Re: Advantest R6581 8.5 digit DMM mini teardown/repair
« Reply #1175 on: May 06, 2025, 08:38:16 am »
As proposed by Kleinstein i adjusted the preregulator replacing the 6K8 adjustment resistor by 1K8 and the first mains transformer now gets 209 Vac, still in 220 V configuration. All low voltage regulators have their headroom.
I implemented many other mods to the meter received from Ismsanmar, including Pomona copper binding posts at the front and a TFT display. Also the reference board was wired with soldered copper wires replacing board connector. Still want to recap the low voltage supplies and mod the input MUX.
In order to test that R6581T i connected it to our ADR1399 and a Keithley 3706. The log at 6 second intervals (100 NPLC with Autozero + 0.76 seconds for R6581T internal temperature measurement) exhibits a point to point standard deviation of about 3.5E-8. In other words: about 250 nVrms noise at this timing (roughly 0.05 Hz).
Internal temperature of the R6581T is about 39 °C at 20 °C ambient. Ambient temperature coefficient is 0.06 ppm/K for the R6581T, yet there are additional deviations depending on speed of temperature change. About 2 uV shift after turning on heaters with a 2 °C ambient temperature rise within 30 minutes.
Image shows one small R6581T malfunction observed during two days of logging, reason unknown (K3706 noise larger than deviation).

Regards, Dieter
« Last Edit: May 06, 2025, 11:11:20 am by dietert1 »
 
The following users thanked this post: Mickle T.

Offline dietert1

  • Super Contributor
  • ***
  • Posts: 2997
  • Country: br
    • CADT Homepage
Re: Advantest R6581 8.5 digit DMM mini teardown/repair
« Reply #1176 on: May 09, 2025, 05:34:30 pm »
Meanwhile i replaced the 4x 2K2 input protection resistor by a depletion mosfet pair like shown here:
https://www.eevblog.com/forum/metrology/ultra-low-noise-short-term-stable-references/msg5699907/#msg5699907. At small currents it has about 200 Ohm resistance and it should resist about 600 V.

Observations:
- Up shift of the 10 V measurement of our ADR1399 reference by about 3 uV.
- Instrument 10 V range TC is now 0.41 ppm/K
- Stability after numerical TC correction improved a lot. No more drifts depending on ambient temperature change speed.
- R6581T internal temperature readings are less noisy

According to Mickle T.s schematics the modded circuit is not involved with reading the LM35 internal temperature sensor, so there is some mystery. Diagram shows period before and after mod. The mod plus some hours of R6581T warm-up are hidden at the step near reading No. 6000.
To be tested: Null meter noise of the R6581T after mod.

Regards, Dieter

Edit: The diagram doesn't show ambient temperature (about 21 °C) but R6581T internal temperature.
« Last Edit: May 12, 2025, 07:00:13 pm by dietert1 »
 
The following users thanked this post: Mickle T.

Offline Kleinstein

  • Super Contributor
  • ***
  • Posts: 17204
  • Country: de
Re: Advantest R6581 8.5 digit DMM mini teardown/repair
« Reply #1177 on: May 09, 2025, 06:14:13 pm »
It could make sense to still have some resistance / PTC in series with the depl. FETs. Under ideal conditions the FETs work well to limit the current, but if they fail, they fail short. So some backup in the protection would be a good idea. ESD spikes may well exceed 600 V and there extra resistors with good peak ratings could help.

One the downside the FET version may be more prone to thermal EMF from thermal gradients. In the slightly different version with enhancement FETs and PV OK to turn them on, I have seen a few µV of thermal drift. Depending on the thermal environent this could also get worse.
Just simple resistors are also not all that perfect in that respect - some may show quite some thermal EMF in a temperature gradient. Ideally the 4 resistors in series should somewhat compensate.

The expeted effect could be a change in the offset drift, but hardly and change in the gain or gain TC.

The current limit part can also effect the charge spike from the AZ switching. In some cases this can be quite some current and the protection resistors can be part of what limits the current.  I don't think the 4x2.2 K would be that relevant for the meters noise - if at all in the 100 mV range (the amps parts uses a different path anyway. The zero path seems to have some 8 K for symmetry. 
 
The following users thanked this post: dietert1

Offline dietert1

  • Super Contributor
  • ***
  • Posts: 2997
  • Country: br
    • CADT Homepage
Re: Advantest R6581 8.5 digit DMM mini teardown/repair
« Reply #1178 on: May 10, 2025, 06:04:34 am »
So, in order to complete this mod one should replace R021 8K2 used for AutoZero by a 200 Ohm resistor and the 1500 V gas discharge tubes E002 and E003 by the 500 V version. Anyway the R6581T front panel indicates "300 V max".
For the LM35 temperature sensor one could insert a RC low pass filter that protects its output circuit from the MUX switching spikes. TI specify a 2 Ohm max output resistance up to 1 mA load without telling the bandwidth.

Regards, Dieter
 

Offline Kleinstein

  • Super Contributor
  • ***
  • Posts: 17204
  • Country: de
Re: Advantest R6581 8.5 digit DMM mini teardown/repair
« Reply #1179 on: May 10, 2025, 07:52:31 am »
I would use more than 200 ohms with the FET protection, more like 1 K to get a lower peak current. There is anyway quite some resistance from the JFET switches. AFAIK the JFETs in the custom switch hybrid are similar to 2N4117 or 8  and thus quite high resistance (e.g. 1-2 K range each). So there is quite some resistance in the input paths anyway. So no real need to go much lower in the input resistance.

Lower voltage gas discharge tubes would make sense.
 

Offline dietert1

  • Super Contributor
  • ***
  • Posts: 2997
  • Country: br
    • CADT Homepage
Re: Advantest R6581 8.5 digit DMM mini teardown/repair
« Reply #1180 on: May 22, 2025, 08:53:00 am »
When thinking about the LM35 temperature sensor noise, i remembered that in our other R6581T meters i modded it from 15 V supply to 5 V supply in order to reduce self heating. After implementing the mod, temperature noise increased, as the mod implies lifting the LM35 supply pin (1 of 3 pins) and that caused a higher thermal resistance to the PCB. Then i used a short piece of shrink wrap tube to protect the sensor from a possible air flow and it really reduced noise to almost zero. The resolution of the ":SENS:ITEM?" readout is 0,01 °C and the standard deviation of point to point differences is 0.0068 °C now.
Seems like some air flow from the fan enters the guarded section of the meter!

Regards, Dieter
« Last Edit: May 22, 2025, 10:52:47 am by dietert1 »
 

Offline Kleinstein

  • Super Contributor
  • ***
  • Posts: 17204
  • Country: de
Re: Advantest R6581 8.5 digit DMM mini teardown/repair
« Reply #1181 on: May 22, 2025, 09:10:21 am »
One can get temperture variations not only from a fan blowing air, but also from oscillations / turbulance from convection flow without a fan.
It can help to avoid extra hot spots (like voltage regulators with relatively small heat sink).
 
The following users thanked this post: Mickle T.

Offline bsw_m

  • Frequent Contributor
  • **
  • Posts: 300
  • Country: by
Re: Advantest R6581 8.5 digit DMM mini teardown/repair
« Reply #1182 on: March 14, 2026, 08:02:56 pm »
I discovered a problem with the voltage reference module connector completely by accident.
The connector soldering looked perfect. The problem was discovered after desoldering the connector and attempting to re-solder it. I desoldered the connector to verify the GND connections on the module.
I am also attaching the corrected module diagram.
 
The following users thanked this post: IanJ, Mickle T., TheDefpom, serg-el, ZhuraYuk

Offline leighcorrigall

  • Frequent Contributor
  • **
  • Posts: 487
  • Country: ca
  • Nuclear Materials Scientist
MASc, EIT, PhD
 
The following users thanked this post: bsw_m

Offline bsw_m

  • Frequent Contributor
  • **
  • Posts: 300
  • Country: by
Re: Advantest R6581 8.5 digit DMM mini teardown/repair
« Reply #1184 on: March 14, 2026, 08:46:15 pm »
During the final stage of connector restoration, blowing it with compressed air—the coating was found to be peeling in other areas. Everyone will have to draw their own conclusions about the problem.
 

Offline bsw_m

  • Frequent Contributor
  • **
  • Posts: 300
  • Country: by
Re: Advantest R6581 8.5 digit DMM mini teardown/repair
« Reply #1185 on: March 15, 2026, 02:43:24 pm »
Another schematic correction for reference module. Attention to R10. Looks like there's another bug in this device.
 
The following users thanked this post: Mickle T., serg-el

Offline dietert1

  • Super Contributor
  • ***
  • Posts: 2997
  • Country: br
    • CADT Homepage
Re: Advantest R6581 8.5 digit DMM mini teardown/repair
« Reply #1186 on: March 15, 2026, 02:58:46 pm »
That missing resistor with wrong wiring is a proposal of the LTZ1000 datasheet for TC fine tuning. I think the correct wiring is from the heater variable output voltage to the 120R resistor R1.
In practice it is quite some work to find the correct value for that resistor. 390K or 400K appears to be an average value. The best value (for TC = 0) depends on the thermal resistance of the LTZ1000 mount, e.g. on PCB slots, PVC covers, location of heater drive transistor.

Regards, Dieter
 

Offline bsw_m

  • Frequent Contributor
  • **
  • Posts: 300
  • Country: by
Re: Advantest R6581 8.5 digit DMM mini teardown/repair
« Reply #1187 on: March 15, 2026, 03:06:59 pm »
I started investigating the reference voltage after measuring its TC, which turned out to be high. This revealed problems with the connector and an error in the connection of resistor R10, which is used to fine-tune the reference voltage's TC.
 

Offline dietert1

  • Super Contributor
  • ***
  • Posts: 2997
  • Country: br
    • CADT Homepage
Re: Advantest R6581 8.5 digit DMM mini teardown/repair
« Reply #1188 on: March 15, 2026, 07:08:35 pm »
Be aware that your LTZ1000 TC is negative (about -0.25 ppm/K) and that most LTZ1000 require compensation of a positive residual TC. The proposed circuit with R10 will certainly decrease TC, as the heater voltage and power will decrease with ambient temperature. There is an alternative linear TC compensation circuit  that many other designs use. It has an about 15 Ohm resistor in series with the LTZ1000 zener and reduces TC as well, as it amplifies Ube of the compensation transistor.

So there may be something wrong in your DUT. Maybe the heater reference voltage divider.

Regards, Dieter
« Last Edit: March 15, 2026, 07:10:27 pm by dietert1 »
 

Offline branadic

  • Super Contributor
  • ***
  • Posts: 2606
  • Country: de
  • www.qoool-sensing.org
Re: Advantest R6581 8.5 digit DMM mini teardown/repair
« Reply #1189 on: March 15, 2026, 07:12:34 pm »
bsw_m, the result of your t.c. measurement looks rather odd and uncommon. I supose that's before you have solved the connector issue?

-branadic-
Measuring is like guessing, but more advanced.
 

Offline bsw_m

  • Frequent Contributor
  • **
  • Posts: 300
  • Country: by
Re: Advantest R6581 8.5 digit DMM mini teardown/repair
« Reply #1190 on: March 15, 2026, 08:49:24 pm »
The soldering of the connector could not have affected the measurement, since the voltmeter connection for monitoring the output module was made in place of the not installed C5.

I also provide a new TC graph, after adjusting the connection point of resistor R10 (along with the old graph). TC better but not enough.
« Last Edit: March 15, 2026, 08:52:23 pm by bsw_m »
 

Offline branadic

  • Super Contributor
  • ***
  • Posts: 2606
  • Country: de
  • www.qoool-sensing.org
Re: Advantest R6581 8.5 digit DMM mini teardown/repair
« Reply #1191 on: March 15, 2026, 09:17:41 pm »
Even this t.c. graph looks odd. It would be good, if you'd provide more information on your temperature profile and run a full sweep, not just a sweep from room-temperature upwards, but down to room temperature again and to temperatures below room temperature as well.

-branadic-
Measuring is like guessing, but more advanced.
 

Offline bsw_m

  • Frequent Contributor
  • **
  • Posts: 300
  • Country: by
Re: Advantest R6581 8.5 digit DMM mini teardown/repair
« Reply #1192 on: March 15, 2026, 09:40:34 pm »
Even this t.c. graph looks odd.
-branadic-
T.C. graph may be looks odd on start since I don't have access to the thermostat right now, heating occurs naturally after the fan in the device being tested is turned off.
 

Offline dietert1

  • Super Contributor
  • ***
  • Posts: 2997
  • Country: br
    • CADT Homepage
Re: Advantest R6581 8.5 digit DMM mini teardown/repair
« Reply #1193 on: March 16, 2026, 12:26:11 am »
Can you describe your setup?
Do you use a second meter and/or an external reference?
What is the source of the temperature parameter?
 

Offline d-el

  • Frequent Contributor
  • **
  • Posts: 253
  • Country: ua
Re: Advantest R6581 8.5 digit DMM mini teardown/repair
« Reply #1194 on: May 15, 2026, 10:48:38 pm »
Please tell me what the problem could be. I connect a reference resistor using Kelvin probes, enter external calibration mode, and set the value to 10.00035kΩ.
After the calibration is completed, I press the Home button and see a measured value of 10.00175kΩ. If I enable OHM COMP, it shows 10.001014kΩ.
What could be causing this issue?

I also calibrated the voltage by applying 10.0000331V, and after calibration it reads exactly 10.0000331V.

Offline Mickle T.

  • Frequent Contributor
  • **
  • Posts: 545
  • Country: ru
Re: Advantest R6581 8.5 digit DMM mini teardown/repair
« Reply #1195 on: May 16, 2026, 06:49:59 pm »
The external resistance calibration procedure's microcode is relatively complex and involves mathematical processing of eight measurements. However, it seems to me that the search for the source of the problem should begin with the cause of the large discrepancy in measurement results with OCOMP mode enabled. A difference of 75 ppm indicates the presence of a 750 µV zero offset in the circuit on the 10 V range, which is completely impossible for a properly functioning multimeter.
 
The following users thanked this post: bsw_m

Offline d-el

  • Frequent Contributor
  • **
  • Posts: 253
  • Country: ua
Re: Advantest R6581 8.5 digit DMM mini teardown/repair
« Reply #1196 on: May 17, 2026, 07:04:08 pm »
I figured it out. The problem was the huge reactive impedance of my reference standard.
I connected a C5-61 resistor and performed the calibration successfully. The values entered during calibration now match the measured values.
With compensation enabled and disabled, the difference is only 0.0000007 kΩ.

The photo shows measurements of an unsuitable resistance standard and measurements of the C5-61.
 
The following users thanked this post: bsw_m, serg-el

Offline Jeffers0n

  • Newbie
  • Posts: 3
  • Country: ca
Re: Advantest R6581 8.5 digit DMM mini teardown/repair
« Reply #1197 on: September 19, 2026, 09:41:26 pm »
Hello everyone,

Please excuse and inform me of anything I should keep in mind in future posts, as this is my first time posting on the forum. If this is not the right place, please direct me toward a more appropriate posting category.
I am posting for a friend of mine, who is a mechanical engineer with a dead R6581D DMM (R6581 without AC measurements). The following is the error message.
"RUNNING SELF TEST  +230, "Internal zero error" "
When the unit is turned on via pressing the power switch, the initialization of the CPU is performed, a beep is heard, and we are greeted with this blinking message.
During boot and when the message is blinking, no front panel buttons seem to affect the operation of the DMM (i.e., unresponsive).

Backstory of this meter:

He bought the meter 10+ years ago, after a repair had been performed on an unknown location on the analog board, which, as I quote, he described to have "left a large amount of flux on the board".
To remedy this, at the time, he thoroughly cleaned and de-fluxed the analog board with IPA/Ethanol.
Whatever the prior fault was that induced this flux-covered repair had apparently been repaired at that time, and, after cleaning, the meter was in full "functional condition" (although I suspect the calibration was voided).
Following this, the meter was fully "functional" for a few more years to come, at which point it was put into storage for an unknown reason and not used for a few more years.
Last month, he turned on the meter to find this error symptom, which has since stayed unchanged.

What we have found so far:

The machine does not smell of magic smoke.
Non of the capacitors appear to be bulged or otherwise in distress.
The machine's analog section power rails seem to be up and accurate at the big power connector near the top right corner (-18V, -27V, +27V, -15V, +15V, +5V).
The lithium battery on the lower logic board closer to the screen measures 3.22V (although it was measured after an unknown period of runtime, during which the battery could have charged?)
The firmware UV-erasable ROM is version A02.
The fan spins up at power on.
There is no sound of relays clicking at boot or during the self test (Suspicious?)

What we have tried so far:
Cleaning and reseating all 4 ends of the two fiber cables.
Measuring the supply voltage of a few chips (e.g, op amps, Analog switches) around the board (all very within spec and present).

A few images of the board and internals have been attached at the bottom.

Do any of you wizards have an idea how to proceed, or any likely causes of this problem?
 

Offline Jeffers0n

  • Newbie
  • Posts: 3
  • Country: ca
Re: Advantest R6581 8.5 digit DMM mini teardown/repair
« Reply #1198 on: September 19, 2026, 10:11:32 pm »
I have done some further digging and found that the only mention of error 230 sees to be reply number 671 of this post, stating that shorting pin 14 and 15 of U214 (a DG442 analog switch) causes this error at boot. Interesting, maybe the cleaning degraded the capacitors near by, or there was some residue?
 

Offline Kleinstein

  • Super Contributor
  • ***
  • Posts: 17204
  • Country: de
Re: Advantest R6581 8.5 digit DMM mini teardown/repair
« Reply #1199 on: Yesterday at 08:24:27 am »
IPA may damage PS type capacitors, but I don't see them in this meter. Other capacitors should be safe.
The meter has largely normal, non latching relays and this way no switching needed for the initialization and much of the self test.
The internal zero error sounds like a pretty basic problem, that may stop the self test relatively early.

One may get an idea of how much of the meter is working by looking at the testpoints with a DSO. TP200 (ADC input) and TP202 (Integrator ouput) would be good points to look at.

If the issue gets more complicated and may get longer, maybe start a new thread under repair with a link to the post here.
 
The following users thanked this post: Jeffers0n