I understand that I'm moving on very thin ice and I don't have enough data to claim something, nor do I own the necessary equipment for doing this. At the moment I'm a bit exploring in which direction I should/want to go and was wondering about the TC stability of the R6851T meters I have and possible improvement. A bit to explore the borders of my R6581T meters.
I've attached below plots of my two meters sampling the same 10V source with 100NPLC over night. In the x axis we have samples. The test setup on my bench is shown in another picture. The sampling was started after about ~5h warm up and an internal calibration of the meters. The temperature in the room is stable to within +/-1°C, I did not have time this morning to download the data of the temp/humidity logger. Maybe the captured data is useful for others. I started this morning another acquisition, let's see how stable it's over the day.
I do not want to go offtopic, but for reference I've attached two plots of my voltage source as measured with the 3458As. The setup was left to warm up for about ~5h and an ACAL was performed. The temperature and humidity in the room is controlled and maintained. The voltage source was primarily connected to the A3458A (Agilent with op02, last cal ~1.5y) and than leads have been switched to the K3458A (Keysight, without opt02, 2 months old) and than back again to the A3458A to capture some more data points. This is why there are small dips in voltage. Data was captured manually and I've left out the warm up phase of the reference from the plots. Reference is LTZ1000CH based. I think overall it works better than I've expected and with respect to the short time stability I was more thinking of ~30min. I do not have long term data, but I'm planning to repeat the measurements at some point.