Hello again, Rax,
The digital board you have, and the one I just acquired, is an older model that uses two pairs of EPROMs (4x 512 kbits). The newer boards use a single pair of EPROMs (2x 1 Mbits). Unfortunately, removing the daughter boards to install 1 MEG EPROMs won't get you far. There is a pin-compatibility issue.
Before you begin, I would highly recommend backing up all of the data. The NVRAM can be dumped with the attached scripts. The NVRAM can be copied remotely over GPIB without disassembly.
Please note that they are untested. I had to make some quick modifications to the previous NVRAM dump script because firmware version 1.07 communication is terribly sensitive. I am leaving for vacation soon and I do not have much time to revise. There is another tested version earlier in this thread:
https://www.eevblog.com/forum/metrology/datron-1281-repairmaintenance/msg3876473/#msg3876473Next, you should copy the EPROMs. I would like to have a copy of version 2.03 (from August 16 1988) if possible. I see that you have them.
From the attached pictures, I will name out the order of the memory from left to right:
- TC5564PL-15 is responsible for the calibration constants, the serial number, and the line frequency (i.e., NVRAM). It relies on a backup battery (1/2AA 3.6V).
- 84256-12LL is static RAM (SRAM) for basic I/O and is hidden beneath the daughter board.
- 410408-A daughter board that hosts 290159 (TMM27512) and 290160 (HN27512) firmware.
- HD68000P8 microprocessor.
- 410408-A daughter board that hosts 290164 (TMM27512) and 290165 (HN27512) firmware.
- 84256-12LL is static RAM (SRAM) for basic I/O and is hidden beneath the daughter board.
Using the firmware available on the xDevs website, you can optionally upgrade:
https://xdevs.com/fix/d1281/#firmwareSome firmware will allow you to use old calibration data. However, in my case where the instrument was calibrated with version 1.07, I was unable to reuse the data. In this case, a blank TC5564PL is needed or an FM1808 adaptor board will have to be prepared either with a universal programmer or by going into the special calibration menu and 'clearing the NVRAM' (CLRMEM ALL, Wavetek 1281 User's Manual page 5-106). If you do this through the menu, then no serial number will be presented. To correct this, you can send a command called SERIAL which can be found on page 5-105.
I would suggest reviewing Appendix A (4-A5) where it shows a list of Corruptions (errors 2013 through 2022) and Non-volatile RAM Checksum Errors (errors 2110 through 2113).
I believe your error (i.e., 2022) means that you need to replace or upgrade your NVRAM as it is bad. Luckily, you can make a FM1808 adaptor board and replace it (pictured as a yellow board). The FM1808B is still available from your favorite electronics distributor. Remember to remove the battery before installing!
I hope this helps,
Leigh