package.
info@siglentna.com
If no joy drop me a PM with your email address.
-
#32 Reply
Posted by
Aph
on 10 Nov, 2020 20:37
-
Wow, sounds like a real common problem!
I too, have one of these (Siglent 1052DL+) just a few months beyond the warranty period.
The Math switch procedure does NOT work. The recovery file does NOT work (got the file right from Siglent).
Does anyone have the correct voltages for the multi-cable that goes from the power supply to the main board?
If it's not that, I will give up... not going to try troubleshooting the main board. I'll just buy another scope... and for sure a different brand!
-
#33 Reply
Posted by
tautech
on 10 Nov, 2020 21:00
-
Wow, sounds like a real common problem!
I too, have one of these (Siglent 1052DL+) just a few months beyond the warranty period.
The Math switch procedure does NOT work. The recovery file does NOT work (got the file right from Siglent).
Does anyone have the correct voltages for the multi-cable that goes from the power supply to the main board?
If it's not that, I will give up... not going to try troubleshooting the main board. I'll just buy another scope... and for sure a different brand!
Welcome to the forum.
Exactly which recovery package have you got ?
USB or SD card ?
Have you got back to Siglent (Ohio I suspect) telling them the recovery didn't work ?
-
#34 Reply
Posted by
Aph
on 11 Nov, 2020 21:39
-
Wow, sounds like a real common problem!
I too, have one of these (Siglent 1052DL+) just a few months beyond the warranty period.
The Math switch procedure does NOT work. The recovery file does NOT work (got the file right from Siglent).
Does anyone have the correct voltages for the multi-cable that goes from the power supply to the main board?
If it's not that, I will give up... not going to try troubleshooting the main board. I'll just buy another scope... and for sure a different brand!
Welcome to the forum.
Exactly which recovery package have you got ?
USB or SD card ?
Have you got back to Siglent (Ohio I suspect) telling them the recovery didn't work ?
Siglent (Ohio) sent me a link for three files to put on a USB stick, attach it to the front of unit, and reboot, etc.
It did not work. When I told them it didn't work, they said I could buy and swap the main board or buy a new unit (both options about the same price!).
Thanks.
-
#35 Reply
Posted by
tautech
on 11 Nov, 2020 22:03
-
Wow, sounds like a real common problem!
I too, have one of these (Siglent 1052DL+) just a few months beyond the warranty period.
The Math switch procedure does NOT work. The recovery file does NOT work (got the file right from Siglent).
Does anyone have the correct voltages for the multi-cable that goes from the power supply to the main board?
If it's not that, I will give up... not going to try troubleshooting the main board. I'll just buy another scope... and for sure a different brand!
Welcome to the forum.
Exactly which recovery package have you got ?
USB or SD card ?
Have you got back to Siglent (Ohio I suspect) telling them the recovery didn't work ?
Siglent (Ohio) sent me a link for three files to put on a USB stick, attach it to the front of unit, and reboot, etc.
It did not work. When I told them it didn't work, they said I could buy and swap the main board or buy a new unit (both options about the same price!).
Thanks.
Really ?
Contact them again and tell them you need the SDS1000DL+ SD card recovery package.
If they won't play ball drop me your email by PM and we'll get you all fixed up.
-
#36 Reply
Posted by
Aph
on 12 Nov, 2020 16:34
-
Okay, thanks for the info.
I have just emailed them today (can't find a phone number).
We will see what their response is... though they do take their time to respond.
-
#37 Reply
Posted by
Aph
on 20 Nov, 2020 17:20
-
Hey, Rob,
Thank you for all the help.
My SDS1052DL+ seems to be working fine now!
-
#38 Reply
Posted by
hdibani
on 02 Dec, 2020 17:20
-
Hi I am having the same issue with an SDS 1102CML+
Can someone help with the recovery package?
Thanks
-
#39 Reply
Posted by
hdibani
on 03 Dec, 2020 10:05
-
Another happy Siglent DSO 1102CML+ owner here.
Rob from Tautech is an absolute hero. He helped me recover my DSO.
Thanks a lot Rob for all the help!!
-
#40 Reply
Posted by
cnkz
on 06 Feb, 2021 13:53
-
I own a SDM3055 with
SW Version: 1.01.01.16R2
HW Version: 02-00-00-12-00
I just wanted to make a FW update and found the front USB not working. It is the second time i am using that USB interface, i had already done one FW update before. Therefore i guess it is a FW issue?
I disassembled the device already but there is no obvious issue with the front USB, all 4 wires are working from front-connector to the main-PCB.
I am wondering if an SD Card recovery might help?
-
#41 Reply
Posted by
tautech
on 06 Feb, 2021 20:18
-
I own a SDM3055 with
SW Version: 1.01.01.16R2
HW Version: 02-00-00-12-00
I just wanted to make a FW update and found the front USB not working. It is the second time i am using that USB interface, i had already done one FW update before. Therefore i guess it is a FW issue?
I disassembled the device already but there is no obvious issue with the front USB, all 4 wires are working from front-connector to the main-PCB.
I am wondering if an SD Card recovery might help?
It could be the USB stick you use.
The SDM should first recognize a USB stick is connected and place a USB symbol on the display.
Does it do this ? If not check with other USB sticks.
USB sticks for FW updates need be formatted in FAT32 and 8GB or smaller.
Copy the FW update .ADS file to the root of the drive and select it from the file manager and Open/Run it.
-
#42 Reply
Posted by
Vinylitic
on 28 Feb, 2021 19:00
-
Hello TO All... New member here...
Tautech... I’m hoping you can help me... I have a Siglent SDM3055 stuck on the boot screen. It was last calibrated March-03-20, so I am assuming it has a most recent firmware. I understand you have helped many members here on eevblog. Although, I am am a new member, I was hoping you could help me as well with a Recovery Package..? Your assistance would be greatly appreciated.
Thank you.
-
#43 Reply
Posted by
tautech
on 28 Feb, 2021 19:23
-
Hello TO All... New member here...
Tautech... I’m hoping you can help me... I have a Siglent SDM3055 stuck on the boot screen. It was last calibrated March-03-20, so I am assuming it has a most recent firmware. I understand you have helped many members here on eevblog. Although, I am am a new member, I was hoping you could help me as well with a Recovery Package..? Your assistance would be greatly appreciated.
Thank you.
Welcome to the forum.
PM received and returned and email sent.
-
#44 Reply
Posted by
Vinylitic
on 28 Feb, 2021 22:40
-
-
-
Now I have the same problem - finding my way here I see hope. PM sent to tautech and email off to Siglent...
-
#46 Reply
Posted by
AlexDel
on 05 Mar, 2021 10:01
-
Hi,
I have an SDM3045X with the same problem - OS crashes, freezes on splash screen.
Can you help me?
Thank you very much.
-
-
Hello,
I've got the same problem with my SDM3055 as darryl.lawler. It suddendly refused to boot and stucks at the bootscreen.
I already contacted my local sales office here in germany but didnt receive an answer. I read about the recovery package here in the forum. Is there any change to get this package?
Greetings from germany.
My siglent stays in the logo, it is sdm3055, who can help me, I live in Uruguay, how can I recover it?
Welcome to the forum.
Oh dear that's no good but Siglent Hamburg should fix you up.
If not send me your email via a PM and I can share the tools we have to fix this.
For others reading......keep your firmware up to date !
-
#48 Reply
Posted by
aheid
on 22 Apr, 2021 21:43
-
Seems my SDM 3055 joined the crowd
I bought the unit in february in 2019, and it had the most up-to-date firmware then (I distinctly recall checking). I'm also fairly sure I updated the firmware at least once since then, which leads me to suspect I have at least 1.01.01.16R2 on it. However I'm not 100% certain.
edit: never mind, didn't fully read the flashing instructions. Seems I need recovery package then?
-
#49 Reply
Posted by
aheid
on 24 Apr, 2021 13:40
-
Seems my SDM 3055 joined the crowd
Thanks to some prompt and excellent assistance from tautech, the unit is now alive again.
I ended up having to do the internal SD-card recovery route, not ideal. However, having gotten it back alive I noticed that the DMM did not recognize my USB drive!
In Windows everything was just fine, and I repartitioned it several times, including a full "plain" partition (FAT32 all times).
edit: after trying to use my brain for a bit, it hit me. The pen drive was using GPT. I followed this[1] guide (using diskpart, GUI option was not available) to turn it back to MBR and now the DMM recognizes the USB drive.
To upgrade the firmware I ended up using the same SD-card I used for recovery, but with a USB card reader. Somehow it wouldn't let me jump straight from the .19 version (from the recovery image) to the latest .22R1 version, had to do a trip via .20 first.
[1]:
https://docs.microsoft.com/en-us/windows-server/storage/disk-management/change-a-gpt-disk-into-an-mbr-disk
-
#50 Reply
Posted by
Joe510070
on 28 Nov, 2021 19:48
-
I’m having a similar problem but with only a dark display. I posted the details in the repair section of this forum. Hopefully there is hope for me too!
-
#51 Reply
Posted by
JimLev
on 16 Jan, 2022 00:18
-
First post. I found this forum yesterday after a quick search. My 18 month old Siglent SDM3055 was stuck on the boot screen.
I sent Tautech a PM and almost instantly he replied back, thanks again Tautech.
He emailed me what I needed to fix my meter and to upgrade the firmware to the latest version.
Great forum, I’ll have to cruise around and see what else I can find.
-
#52 Reply
Posted by
tautech
on 16 Jan, 2022 09:42
-
To all those following, latest firmware should put an end to all this as the important feature in it is:
Optimize flash R&W design.Anyone with a SDM model please install the latest version.
Pick the version to match your model from here:
https://int.siglent.com/download/firmwares/?CateIdss=5
-
#53 Reply
Posted by
Neox_Dev
on 03 Feb, 2022 12:36
-
Hi guys, the same happened here with my SDM3045X does anyone could provide me the recovery tool please?
-
#54 Reply
Posted by
tautech
on 04 Feb, 2022 06:53
-
Hi guys, the same happened here with my SDM3045X does anyone could provide me the recovery tool please?
Welcome to the forum.
Received your PM and will contact you when back from holiday next week.
-
#55 Reply
Posted by
MWilliams
on 10 Feb, 2022 04:15
-
New to the forum with a new Siglent SDM3055X and right out of the carton it powers up and freezes on the splash screen. I've read in this posting that the new firmware no longer does this so I'm guessing this unit must have been sitting in an Amazon warehouse for quite while. If I can get that repair/fix it would be most appreciated.
-
#56 Reply
Posted by
Kleinstein
on 10 Feb, 2022 06:37
-
If it is dead right out of the box, this would be something to send back for an exchange on warranty.
-
#57 Reply
Posted by
idris5416
on 25 Feb, 2022 23:47
-
Hi everyone
i got my brand new SDM3055 and press on button then dada siglent logo, almost 2h still there, looks like stuck. Any possible way to recover?
Thank you.
-
#58 Reply
Posted by
farj
on 04 Mar, 2022 19:29
-
Hello everybody,
I’m a brand new member.
I had the same renowned problem with my SDM3055x and member Tautech (Rob) kindly helped me out to bring it back to life (with the recovery package). It’s now working with the latest firmware 1.01.01.25.
Thank you very much Rob. Excellent person and excellent forum.
-
-
How’s that Siglent stock price doing after their IPO? Whatever the current stock price is it would be higher if they hired tautech.
-
#60 Reply
Posted by
WillyCrash
on 04 Apr, 2022 22:29
-
My SDM3055 stuck on boot screen, contacted tautech within a very short time I was back in operation.
Thank you tautech
-
#61 Reply
Posted by
sjb
on 13 May, 2022 21:35
-
My SDM3055 has been running 1.01.01.25 since last September. Used it more times than I can remember since then. Just powered it on and it's hung at the splash screen. So, not convinced that the latest firmware is bug free!
I've messaged @tautech in the hope he can help me.
Back to using an old Meterman 37XR handheld 3.5 digit for the weekend!
-
#62 Reply
Posted by
tautech
on 14 May, 2022 02:30
-
My SDM3055 has been running 1.01.01.25 since last September. Used it more times than I can remember since then. Just powered it on and it's hung at the splash screen. So, not convinced that the latest firmware is bug free!
I've messaged @tautech in the hope he can help me.
Back to using an old Meterman 37XR handheld 3.5 digit for the weekend!
Check your emails.
-
#63 Reply
Posted by
sjb
on 14 May, 2022 15:05
-
Thanks @tautech, I tried the recovery on numerous USB sticks (all FAT32 formatted and less than 16GB), USB LED just flickers a few times at start-up then nothing, it still just sits there on the splash screen. You can see the green LED of the micro through the back panel slots and that is on solid flashing off twice about every second, so there is activity, not sure what it should be doing under normal operation or during boot(?).
Has this recovery been tested with 1.01.01.25? I'm a bit sceptical that the last update (25) was to "Optimise the Flash R/W" ... has that had any effect on the ability to do a recovery? Has it made the hang more prone rather than fixing it?
Any Idea's what else to try? I thought from the previous posts it would be fairly straight forward, but I'm not beaten yet!
Can a recovery be done using the internal SD card slot? Is there a recovery package available for that? I guess that would wipe the calibration though!
-
#64 Reply
Posted by
sjb
on 21 May, 2022 01:45
-
Can someone do me a VERY BIG favour!
Can someone Telnet into an SDM3055 (root,ding123) and copy the filesystem to a USB stick:
cp -R /usr/bin/siglent /usr/bin/siglent/usr/mass_storage/U-disk0/
... then message me with (or post on here) a file listing of your /usr/bin/siglent/firmdata0 directory and the contents of these 2 files:
/usr/bin/siglent/firmdata0/DMM_Calibration.xml
/usr/bin/siglent/firmdata0/NSP_System_Info.xml
I've restored my Serial Number after a failed U-Disc recovery (ended up doing a SD U-Disc Recovery!), but the Info Screen now doesn't show 'Production ID : SDM3055' anymore, the line just isn't there! So, I'm wondering where that is stored in the system files.
NSP_System_Info.xml:
<?xml version="1.0" encoding="UTF-8"?>
<nsp_system_info_root>
<device>
<system_information>
<serial_number>
<chip>SERIAL NUMBER GOES HERE</chip>
</serial_number>
</system_information>
</device>
</nsp_system_info_root>
Also, My calibration file is empty:
DMM_Calibration.xml:
<?xml version="1.0" encoding="UTF-8"?>
<calibration/>
Not sure if this is the user calibration or factory calibration?? If it's the latter, it would be good to know the format of the calibration data from a good file so I can get it baseline cal'd locally, although it doesn't look too bad calibration wise as-is.
Thanks in advance!
-
#65 Reply
Posted by
tv84
on 21 May, 2022 07:51
-
Usually model names are stored in .cfg file and are reloaded by a siglent.cfg update file.
I don't remember if that is exactly the case in a SDM.
-
#66 Reply
Posted by
sjb
on 21 May, 2022 15:07
-
@tv84, do you have any info on the format or content of the siglent.cfg file?
@tautech, have you got a SDM3055 that you can extract the /usr/bin/siglent/firmdata0/ files from ... just as a reference to compare with what I have in mine.
-
#67 Reply
Posted by
tv84
on 21 May, 2022 15:57
-
@tv84, do you have any info on the format or content of the siglent.cfg file?
Not sure it exists. We would need a dump from someone's filesystem to be sure.
-
#68 Reply
Posted by
dibro
on 21 May, 2022 16:39
-
@sjb, attached the dump of my SDM3055 with Serial removed.
This is from when I received the device in August last year.
The calibration file may not be correct for you device.
-
#69 Reply
Posted by
sjb
on 21 May, 2022 18:00
-
Big thanks @dibro, I'll check it it later. One question, is your SDM3055 just factory calibrated or has it had a user cal applied? ... just so I know what the cal is possibly referring to.
-
#70 Reply
Posted by
dibro
on 21 May, 2022 18:03
-
Factory calibrated.
-
#71 Reply
Posted by
sjb
on 21 May, 2022 18:53
-
Well, I've now restored the serial number (in NSP_system_info.xml) and the device descriptor (in NSP_trends_config_info.xml). So the system information now reports all the right stuff.
Just need to get the device factory calibrated now!
-
#72 Reply
Posted by
etiboss
on 17 Sep, 2022 15:35
-
Hello All,
My SDS 1052DL+ froze on the boot screen, I tried the hard reset with the math button but it's not working. Is this possible to have the usb recovery package suggested in the previous posts?
Thanks for the Help.
-
#73 Reply
Posted by
foxbib
on 21 Nov, 2022 16:32
-
Hello everyone,
I'm new here !
As others, i have a Siglent SDM 3055 stuck on boot screen.
@tautech: Can you send me plz the magic kit
Thank you
-
#74 Reply
Posted by
foxbib
on 25 Nov, 2022 10:46
-
Hello,
Thank to tautech, i was able to recover my SDM 3055.
And now i'm upgrading all my devices
Have a good day.
++
-
-
Bonjour
I have this problem which appeared on my SDM3045x.
At first, the screen was stuck on the Siglent logo, I called the after-sales service of the distributor who sent me by email with the recovery file (for version 5.01.01.03 or more)
I used a 4G USB key but the boot didn't do anything, so suddenly I used an 8G micro SD card to attempt to boot from the main board PCB socket... and now the multimeter turns on with a black screen.
- the LED of the analog card flashes 2x per second.
-The microprocessor LED is on.
I also noticed that the label of the analog card is noted "SDM3055X-E"
the serial number of my SDM3045X is SDM34FAD2R1804 and the 3 year warranty has expired.
1: Is the recovery file corrupted or not at the correct version ?
2: Does the recovery file have to be different between a USB key and a µSD card ?
3: Is it the main board that is dead ?
I contacted SIGLENT France for an estimate of the cost of an intervention and I am waiting.
Thanking you.
Cordially
-
-
-
#77 Reply
Posted by
HamDancer
on 08 Jan, 2023 19:40
-
I've restored my Serial Number after a failed U-Disc recovery (ended up doing a SD U-Disc Recovery!), but the Info Screen now doesn't show 'Production ID : SDM3055' anymore, the line just isn't there! So, I'm wondering where that is stored in the system files.
[...]
Also, My calibration file is empty:
Hi - I had the exact same problem you did. My U-disk recovery failed and left me with an SDM3055 with no model number, a serial number of 0123456789, and no calibration. I'm still waiting for Siglent technical support to give me any useful information. How did you get your model number and serial number back? Were you ever able to get the device recalibated?
Thanks!
-HD
-
#78 Reply
Posted by
sjb
on 08 Jan, 2023 20:37
-
@HamDancer, post #12 in this thread shows you how to add/restore the Serial Number and Model Number.
The calibration for my SDM3055 was supplied to me by the Siglent tech team after a conversation with them. They supplied it as a .bin file (from their production records). They also supplied a modified version of the latest firmware that allows the calibration .bin to be loaded in the same way you would import a firmware update file. Not sure why this can't be standard feature or at least give the user the ability to backup/restore this info up to/from a USB drive! I guess they need some work for the support team to do!
My SDM3055 is now back to a good state with correct factory calibration.
-
#79 Reply
Posted by
HamDancer
on 08 Jan, 2023 21:21
-
@HamDancer, post #12 in this thread shows you how to add/restore the Serial Number and Model Number.
The calibration for my SDM3055 was supplied to me by the Siglent tech team after a conversation with them. They supplied it as a .bin file (from their production records). They also supplied a modified version of the latest firmware that allows the calibration .bin to be loaded in the same way you would import a firmware update file. Not sure why this can't be standard feature or at least give the user the ability to backup/restore this info up to/from a USB drive! I guess they need some work for the support team to do!
My SDM3055 is now back to a good state with correct factory calibration.
[Edited...]
OK it took some searching but I found at least one telnet ADS file for the SDM3055 over in the "Siglent .ads firmware file format" thread that let me login as root without a password. With a little experimenting I was able to remount the firmdata0 filesystem RW and update the model and serial numbers. Now if only I had kept a backup of the calibration file!
Thanks for the help.
-HD
-
#80 Reply
Posted by
HamDancer
on 10 Jan, 2023 19:07
-
@HamDancer, post #12 in this thread shows you how to add/restore the Serial Number and Model Number.
The calibration for my SDM3055 was supplied to me by the Siglent tech team after a conversation with them. They supplied it as a .bin file (from their production records). They also supplied a modified version of the latest firmware that allows the calibration .bin to be loaded in the same way you would import a firmware update file. Not sure why this can't be standard feature or at least give the user the ability to backup/restore this info up to/from a USB drive! I guess they need some work for the support team to do!
My SDM3055 is now back to a good state with correct factory calibration.
This is the message I got from Siglent NA tech support:
I talked with the factory and they said the USB recovery does not cause calibration data loss but the SD card recovery does.
Also they said they do not store the calibration data. We can perform calibration in our office. The cost for the SDM3055 calibration is $195.00.
Obviously they are wrong about the SD card recovery.
@sjb - Can you give me more information on how you got the calibration file? Was the file specific to your unit or a generic one? Who did you talk to?
I took the calibration file from the firmdata0 dump earlier in this thread and installed it in my unit. The measurements are pretty good now; I don't have accurate enough equipment to determine whether or not they're perfect. So I'm wondering if each unit is individually calibrated or if there is a single factory calibration for all of them.
-HD
-
#81 Reply
Posted by
tautech
on 10 Jan, 2023 20:08
-
-
-
Hello,
For my part, I use the "user" calibration using Defpom's excel solver (
https://www.thedefpom.com/siglent_sdm_calibration.php).
I made the adjustment with the help of a JF187 and a power supply connected to a multi-turn potentiometer.
It worked fine, except for the 2-wire and 4-wire ohmmeter (Errror in recalibrated mode or range).
It's a pity that the calibration format for capacitances is missing on Defpom's website.
Have a good day
-
#83 Reply
Posted by
HamDancer
on 12 Jan, 2023 19:36
-
Hello,
For my part, I use the "user" calibration using Defpom's excel solver (https://www.thedefpom.com/siglent_sdm_calibration.php).
I made the adjustment with the help of a JF187 and a power supply connected to a multi-turn potentiometer.
It worked fine, except for the 2-wire and 4-wire ohmmeter (Errror in recalibrated mode or range).
It's a pity that the calibration format for capacitances is missing on Defpom's website.
Have a good day
What is a JF187? I can't find that anywhere on the internet related to instrumentation.
Isn't the whole issue that you need a meter that is already more precise than the SDM3055 in order to do the calibration? Not to mention a power supply that can go to 1000V. That's not equipment that most people have available.
And why didn't the resistance calibrate properly? That seems like a major problem...
Thanks!
-HD
-
-
Hello,
JF187 is a John Fluke 187, it has the same resolution as the SDM3045x, a basic accuracy of 0.025% (0.02 for the SDM3045x) and with up-to-date metrological control.
For poor man's calibration
I made an electronic voltage booster assembly for the 200v (the 1000v, I will do it later).
For the resistance and capacity calibration file, I will ask SIGLENT France (perhaps Defpom can pass it on to me) if they do not have a manual because it is a user function and it is not normal not to not have user information for that.
In March, I have a metrology team that comes to my workshop for two months for checks and calibrations, I would ask them if they know the calibration protocol in GPIB with a Fluke 5522A for the SIGLENT brand, for try to recover a factory calibration (Calibration.xlsx file).
-
#85 Reply
Posted by
Ruediger
on 24 Jan, 2023 10:14
-
Hello,
I have the same problem as you with the SDM3045x at work.
According to a colleague, who unfortunately no longer works for us, the method with the USB stick was not successful.
I have been waiting for an answer from Siglent support for 1.5 weeks now.
Could someone send me the data and a short guide to the SD card method?
thanks advance
Ruediger
-
#86 Reply
Posted by
tautech
on 24 Jan, 2023 11:21
-
Hello,
I have the same problem as you with the SDM3045x at work.
According to a colleague, who unfortunately no longer works for us, the method with the USB stick was not successful.
I have been waiting for an answer from Siglent support for 1.5 weeks now.
Could someone send me the data and a short guide to the SD card method?
thanks advance
Ruediger
USB recovery is very dependant of correct format of the USB stick.
It must be 8GB max and formatted in FAT32 with 4k clusters.
Nothing else can be guaranteed to work.
Please try again a time or 2 and then contact me via PM with your email if necessary.
Be sure to get it updated to the latest firmware from here:
https://int.siglent.com/download/firmwares/?ProId=25
-
#87 Reply
Posted by
TheDefpom
on 02 Mar, 2023 07:47
-
Hello,
For my part, I use the "user" calibration using Defpom's excel solver (https://www.thedefpom.com/siglent_sdm_calibration.php).
I made the adjustment with the help of a JF187 and a power supply connected to a multi-turn potentiometer.
It worked fine, except for the 2-wire and 4-wire ohmmeter (Errror in recalibrated mode or range).
It's a pity that the calibration format for capacitances is missing on Defpom's website.
Have a good day
I just stumbled upon this comment, I am now working to add capacitance mode calibration, it will be untested and as such it may not work.
I have very limited information from Siglent about the calibration files, and even suspect some to not be correct, I only have the SDM3065X so that is all I can test with, capacitance mode isn't even included in the user cal files for the SDM3065X, but I do have it for the 3045X/3055 so it will be based upon that, it MIGHT work.
It is interesting that the resistance calibrations didn't work for you, I will have to look at that more closely to see if I can figure out why.
-
#88 Reply
Posted by
TheDefpom
on 02 Mar, 2023 21:17
-
-
-
Hello to all,
I have the same problem with my Siglent sdm3055. It is blocked on the boot screen with the Siglent logo. I'm working at Rome University (Italy).
I'll thank you in advance if you can help me somehow.
-
#90 Reply
Posted by
tobi1024
on 02 May, 2023 20:31
-
Hello,
my SDM3055 is now also stuck on the boot screen all of the sudden. It's a bit more than 3 years old now and of course out of warranty. Installed firmware should be V1.01.01.25.
Any help is greatly appreciated!
Thank you
-
#91 Reply
Posted by
catsgsm
on 19 Jun, 2023 21:03
-
Hello,
Today, my SDM3055 also stays at the start logo...
A few months ago it was updated to 1.01.01.25.
Can someone help me with the recovery file?
-
#92 Reply
Posted by
tautech
on 19 Jun, 2023 21:18
-
Hello,
Today, my SDM3055 also stays at the start logo...
A few months ago it was updated to 1.01.01.25.
Can someone help me with the recovery file?
Welcome to the forum.
Send me a PM with your email address and we'll have you up and running again in no time then you can install the later 1.27R2 firmware.
-
#93 Reply
Posted by
Machtl
on 30 Aug, 2023 17:08
-
Hello tautech!
All of a sudden i ran into the same problem today with my SDM3055. It booted normally a few days ago,
but today it hangs at the bootscreen. I can't remember that i ever did a fw upgrade, it worked well all the time.
Can you maybe help me with this problem?
Best regards from Austria, Martin
-
#94 Reply
Posted by
ozzy
on 02 Sep, 2023 23:03
-
Hello,
I have the same problem with my sdm3055 I hope you can help me too thanks
-
#95 Reply
Posted by
paulc_6969
on 25 Sep, 2023 17:03
-
Looking at the SDM3055 as a bench MM, but after reading so much about the "SDM3055 stuck", I see the solution:
BUY A RIGOL!
-
#96 Reply
Posted by
tautech
on 25 Sep, 2023 18:46
-
Looking at the SDM3055 as a bench MM, but after reading so much about the "SDM3055 stuck", I see the solution:
BUY A RIGOL!
Right, so to put the record straight, all problems are due to old FW and poor NAND management.
With recent FW this is no longer an issue.
Even Keysight had their problems with NAND management in DSO's worth much more than any of these SDM models.
-
#97 Reply
Posted by
KungFuJosh
on 25 Sep, 2023 20:13
-
Looking at the SDM3055 as a bench MM, but after reading so much about the "SDM3055 stuck", I see the solution:
BUY A RIGOL!
You might be shocked to know that early firmware versions of Rigol DMMs also had boot issues. It's Mind blowing to think that outdated firmware might have bugs on any brand. 🙄
I've never had any issues with my Siglent DMMs.
-
#98 Reply
Posted by
vitalff
on 27 Sep, 2023 09:39
-
Hi.
I also have a problem with boot SDM3055 (screen - siglent logo). Version after last update SDM3055_1.01.01.20R2.
Please help me. Thx.
-
#99 Reply
Posted by
tautech
on 27 Sep, 2023 09:49
-
Hi.
I also have a problem with boot SDM3055 (screen - siglent logo). Version after last update SDM3055_1.01.01.20R2.
Please help me. Thx.
Welcome to the forum.
PM received and returned.
-
#100 Reply
Posted by
ODJeetje
on 30 Sep, 2023 18:31
-
Hello forum,
I ran into trouble with my SDM3055 at work. It's an early model with the old type rubber corners.
Up till recently it took all the firmware updates without any problems. (Latest uploaded version was V1.01.01.25)
When I tried to update to the newer version V1.01.01.27R2 it did not detect the USB stick, although it's the stick I have been using before.
After some failing attempts I tried to update bij putting the .ADS file onto a uSD card and used the slot on the main board.
I wish I hadn't done that...
Now the unit won't even boot anymore.
I'm left with a black screen, a blinking LED on the mainboard and a blip from the piezo every few seconds...
Is there a way to bring this one to life again?
-
#101 Reply
Posted by
tautech
on 30 Sep, 2023 18:35
-
Hello forum,
I ran into trouble with my SDM3055 at work. It's an early model with the old type rubber corners.
Up till recently it took all the firmware updates without any problems. (Latest uploaded version was V1.01.01.25)
When I tried to update to the newer version V1.01.01.27R2 it did not detect the USB stick, although it's the stick I have been using before.
After some failing attempts I tried to update bij putting the .ADS file onto a uSD card and used the slot on the main board.
I wish I hadn't done that...
Now the unit won't even boot anymore.
I'm left with a black screen, a blinking LED on the mainboard and a blip from the piezo every few seconds...
Is there a way to bring this one to life again?
Welcome to the forum.
I too wish you hadn't done that.
But we can fix it.
Send me PM with your email.
-
#102 Reply
Posted by
rooduction
on 01 Oct, 2023 07:56
-
Dear Forum
I found all your entries searching for the same topic on the web, and I think I am in the right place.
Beeing a hobby electronics guy, I work on some projects from time to time only.
During the last years I rarely touched my equipment and as I wanted to fire up my osciloscope yesterday it was stuck on the boot screen.
I already tried the Math-Button-Trick, and it changed the lighting of the buttons, in combination with some internal relay switching.
It is a "SIGLENT SDS 1102CML+" and when stuck it shows the logo on screen and the LEDs of the Channel 1-, Utility- and Run/Stop-Button glow. (Probably some kind of error code?)
What other information do I need to provide for proper analysation?
Can somebody probably help me with this?
Greetings from Austria
Yours, Robert
-
#103 Reply
Posted by
tautech
on 01 Oct, 2023 07:59
-
Dear Forum
I found all your entries searching for the same topic on the web, and I think I am in the right place.
Beeing a hobby electronics guy, I work on some projects from time to time only.
During the last years I rarely touched my equipment and as I wanted to fire up my osciloscope yesterday it was stuck on the boot screen.
I already tried the Math-Button-Trick, and it changed the lighting of the buttons, in combination with some internal relay switching.
It is a "SIGLENT SDS 1102CML+" and when stuck it shows the logo on screen and the LEDs of the Channel 1-, Utility- and Run/Stop-Button glow. (Probably some kind of error code?)
What other information do I need to provide for proper analysation?
Can somebody probably help me with this?
Greetings from Austria
Yours, Robert
Welcome to the forum.
Not really the correct thread but nevermind.
Send me a PM with your email address.
-
#104 Reply
Posted by
dfharman
on 06 Oct, 2023 19:00
-
Hi,
I have an SDM3055 and turned it on this morning to get the dreaded stuck on boot screen issue! I've my SDM3055 for a couple of years, use it every few weeks, but haven't updated the firmware on it for a good few months - I think I last updated it to 20R2.
Help! Would it be possible to get the recovery file please??
-
#105 Reply
Posted by
tautech
on 06 Oct, 2023 19:05
-
Hi,
I have an SDM3055 and turned it on this morning to get the dreaded stuck on boot screen issue! I've my SDM3055 for a couple of years, use it every few weeks, but haven't updated the firmware on it for a good few months - I think I last updated it to 20R2.
Help! Would it be possible to get the recovery file please??
Welcome to the forum.
More than a couple of years as 20R2 is from late 2019 !
Send PM with your email.
-
#106 Reply
Posted by
dfharman
on 07 Oct, 2023 07:17
-
Thanks Tautech, pm sent.
-
#107 Reply
Posted by
tautech
on 07 Oct, 2023 07:27
-
Thanks Tautech, pm sent.
Email returned.
-
#108 Reply
Posted by
maralb
on 07 Oct, 2023 11:39
-
Forgive me if this is already asked before: is this problem related to certain hardware revisions or firmware versions or can this happen to all SDM3055's?
-
#109 Reply
Posted by
tautech
on 07 Oct, 2023 18:06
-
Forgive me if this is already asked before: is this problem related to certain hardware revisions or firmware versions or can this happen to all SDM3055's?
It's a NAND management issue like HPAK had with their DSO's.
Latest firmware versions solve these issues.
-
#110 Reply
Posted by
dfharman
on 09 Oct, 2023 19:19
-
Thanks Tautech, all sorted!
-
#111 Reply
Posted by
ds_user
on 12 Oct, 2023 14:16
-
Thanks for sharing info about the SDM3055 firmware issue. I did get "5.5 digital DMM USB recovery and instruction.zip" but it does not work on a 2GB, 8GB flash drive, freshly formatted to FAT32 and copied the 8 files (no matter with or without the instruction PDF). I also tried a 8GB micro sd card. Not working either.
Then I tried to open the case and push the microsd into the internal card slot. Still doesn't work. Did I do anything wrong?
Also I tried to prevent the bug from happening on another SDM3055, presumably the same batch. It shows:
SW Version ID: 1.01.01.19
HW Version ID: 03-00-00-13-00
But the meter does not show any thing when I plug flash drive in the front, nor will it be recognized as a hardware on computer from the USB on the back with a cable, how can I update firmware on that?
-
#112 Reply
Posted by
KungFuJosh
on 12 Oct, 2023 14:22
-
Thanks for sharing info about the SDM3055 firmware issue. I did get "5.5 digital DMM USB recovery and instruction.zip" but it does not work on a 2GB, 8GB flash drive, freshly formatted to FAT32 and copied the 8 files (no matter with or without the instruction PDF). I also tried a 8GB micro sd card. Not working either.
Then I tried to open the case and push the microsd into the internal card slot. Still doesn't work. Did I do anything wrong?
Also I tried to prevent the bug from happening on another SDM3055, presumably the same batch. It shows:
SW Version ID: 1.01.01.19
HW Version ID: 03-00-00-13-00
But the meter does not show any thing when I plug flash drive in the front, nor will it be recognized as a hardware on computer from the USB on the back with a cable, how can I update firmware on that?
Try formatting your flash drive with this first:
https://www.sdcard.org/downloads/formatter/For your other 3055 that still works, just update the normal way to the latest firmware.
-
#113 Reply
Posted by
tv84
on 12 Oct, 2023 14:38
-
Then I tried to open the case and push the microsd into the internal card slot. Still doesn't work. Did I do anything wrong?
AFAIR the software for the USB is not the same to put in the internal slot. @tautech is better to explain.
-
#114 Reply
Posted by
tautech
on 12 Oct, 2023 18:31
-
I did get "5.5 digital DMM USB recovery and instruction.zip" but it does not work on a 2GB, 8GB flash drive, freshly formatted to FAT32 and copied the 8 files (no matter with or without the instruction PDF). I also tried a 8GB micro sd card. Not working either.
From whom ?
Then I tried to open the case and push the microsd into the internal card slot. Still doesn't work. Did I do anything wrong?
Yep, big time.
Contact NA HQ to return it and be properly recovered.
Also I tried to prevent the bug from happening on another SDM3055, presumably the same batch. It shows:
SW Version ID: 1.01.01.19
HW Version ID: 03-00-00-13-00
But the meter does not show any thing when I plug flash drive in the front, nor will it be recognized as a hardware on computer from the USB on the back with a cable, how can I update firmware on that?
Only the front USB is used with USB sticks and the GUI should flag when a USB stick is inserted with the USB symbol. Access to the FW update menu is within the Utility menu.
Download this latest firmware, unzip and study the update instructions for any special steps required to update from this now very old 1.01.01.19 firmware.
https://int.siglent.com/upload_file/zip/firmware/A_multimeter/SDM3055_V1.01.01.27R2_EN.zipDownload any other versions required from here:
https://int.siglent.com/download/firmwares/?ProId=26
-
#115 Reply
Posted by
rooduction
on 12 Oct, 2023 20:38
-
Dear Forum
I found all your entries searching for the same topic on the web, and I think I am in the right place.
Beeing a hobby electronics guy, I work on some projects from time to time only.
During the last years I rarely touched my equipment and as I wanted to fire up my osciloscope yesterday it was stuck on the boot screen.
I already tried the Math-Button-Trick, and it changed the lighting of the buttons, in combination with some internal relay switching.
It is a "SIGLENT SDS 1102CML+" and when stuck it shows the logo on screen and the LEDs of the Channel 1-, Utility- and Run/Stop-Button glow. (Probably some kind of error code?)
What other information do I need to provide for proper analysation?
Can somebody probably help me with this?
Greetings from Austria
Yours, Robert
Welcome to the forum.
Not really the correct thread but nevermind.
Send me a PM with your email address.
Thank you, tautech - Recovery worked like a charm - Thank you very much for all your help!
Please keep up the great work.
Greetings from Austria
Yours, Robert
-
#116 Reply
Posted by
ds_user
on 12 Oct, 2023 21:09
-
I did get "5.5 digital DMM USB recovery and instruction.zip" but it does not work on a 2GB, 8GB flash drive, freshly formatted to FAT32 and copied the 8 files (no matter with or without the instruction PDF). I also tried a 8GB micro sd card. Not working either.
From whom ?
Then I tried to open the case and push the microsd into the internal card slot. Still doesn't work. Did I do anything wrong?
Yep, big time.
Contact NA HQ to return it and be properly recovered.
Also I tried to prevent the bug from happening on another SDM3055, presumably the same batch. It shows:
SW Version ID: 1.01.01.19
HW Version ID: 03-00-00-13-00
But the meter does not show any thing when I plug flash drive in the front, nor will it be recognized as a hardware on computer from the USB on the back with a cable, how can I update firmware on that?
Only the front USB is used with USB sticks and the GUI should flag when a USB stick is inserted with the USB symbol. Access to the FW update menu is within the Utility menu.
Download this latest firmware, unzip and study the update instructions for any special steps required to update from this now very old 1.01.01.19 firmware.
https://int.siglent.com/upload_file/zip/firmware/A_multimeter/SDM3055_V1.01.01.27R2_EN.zip
Download any other versions required from here:
https://int.siglent.com/download/firmwares/?ProId=26
Thanks for your detailed reply, I got recovery from siglentna.com. And it doesn't work in my case.
Also unfortunately the USB drive does not show anything in the working unit. I tested there are 5V on USB port and steady 0.0022V on D+ and D-, through a new USB extension cable. So it shouldn't be a connection issue. Maybe that unit also need to be shipped back for repair?
-
#117 Reply
Posted by
KungFuJosh
on 12 Oct, 2023 21:13
-
Also unfortunately the USB drive does not show anything in the working unit. I tested there are 5V on USB port and steady 0.0022V on D+ and D-, through a new USB extension cable. So it shouldn't be a connection issue. Maybe that unit also need to be shipped back for repair?
Did you try formatting a USB drive with the formatter I linked above?
It sounds like your flash drives are either incompatible, or more likely, poorly formatted. Or you might need to wait longer for the DMM to register the USB drive.
-
#118 Reply
Posted by
tautech
on 12 Oct, 2023 22:16
-
I did get "5.5 digital DMM USB recovery and instruction.zip" but it does not work on a 2GB, 8GB flash drive, freshly formatted to FAT32 and copied the 8 files (no matter with or without the instruction PDF). I also tried a 8GB micro sd card. Not working either.
From whom ?
Then I tried to open the case and push the microsd into the internal card slot. Still doesn't work. Did I do anything wrong?
Yep, big time.
Contact NA HQ to return it and be properly recovered.
Also I tried to prevent the bug from happening on another SDM3055, presumably the same batch. It shows:
SW Version ID: 1.01.01.19
HW Version ID: 03-00-00-13-00
But the meter does not show any thing when I plug flash drive in the front, nor will it be recognized as a hardware on computer from the USB on the back with a cable, how can I update firmware on that?
Only the front USB is used with USB sticks and the GUI should flag when a USB stick is inserted with the USB symbol. Access to the FW update menu is within the Utility menu.
Download this latest firmware, unzip and study the update instructions for any special steps required to update from this now very old 1.01.01.19 firmware.
https://int.siglent.com/upload_file/zip/firmware/A_multimeter/SDM3055_V1.01.01.27R2_EN.zip
Download any other versions required from here:
https://int.siglent.com/download/firmwares/?ProId=26
Thanks for your detailed reply, I got recovery from siglentna.com. And it doesn't work in my case.
If the FW was 1.01.01.19 it will work but it seems instructions from NA were not detailed enough.
Also unfortunately the USB drive does not show anything in the working unit. I tested there are 5V on USB port and steady 0.0022V on D+ and D-, through a new USB extension cable. So it shouldn't be a connection issue.
With latest firmware ?
Maybe that unit also need to be shipped back for repair?
If you're struggling with FW updates you should and maybe get the Cal checked too.
-
#119 Reply
Posted by
ds_user
on 13 Oct, 2023 10:47
-
Also unfortunately the USB drive does not show anything in the working unit. I tested there are 5V on USB port and steady 0.0022V on D+ and D-, through a new USB extension cable. So it shouldn't be a connection issue. Maybe that unit also need to be shipped back for repair?
Did you try formatting a USB drive with the formatter I linked above?
It sounds like your flash drives are either incompatible, or more likely, poorly formatted. Or you might need to wait longer for the DMM to register the USB drive.
Thanks for sharing the tool. Unfortualy it did not work in my case. I tried the tool, format in win11, format in Ubuntu, nothing works.
-
#120 Reply
Posted by
tv84
on 13 Oct, 2023 14:31
-
1. Change USB disk.
2. Format in linux.
3. Try flashing FW.
4. If still doesn't work, go to Step 1.
-
#121 Reply
Posted by
KungFuJosh
on 13 Oct, 2023 15:19
-
What tv84 said. If flashing with that utility or in Linux doesn't work, the drive is very likely the problem.
-
#122 Reply
Posted by
CobblePro
on 29 Oct, 2023 15:17
-
I have the same problem. Started 2 days ago. I wonder if there was a bad batch of meters that all started failing in the last few months?
-
#123 Reply
Posted by
Peter_O
on 31 Oct, 2023 12:19
-
+1
I bought my 3055 end of 2020.
Since then I did some FW updates, last was most probably SDM3055_V1.01.01.27R2_EN from April 2023.
After that , the 3055 has been in use several times without any problem.
Today it stucks on boot screen after power on. Power off/on does not help.
Do I need to contact Siglent, or is there a more efficient way to solve the issue?
-
#124 Reply
Posted by
Martin72
on 31 Oct, 2023 12:44
-
Wait until Tautech is back online and read your post.
-
#125 Reply
Posted by
Peter_O
on 01 Nov, 2023 13:18
-
Problem is fixed with the kind support of tautech.
Thx a lot!
-
#126 Reply
Posted by
JirkaJ
on 03 Dec, 2023 11:20
-
Hi,
I have an SDM3055 and turned it on this morning to get the dreaded stuck on boot screen issue! I've my SDM3055 for a couple of years, use it every few weeks, but haven't updated the firmware on it for a good few months.
Help! Would it be possible to get the recovery file please??
-
#127 Reply
Posted by
tautech
on 03 Dec, 2023 11:54
-
Hi,
I have an SDM3055 and turned it on this morning to get the dreaded stuck on boot screen issue! I've my SDM3055 for a couple of years, use it every few weeks, but haven't updated the firmware on it for a good few months.
Help! Would it be possible to get the recovery file please??
Welcome to the forum.
Sure, send PM with your email address.
-
#128 Reply
Posted by
JirkaJ
on 05 Dec, 2023 05:14
-
Thank you for the help from tautech, he tried to help me, but I have a USB input without power. I have to send it to Siglent service.
-
#129 Reply
Posted by
tautech
on 05 Dec, 2023 06:38
-
Thank you for the help from tautech, he tried to help me, but I have a USB input without power. I have to send it to Siglent service.
Not so fast, check your emails.
-
#130 Reply
Posted by
JirkaJ
on 05 Dec, 2023 10:14
-
Thank you for your cooperation tautech, he is very willing and communicative in solving my SDM 3055 problem.
We solved everything with the help of an SSD, again the SDM 3055 works as it should.
Once again, I thank this forum, where I got help.
Sorry for the translation.
JirkaJ
-
-
eventually, I could fix my SDM3055 with the valuable hints in this thread.
After re-formatting a 8GB-USB Stick on raspy linux and copying the 1.01.01.19 FW from a rescue-pack of the German Seller Batronix , the recovery on the front USB succeeded.
Now I wanted to further upgrade to the latest FW version, but without success.
After deleting all files on the USB stick, I copied the latest SDM3055_V1.01.01.27R2.ADS to the root of the stick and followed the instructions.
But the progress bar of the upgrade procedure finished in a second and shortly displayed "Update Fail"
An attempt to upgrade in a sequential way with the .ADS file versions following 1.01.01.19 chronological failed in the very same way.
How can I fix this ?
Thanks for any suggestion,
Joe
-
#132 Reply
Posted by
tautech
on 07 Dec, 2023 21:23
-
eventually, I could fix my SDM3055 with the valuable hints in this thread.
After re-formatting a 8GB-USB Stick on raspy linux and copying the 1.01.01.19 FW from a rescue-pack of the German Seller Batronix , the recovery on the front USB succeeded.
Now I wanted to further upgrade to the latest FW version, but without success.
After deleting all files on the USB stick, I copied the latest SDM3055_V1.01.01.27R2.ADS to the root of the stick and followed the instructions.
But the progress bar of the upgrade procedure finished in a second and shortly displayed "Update Fail"
(Attachment Link)
An attempt to upgrade in a sequential way with the .ADS file versions following 1.01.01.19 chronological failed in the very same way.
How can I fix this ?
Thanks for any suggestion,
Joe
Welcome to the forum.
1.01.01.19 FW from a rescue-pack
Is the very first recovery pack and now severely outdated.
Latest version we have recovers to V1.01.01.25 which still requires installing V1.01.01.27R2 to get to latest.
Contact us via PM with your email address then send one to Batronix asking them to update their resources.
-
#133 Reply
Posted by
tibimakai
on 20 Dec, 2023 03:59
-
I'm the latest victim of seeing only the splash screen, of my SDM3055.
I would appreciate it a lot, if you can help me out.
Thank you,
Tibor
-
#134 Reply
Posted by
tibimakai
on 22 Dec, 2023 03:14
-
Contacted tautech, and he is on holiday brake until January.
I was advised to ask help from another forum member, who received this software.
I would appreciate it a lot, if somebody would be kind enough to share this software with me.
Thank you in advance.
Tibor
-
#135 Reply
Posted by
rf-loop
on 22 Dec, 2023 05:04
-
You have PM
-
#136 Reply
Posted by
tibimakai
on 22 Dec, 2023 20:34
-
Thank you very much for the prompt help.
I have managed to revive my meter!!
Also upgraded to the latest firmware 27R2.
Now I have to find out, how to get back my Rigol DP711, because has a fan error and the output is disabled.
-
-
Hi all,
I know this thread is a bit older, but my SDM 3055 also is stuck @ boot screen.
I already got the USB recovery package from Siglent.eu but it didn't work (also created multiple USB sticks on Linux).
Maybe you can supply me with the mentioned "Magic" Rescue package ?
Thank you very much
-
-
I too am afflicted with this terrible nuisance. SDM3065X stuck on boot screen
@tautech ... I will send you a PM
-
-
I have an SDS 1102CML... totally corrupted boot, freezes on splash screen. Factory reset just has it freeze 1-2 seconds after boot...
wondering what I can do to suck up to @tauetech or someone else to get it fixed?
-
#140 Reply
Posted by
jeroen
on 12 Feb, 2024 22:07
-
I had the same issue with my SDM3045X, however I somehow got out of the bootloop by doing the following:
1. Switch off the power supply (or unplug the power cable)
2. Press the on/off button and keep it pressed while switching on the power supply
After this it would boot up and I could install the latest firmware.
-
#141 Reply
Posted by
tautech
on 12 Feb, 2024 22:37
-
I had the same issue with my SDM3045X, however I somehow got out of the bootloop by doing the following:
1. Switch off the power supply (or unplug the power cable)
2. Press the on/off button and keep it pressed while switching on the power supply
After this it would boot up and I could install the latest firmware.
Boot loop or boot freeze ?
Wow, #2 is very interesting and please share more if you can.
-
#142 Reply
Posted by
KungFuJosh
on 12 Feb, 2024 22:42
-
I had the same issue with my SDM3045X, however I somehow got out of the bootloop by doing the following:
1. Switch off the power supply (or unplug the power cable)
2. Press the on/off button and keep it pressed while switching on the power supply
After this it would boot up and I could install the latest firmware.
Boot loop or boot freeze ?
Wow, #2 is very interesting and please share more if you can.
That's interesting. #2 is a common reset feature on some networking equipment and sometimes other devices.
-
#143 Reply
Posted by
tautech
on 12 Feb, 2024 22:48
-
I had the same issue with my SDM3045X, however I somehow got out of the bootloop by doing the following:
1. Switch off the power supply (or unplug the power cable)
2. Press the on/off button and keep it pressed while switching on the power supply
After this it would boot up and I could install the latest firmware.
Boot loop or boot freeze ?
Wow, #2 is very interesting and please share more if you can.
That's interesting. #2 is a common reset feature on some networking equipment and sometimes other devices.
Still, the real question is why a boot loop came about.....maybe a conversion attempt to 3055X-E ?
-
#144 Reply
Posted by
Patoche92
on 15 Feb, 2024 11:22
-
Hello
I own a Siglent SDM 3055
I have the same problem it is also stuck on the boot screen, it is from 2019 with the latest firmware V1.01.01.27R2 I have tried all the manipulations but nothing helps.
please excuse me for my very approximate English.
This is a google translation
Thanks in advance
-
#145 Reply
Posted by
Patoche92
on 15 Feb, 2024 13:56
-
A big thank you to Tautech who repaired my SDM3055 in less than 1 hour
Excuse my English
-
#146 Reply
Posted by
DonB
on 16 Feb, 2024 02:17
-
Hi,
I have joined the "Stuck on Splash Screen" club for the second time. The first time Siglent said I had to send my SDM-3055 back to them for repair.
tautech, would you be able to help me? I don't know what version of firmware is installed but I vaguely remember checking it when I got it back the last time(I think around 11/2019) that it was the latest then.
Thanks!
-
#147 Reply
Posted by
jeroen
on 03 Mar, 2024 11:45
-
I had the same issue with my SDM3045X, however I somehow got out of the bootloop by doing the following:
1. Switch off the power supply (or unplug the power cable)
2. Press the on/off button and keep it pressed while switching on the power supply
After this it would boot up and I could install the latest firmware.
Boot loop or boot freeze ?
Wow, #2 is very interesting and please share more if you can.
That's interesting. #2 is a common reset feature on some networking equipment and sometimes other devices.
Still, the real question is why a boot loop came about.....maybe a conversion attempt to 3055X-E ?
It just happened after I turned it on after not using it for some time.
I hadn't even thought about updating or changing any software on it (if it ain't broken).
I'm not entirely sure what kind of issue it was and if it is the same problem others have had here.
The meter would boot up after a while, but once the meter screen came up nothing worked anymore.
No controls and no measurements were shown.
-
#148 Reply
Posted by
freeky
on 26 Mar, 2024 09:37
-
Hallo, bin neu im Forum und habe auch das Problem mit dem eingefrorenem Startbildschirm
-
#149 Reply
Posted by
Peter_O
on 26 Mar, 2024 10:11
-
Hallo, bin neu im Forum und habe auch das Problem mit dem eingefrorenem Startbildschirm
Hello freeky
welcome here.
This forum is a global one, so english will help.
Here's freekies post:
"Hello, I'm new here and I face the same problem with the stuck bootscreen."
I can imagine tautech posting you a private note.
(Möglicherweise bekommst Du von tautech eine private Nachricht.)
-
#150 Reply
Posted by
tautech
on 26 Mar, 2024 10:16
-
Hallo, bin neu im Forum und habe auch das Problem mit dem eingefrorenem Startbildschirm
Hello freeky
welcome here.
This forum is a global one, so english will help.
Here's freekies post:
"Hello, I'm new here and I face the same problem with the stuck bootscreen."
I can imagine tautech posting you a private note.
(Möglicherweise bekommst Du von tautech eine private Nachricht.)
freeky has already sent a PM that he/her needs to translates and send again along with his/her email address.
-
#151 Reply
Posted by
lordheavy
on 15 Apr, 2024 12:59
-
Hi, i have a similar issue with a Siglent 1202X-E, it is stuck on boot with the logo, i plugged an usb key to see if it is detected, and it is. I tried to do the reset with the 'math' button with no luck.
Any more help appreciated. Thanks.
-
#152 Reply
Posted by
KungFuJosh
on 15 Apr, 2024 14:04
-
Hi, i have a similar issue with a Siglent 1202X-E, it is stuck on boot with the logo, i plugged an usb key to see if it is detected, and it is. I tried to do the reset with the 'math' button with no luck.
Any more help appreciated. Thanks.
Did you disconnect everything besides the power cable? Including any probes or network or usb cables?
-
-
I'm also stuck at the boot screen with my SDM3055 :/
...save me Tautech...you're my only hope.