Thanks - As a NOOB on this calibration stuff, is it possible to increase the apparent accuracy of the calibrator through measuring with additional equipment?
For example, the 5100B is stated to be for calibrating 4.5 digit equipment - If I had a DMM that was calibrated and could say read to 85 digits, would I be able to use the two of them to calibrate something rated to 7.5 digits?
While it clearly is a lot more complicated than this, isn't the basic idea:
1. Create voltage
2. Measure that voltage to a given accuracy
3. Measure that voltage again with the DUT
4. Adjust DUT until it's reading is within it's accuracy spec
TonyG
This may work, but it depends on the specification of the calibrator.
A first trick is, to calibrate these calibrators in shorter time periods than the usual 1 year, like every 3 months, or each month, to calibrate the DMM regarding its 1 year specification.
So the calibrator has a better uncertainty, but only for the case that it's already specified for 1 year, 3 months, 1 month, 24h and 10 min (Transfer stability), and if the calibration of the calibrator itself is sufficiently uncertain, of course.
In the extreme case, what you describe, is using the Transfer stability (10 min.) to transfer the 8.5 digit DMMs uncertainty to the 7.5 digit DUT - DMM.
So, if you look into the 5100B specification, you will find that this calibrator is relatively unstable, and the specification is incomplete or really bad, as no 24h, 3 mo., specs are given.. They did not even specify, over which time the spec. is valid, I assume 1 year, maybe.
But they give a 10 min. Transfer Stability, for DCV it is about 10ppm plus 2 ppm of range, 25ppm for 500V.
So to answer your question:
Nope, this is NOT sufficient for 7.5 digit DMM, even not for 6.5 digits.
These have 20ppm or better 24h specification, and the T.U.R. (test uncertainty ratio) is too low, about 20ppm/12ppm ~ 1.7, so you would need something more stable or more uncertain, to have a TUR of 3..4 or better.
What you could try instead, is to hook the 8.5 digit DMM and the DUT in parallel and measure the calibrators output in parallel, over the same time window.
The crucial thing is, you need also to determine the standard deviation of the calibrators output, using the 8.5 digits DMM statistics function, and also the DUTs'.
3 times the STD will give the 95 % transfer uncertainty of this setup, and MIGHT be better than the spec. of the calibrator.
If the calibrator is too noisy, or too unstable, you will fail, anyhow, as the STD might be too high.
Also, the T.U.R. between the 8.5 digit DMM and the DUT might again be too low.
But this may also be handled, by using a "Golden Standard", that is an instrument, with tested tighter uncertainty limits, than specified.
Fluke and Keysight do this, they use "Golden" 5720A/30A calibrators and also "Golden" 3458As to calibrate 3458A DUTs.
Sounds confusing, I fear.. I can recommend the Fluke and Keysight Web Seminars about these techniques, and also Flukes nice ANs about this subject.
Frank
PS: I also own a 5442A and a 5200A, and a 3458A.
I use the 3458A together with Swerleins Algorithm, to create precise AC voltages for calibration of 6.5 digit DMMs, the 5200A used as an AC source only.