I can add present-day deviation measurements from seven SR10xx boxes made between 1979 and 1993, and also make some general comments. Individual step deviations were measured at 21-23C using a (mostly) calibrated ESI 240 bridge and using a SR10xx step(s) of another box as the Standard for comparison. History of the boxes is unknown, and they came from different labs.
The Factory spread data for each box is from the factory measured table fixed to each box, and afaik are calibrated absolute values, but noting that measurement temperatures ranged from 20C to 24C, and there has been a subsequent 1990 ohm correction.
The 2026 spread results could show the influence of unknown stresses as well as ageing, and the 1k box was noticeably 'worse' than the others, whereas the SR1030 10 and 10k were pretty damn good.
Given each box has 12 resistors, it may be worthwhile to select a 'better' series sequence of steps when using a box as say a 1:2, 1:5 or 1:10 or 10:11 divider. That selection could minimise the ppm spread quite noticeably, especially if the 'end' step has low average deviation from the other series steps in the divider. Similarly, an obvious outlier step(s) may conveniently be located at step #1, or #2, or #11 or #12, and so be excluded from typical use, although the series and series-parallel ESI accessories don't allow any flexibility in step choice.
It would be enticing to hope that an SR1030 box showing a relatively low ppm spread nowadays has a better liklihood of steps aging in a symmetric distribution, such that a series, or series-parallel, or parallel connection of steps, or a selected step, was within a few ppm of original factory measurement.