Hi,
I attached some measured points from our 3458As between the years.
HP3458A#1 -0.07ppm/day
HP3458A#2 -0.02pppm/day
(calculated with the formular from the service note. I think one should use a linear regression with all data points).
HP3458A#2 has the 4ppm option, but the stability from both units over many years of calibration is the same. Both meters are quite old HP ones.
@branadic: Think about the number of digits, please
. As you can see in the graphic attached, the drift isn't linear, it is not even monotonic. If I would choose a 10 day intervall (like you) and use the data from day 20 to day 30 for example. My values would be much smaller.