This scope is really good at measuring differential voltages inside a circuit.
Sure, you can do that - with one channel - and it's still somewhat problematic. A scope running from a battery is not the same as an isolated scope.
If the ground of the scope is actually on a high level, the bare metal metal BNC connectors are a safety hazard.
Yes, that's high sensitivity + reliable and fast trigger + real time sampling + high acquisition rate.
Well, a scope without a "reliable" trigger is not a scope. Not sure what a "fast" trigger is. A trigger is a trigger. And again, the trigger system is about as basic as can be.
Side note: there is no trigger jitter specified. Actually, I couldn't find any specification for the accuracy of the time base, either.
"Real time sampling" is something that every normal scope does. Most modern scope also perform some kind of "equivalent time base sampling" for cyclic signals (and if the real time sample rate is not sufficient), but that doesn't change the fact that they are real-time sampling scopes.
And about the "high acquisition rate": again there is no (maximum) number specified for waveforms/s. Most modern scopes sit somewhere between 50k and 1 Million, so if I had to guess, the DsTouch scopes would be more on the low end of this.
But sure, the DS4T252 is probably better than most offerings in the ~100€ range. But this is mainly because most of them are total garbage or outdated.
Still, the form factor is the most interesting feature but also kinda limits the actual use cases. Fiddling around on such a small screen is problematic, protocol decoding is nearly useless due to the screen size, the limitation to two channels and the lacking trigger options.
I fear that even though it's kinda cool, I would mainly buy it to buy it but would rarely actually use it.