Congratulations on your 732A. I hope it is stable which most appear to be after a good warmup and settling time.
The recommended way to monitor the drift of a reference is to monitor the series-opposition voltage against a known good reference (or more).
A direct reading with the 3458A is not the best way due to its own drift potentially skewing your measured data.
If you can find someone local to you with a reference or two, I would recommend monitoring against their 10V output. The other reference should have a calibration history if possible.
NBS430 is the document that describes monitoring 3 or more references.
https://dataproof.com/Publications/NBS430.PDF732A/B are rabbit holes because of the need for at least 3 if you want to continually monitor the drift. I have been down that hole for several years now and it got expensive real quick. A 34420A and Data Proof scanner added to the cost.
I also replace the electrolytics on the pre-regulator and regulator PCB. The last few 732A have also had all the carbon comp resistors replaced. At least half were out of tolerance after 40+ years of service. The schaffner input power filters have all been replaced.
You can use shielded cables with twisted copper wires. Networking cable is fine. Just don't use anything with nickel plating as you will be measuring into the nV range.