For an old instrument even slow aging should have come to an end, except if it depends on operating temperature and the instrument was essentially off all those years. Did you think about trying a TEC pile on that ADC hybrid U180? I mean if the firmware of the 3458A does not support temperature and drift corrections, maybe you can do it by chip temperature control. If aging depends on chip temperature, one could try running that chip a little cooler.
Before trying mods on an expensive instrument i would decide upon a certain goal in terms of ppm/year or so.
Regards, Dieter
Hello Dieter,
Midi already wrote it as it is: No DMM is intended and accepted as (secondary) standard.
The 3458A is used in metrology mainly as a ratio machine for transfers, due to its ultra linear A/D.
Its voltage reference is neither optimized for timely, nor for temperature drifts.
As it's got 90°C oven temperature, drift would be usually about -30ppm/yr, but the complete modules are monitored and selected for annual drift of 8ppm, 4ppm or until recently 2ppm/yr.
The standard reference will usually drift less, like 2ppm/yr., after one or two years, it's rumored.
But it will not approach zero.
All drift specifications are valid for continuous operation, so it's probable, that the drift is very low when not in use, and if no hysteresis effects occur.
But even real voltage references, like the 732A/B/C, or the M7000, running only at 45..55°C, mostly show a constant drift, after years of operation. The LTZ1000 drifts typically -0.8ppm/yr., and might settle down to - 0.5 ..-0.2 ppm/yr.
hp did not even trim the module for zero T.C., so the, I think, 0.3ppm/K are quite mediocre.
With a little bit of effort, it could have easily be tuned to 1/10 of that value.
The U180 has additionally about 0.5ppm/K, which can easily be compensated by the ACAL procedure.
hp did not intend to design a real metrology instrument, that's obvious.
So a bit of tuning can be done, like decreasing the oven temperature, maybe compensating the T.C. of the LTZ circuit. But then the story ends, and you better get a real 10V standard, which you might use for comparison, or for frequent calibration of the DCV mode.
Frank