Hello everyone
We have received the device from customer and the device was in off mode, we just connected in our lab for 5 days to get suitable temperature, but still  oven indicator working  and the output of 10V was 7V.
please advice me to solve this.
 

Offline branadic

  • Super Contributor
  • ***
  • Posts: 2390
  • Country: de
  • Sounds like noise
Re: Fluke 7001 Volt reference
« Reply #1 on: August 14, 2022, 08:04:55 am »
You can find reverse engineered schematics by user chekhov over here:

https://www.eevblog.com/forum/metrology/(3300)-wavetek-7000-the-hidden-gemstone/msg3452674/#msg3452674

Fluke 7001 is similar to Wavetek 7000 except for the DC/DC converter.