You can see the difference between the two 3458A in the Fluke 5440B plot. Both plots are measured simultaneously.
Hello,
I propose, that you run both 3458A against each other for several more passes, as you did with the other instruments. (see also hp journal 4/89)
These linearity tests are very sensitive to fluctuations / noise, especially, when you approach the 3458As region of resolution and linearity, on the order of 0.01ppm.
Finally, it would be great, if you would draw the difference between both 3458A, but not vs. the 5440B.
That would better display the non-linearity of the A/D of the 3458As, so that you get a real quantitative measure of the INL (is it really about 0.02ppm??).
Frank