Author Topic: Datron/Wavetek 4910 restoration  (Read 10608 times)
Offline nikonoidTopic starter
« on: April 27, 2018, 03:38:20 am »
I am just starting a restoration for Datron 4910.
The unit is from 1993 or 1994. It looks fairly clean on the inside but has some electrical fire smell to it, so I have not tried to power it up yet.
Right now I am on a search for a service manual and also a source for replacement SLA batteries.
Can someone help me there?
My unit came with 15 original 6V 2Ah batteries that are completely dead. It is very hard to find replacements on the same size. Worse come to worse I can rebuild the battery compartment to accommodate different batteries, but would prefer not to do that if possible.
I am planning to document the restoration process here.
The following users thanked this post: quarks, MegaVolt
Offline CalMachine
  • Metrology Nut
Re: Datron/Wavetek 4910 restoration
« Reply #1 on: April 27, 2018, 03:44:15 am »
Very nice!!!  I'm excited to see this  :popcorn:
Here are the batteries you will need.  There might be cheaper ones around, but it's what I picked up for our 4911 and will get for my own 4911, in the near future.
https://www.mouser.com/ProductDetail/Power-Sonic/PS-621F1/?qs=sGAEpiMZZMso%252bSMtTqBO3073LIrlST7LWiXQ%252beQmRkk%3D
I am also in search of the service manual
« Last Edit: April 27, 2018, 03:46:10 am by CalMachine »
All your volts are belong to me
« Reply #2 on: April 27, 2018, 04:25:13 am »
Thank you very much, CM. I will give them a shot.
The size is a little bit smaller than original and placement of contacts is different. Did you have to mount them on their side?
Offline TiN
  • Country: ua
    • xDevs.com
« Reply #3 on: April 27, 2018, 04:35:50 am »
This will be interesting, new Quad LTZ bonanza :)
I don't have much info on 491x, just the shameless plug of MM's work.
« Reply #4 on: April 27, 2018, 04:59:06 am »
That is an excellent article. I actually read it few hours ago. That article is literally the first thing that comes up in google when searching for Datron 4910.
Thanks MM!
Offline ap
    • ab-precision
« Reply #5 on: April 27, 2018, 05:57:31 am »
It is strongly recommended to replace all Tantalum caps, those started to blow up on mine one after the other, so I changed them all.
Metrology and test gear and other stuff: www.ab-precision.com
Offline quarks
« Reply #6 on: April 27, 2018, 08:59:42 am »
very nice, thanks for sharing
about the batteries, afaik they are from Powersonic
http://www.power-sonic.com
so you can ask them for a replacement type
« Last Edit: April 27, 2018, 09:07:22 am by quarks »
Offline ManateeMafia
« Reply #7 on: April 27, 2018, 11:32:41 am »
There are some additional pics that didn't make it into the article.
I found a bad regulator that was hidden below the pcb with the row of 15-pin connectors.
I will see if my editor ever got them and maybe post a few to the article.  ::)
Offline Dr. Frank
« Reply #8 on: April 27, 2018, 12:19:50 pm »
Would you mind to determine the oven set point & temperature of each of the LTZ modules?
Although there exists a roughly measured schematic reproduction of the VISHAY resistor array, I doubt that it is fully correct, as probably being measured 'in circuit'.
The divider ratio can best be determined by a voltage measurement on the - input of the LT1013, and the real temperature by fast monitoring of the BE diode of the reference during warming up.
Thanks.
Frank
« Last Edit: April 27, 2018, 06:18:37 pm by Dr. Frank »
Offline CalMachine
  • Metrology Nut
« Reply #9 on: April 27, 2018, 03:40:40 pm »
Yeah the sizing isn't exactly the same.  This size is what was included with the unit when it was purchased, and they all needed to be replaced, so I just ordered the direct replacements.  One of the thermistors that monitors the temperatures of the battery banks had to be replaced, as well.
https://pasteboard.co/HizvlUj.jpg
https://pasteboard.co/HizvV1f.jpg
« Reply #10 on: April 27, 2018, 04:11:37 pm »
Although there exists a roughly measured schematic reproduction of the VISHAY resistor array, I doubt that is fully correct, as probably being measured 'in circuit'.
Thanks.
Frank
Dr. Frank, I can definitely do the measurements, once I get the beast working. I have a feeling that it had not been on for at least 10 years.
I will PM you for details on measurements.
« Reply #11 on: April 27, 2018, 04:22:03 pm »
CalMachine, I just ordered Power Sonics. One of you pictures look like an acid damage to the connector. Scary!
I am a little bit concerned about putting foam in between batteries, as it will trap the heat. I am considering making thin rubber rings outside of each battery to build it up to the size of original batteries.
Strangely in my unit not only two ribbon cables were disconnected but also each of battery banks was disconnected at a thermistor and clip held in place with tape. Someone really prepped the unit for long term storage.
Guys, thank you for suggestions. My next step is to disassemble and inspect all boards while I am waiting for batteries.
Offline CalMachine
  • Metrology Nut
« Reply #12 on: April 27, 2018, 04:36:51 pm »
Unfortunately our 4911 didn't come with the black plastic battery separator below the metal top, so I had to figure out a way to keep the batteries from moving around too much.  You're right about the extra heat, and something that I was contemplating.  hopefully, the thermistors will compensate the battery supply voltage properly as the temperature fluctuates.
I just got the idea to 3d scanning/printing up the one in our 4910, so I can remove the foam in the 4911
« Reply #13 on: April 27, 2018, 05:10:35 pm »
That is a great idea with 3D printing, CalMachine!
The file can even be customized to fit PS-621 batteries perfectly.
Offline CalMachine
  • Metrology Nut
« Reply #14 on: April 27, 2018, 05:43:14 pm »
Very much so!  Looks like I've got another project on my hands as well :P
Offline quarks
« Reply #15 on: April 27, 2018, 08:28:27 pm »
I guess I already posted what you are looking for, have a look here
https://www.eevblog.com/forum/metrology/ultra-precision-reference-ltz1000/msg273274/#msg273274
« Last Edit: April 27, 2018, 08:30:13 pm by quarks »
Offline Pipelie
  • Country: cn
« Reply #16 on: April 28, 2018, 10:36:11 am »
I removed and measured the vishay 315532  six years ago.
and here is the result:
there are at least five resistors in the TO-CAN.
PIN:7-8->21Ω
1-7->39k
2-3->127Ω
4-5->659Ω
5-6->9k
315532 measured by hp34401a, 4Wire
PIN 1-2:48.7795K  1-3:48.6536K  1-4:48.6543  1-5:47.9947K  1-6:38.9950K  1-7:38.9952K  1-8:39.0165K
2-3:127.422ohm  2-4:127.326ohm  2-5:786.248ohm  2-6:9.78619K  2-7:9.78629K  2-8:9.80740K
3-4:0.3380ohm  3-5:659.267ohm  3-6:9.65917K  3-7:9.65927K  3-8:9.68048K
4-5:659.163ohm  4-6:9.65910K  4-7:9.65920K  4-8:9.68033K
5-6:9.00017K  5-7:9.00028K  5-8:9.02140K
6-7:0.3355ohm  6-8:21.5500ohm
7-8:21.4520ohm
Offline GEOelectronics
« Reply #17 on: April 28, 2018, 09:29:14 pm »
Would the Daytron 4910/4911 Users Handbook help? Its an 80 page document.
Is there a FILE storage area on this chat group, can I post a 5.7M file to a regular post or should it go on my web page and post a link here?
Thanks for the help.
George Dowell
« Last Edit: April 28, 2018, 09:33:29 pm by GEOelectronics »
« Reply #18 on: April 29, 2018, 01:48:45 am »
George,
The files are already here. https://doc.xdevs.com/doc/Datron/4910_4911/ but it never hurts to have an alternate location.
Cheers.
Offline texaspyro
« Reply #19 on: April 29, 2018, 03:08:26 am »
George,
Cheers.
Like ko4bb.com
« Reply #20 on: April 30, 2018, 03:46:27 am »
George, I think we already have the document you are referring to and this is the same document on xdevs server and ko4bb.
I am slowly progressing with the project.
I started with power supply. Taking it apart was not difficult. Under metal shield I found a toroid transformer, generating 5 isolated 26V outputs plus another line out of about 11V. Here are few quick pictures. I will take high resolution photos and post them later on.
This is my current understanding of pinout of connectors coming to battery compartment. Please note that I numbered pins in a way how they correspond to pins on IDC connectors on power supply board. This is not a standard numbering for DB9 and DB15 connectors.
There are 5 channels of batteries with 3 batteries in each. Each channel has a 10k thermistor at positive terminal of the first battery (all connected to DB15 connector). First channel also has a thermistor on negative terminal of the last battery. I have not traced how it is used yet. Solid dots on DB9 diagram are negative terminals for each of 5 battery channels.
Without service manual I was not really sure what to expect for power output from supply. There are two output power IDC connectors on power supply: 10 pin and 16 pin.
For now I recorded all pairwise voltages between them:
20 VDC is directed at charging each battery channel. 24.5 VDC is actually 4.5 VDC relative to 20 VDC that comes to each thermistor. So far this makes sense. However areas in yellow look a little bit out of the pattern to me, so this is what I will be investigating next. I want to be sure the power supply is not going to destroy something before I move further.
Offline Pipelie
  • Country: cn
« Reply #21 on: May 03, 2018, 05:49:03 am »
hello, nikonoid,
my 4912 didn't come with the 'connector board', I asked my friend to help me with the circuit recovering. and then I can rebuild the connector board,
in order to replace the SLA battery to 18650B battery, I remove the NTC resistor and replaced it with a resistor network,  show as the picture below.
this NTC resistor is part of the feedback resistor of the regulator and causes the output voltage of the charger circuit to decrease as the battery temperature increase. but I'm NOT sure the purpose of the extra NTC.
« Reply #22 on: May 03, 2018, 04:42:26 pm »
Thank you, Pipelie. This is a clever approach in switching batteries to 18650.
I am still waiting for replacement SLA batteries. I will be working on it next week and will compare your schematics against my connector board in battery compartment.
Sent from my iPhone using Tapatalk
Offline Dr. Frank
« Reply #23 on: May 04, 2018, 08:09:14 am »
PIN:7-8->21Ω
1-7->39k
2-3->127Ω
4-5->659Ω
5-6->9k
7-8:21.4520ohm
Thank you.
Then the divider ratio of the oven controller would be 9k over 659 = 13.7, which would yield about 75°C oven temperature, which is way too high.
The Fluke 732A/B and the Datron 7001 are all running on about 45°C, so I would have assumed that the 491x ref would do the same, to achieve a stability of around 1ppm/yr.
In the manual for the 4910, page 1-4, they describe the oven temperature as being ' a few °C above the instruments maximum ambient operating temperature.', which is specified on page 4-1: 'Temperature: Operating:  0° to 40°C'.
So also from that numbers, a 50°C oven temperature is much more probable.
Maybe there's a detail in the schematic, been overlooked up to now, which reduces the oven temperature.
So I would like to ask you, to really measure the oven temperature.
Thanks.
« Last Edit: May 04, 2018, 08:11:23 am by Dr. Frank »
« Reply #24 on: May 09, 2018, 04:42:22 am »
Batteries came. I ordered powersonic PS-621, like CalMachine suggested. Surprisingly new battery is about half the volume of old, while caring the same 6V 2Ah spec:
I have finished checking out the power supply and found nothing wrong with it.
Then while testing the connector on the back of the unit responsible to monitoring battery voltages from outside, I found that all negative terminals of all 5 batteries are tied together. This is happening on main board (not on the battery compartment side or power supply side).
Is there any good reason why 5 isolated channels would have all their negative sides tied together? Or is it an indication of a possible fault?
At this points I am ready to dig into main board. I can see from the side view that I have at least two capacitors that are leaking. I am not 100 percent sure how to remove main board. Any recommendations on the procedure?