I must admit, this sort of input protection on a DMM worries me. Sure, I know that precision bench meters don't often have a Cat rating, but this sort of 'brute force' input clamping isn't exactly going to do the operator any favours in an accidental / gross stupidity situation. I know at least one older Solartron meter that just has a 1.2kV GDT soldered straight across the input terminals.
As you say there are other ways - I keep coming back to Datron and their series string of big carbon composition resistors, mounted on PTFE standoffs, adding up to 88k at 12W dissipation, before clamping with zener+JFET combinations. This was used on the 1040/1050/1060 (/1070/1080???) series. This gives 1kV continuous protection even on the 10mV range [Edit: and very low leakage]. Only for DC of course (is this the hang-up?).
I realize that, at some voltage, something is going to flash over to Guard and/or Chassis in a low impedance way, but it does seem a more 'public spirited' form of input protection.