Others use active current sources. The Advantest R6581T schematic shows active current sources for the unknown and for the principal rundown current. In the Prema 6048 the unknown is a current source, too. In some sense the rundown current source is active, too, as it includes a sense line to correct for the switch behaviour. Looks like active current sources are appropriate for a 8.5 meter.
To convert the run-down current source of a Prema 6048 into an active one for near ground current routing, one can use a dual opamp, a small mosfet and 2x 1K and 1x 10K resistors of an LT5400. The divider raises the reference from 7 to 7.64 V and then senses the current at 0.64 V. That is the headroom for the mosfet, the switch and maybe some kind of filter. One might use something similar for the unknown. A little more complicated if it needs to be bipolar.
Regards, Dieter