All you need for this test are two pieces of BNC cable and a BNC T adapter. Connect the T adapter to Channel 1. Set the impedance of Channel 1 to 1M ohm. Connect a BNC cable from one side of the T adapter to Channel 2. Make sure the cable can't flop around and make sure that the temperature won't change during the test. The length doesn't matter. A meter or two is fine. Set the impedance of Channel 2 to 50 ohm. Feed a square wave from your 8040 to the open connector on the T adapter. I think the 8040 has a 1 PPS output. That would be perfect. Set the triggering on both channels appropriately for the input signal. Now measure Time Interval and collect the data with Timelab. Since there's nothing between Channel 1 and 2 but a piece of cable, the delay should be very stable. Any variation will be due to noise or limitations inside the counter. The resulting ADEV graph will show you the best results you'll ever see from that counter.
A good rule of thumb is that the ADEV @ 1 sec. should be about the same as the minimum time interval resolution of the counter. For the 53131A, that would be about 5e-10 @ 1 sec. Since most good quality counters exceed their specs, you might do a bit better.
If your results are much better than spec, it means your counter is making multiple measurements and averaging them. You have to disable that 'feature' before any of your measurements will make sense. If your results are much worse than spec, you may have a setting wrong on your counter or a fault in your counter or your 8040.
This isn't a long test. If you watch the ADEV graph as the data is being collected, you'll see that the 1 sec. value flops around to begin with, but stabilizes within a few minutes. You can stop the test there. Sometime when it's convenient, you can let the test run longer to see how far down you can go. Eventually, the graph will become a flat, horizontal line, but that's usually so far down that there isn't any point in measuring it.
Ed
Ed-
When you said time interval for this test doesn't it meant the time interval port 1 to port 2? I assumed that's what you mean since I have both ports hooked up with the cabled between them and T on port 1 with the Datum 8040 feeding it. Have port 1 and 2 set to auto trigger. My 8040 doesn't have a 1 PPS output only the 10MHz output so that's what I used. I don't see any averaging settings. It looks like hooked up this way the live display shows time interval bouncing around between .0001 µs and .001 µs. I'm running the time interval again with the 50Ω Pomona cables I have now. I'm afraid it's going to be better than when I ran the test with the video BNC cable
Bill
Bill, don't get confused between a normal measurement and the noise floor measurement with the BNC T adapter. The noise floor test is a one-time thing that tells you that the counter is working properly. It tells you nothing about the signal that you're feeding into the BNC T adapter. You did that test, you got a good result showing a 1 sec. value of about 5e-10. Now lets move on to normal measurements.
I dug out my 53131A, dropped it on the bench, powered it up, and connected it to GPIB. I did a noise floor test as shown. Then I did a quick data run with my 5065A to Ch. 1 and Tbolt to Ch. 2. Measured Time Interval as shown. I made no attempt to filter or optimize either of these tests. I changed nothing in any of the boxes on the Timelab Acquisition screen. I used the 'monitor' function to set the sampling interval. Since both inputs were 10 MHz, the 'Input Frequency' was left at the default value of 10 MHz.
As expected, the noise floor came out slightly lower than the live measurement. These are the kinds of results that you should see with either a Rb or GPSDO. You should not see 1e-2 or 1e-16. Neither is realistic. The sampling interval would have only a small effect on the graph.
Then I looked at the Acquisition screen you posted.
Houston, we have a problem.
My Acquisition screen is the same as yours except that where you are showing numbers like 1.000000042E-07, I'm seeing things like 8.7E-09. That makes me suspicious because you've got far too many digits. There has to be averaging going on because the values shown in the Incoming Data window are only changing by 1e-15. That's why your graph was so good. My numbers are varying by ~5e-10.
Is the yellow LED beside the STATS button lit?
Ed