Here i have a result for our TEC chamber.
The Dr. Neumann TEC cooler is specified at 240 W. It takes about 10 A at 24 V.
Inner and outer fans currently run on 12 V (half speed).
An Advantest R6581T DVM is operating inside the TEC chamber, so the TEC
cooler is always in cooling mode in order to take out the power dissipation
of the DVM. During the data taking period of several days the inner
temperature sensor of the DVM was 38.5 +/- 0.1 °C while ambient temperature varied.
The TEC cooler runs at voltages of up to 7 V. As far as i understand i could
put another meter into the chamber, maybe even three of them when
using a mains voltage stabilizer with reduced output voltage.
The extrapolated measurement shows that without TEC drive, the inner
cooler temperature will be above the outer cooler, but only about 1 °C
or so. This is good for operational safety: It means that the DVM will
not overheat as long as the fans are running and as long as the TEC is
protected from heating (inhibit voltage reversal).
One disadvantage of the COTEK AE-800 power supply used for the TEC cooler
are its two small fans - they sing like dentist tools. Need to find
a quiet solution. Today i got chamber temperature control depending on
R6581A inner temperature sensor, will show results later.
Regards, Dieter