Author Topic: HP 4271B LCR Question  (Read 1016 times)
Online Wallace GasiewiczTopic starter
« on: October 22, 2021, 11:15:16 pm »
Hope this is the appropriate board for this question:
I recently rescued and repaired an HP 4271B digital LCR from a hamfest. Appears to work pretty well now.
There is a switch on the A11 board that grounds the shield of the L pot lead. Should this switch be open or closed????
I do have the service manual and am unable to find the answer to my question after a lot of reading.
The only other outer conductor that is grounded is the H Pot, all outer conductors are all isolated at the chassis mount
I am also looking to acquire or build a test fixture.
Wally
Online wn1fju
Re: HP 4271B LCR Question
« Reply #1 on: October 23, 2021, 12:25:56 pm »
Hmm, never noticed that switch on my unit.  I suppose you could do a measurement and flick the switch to see if it has any effect.  Perhaps that will give you a clue.  I don't think it will break anything.
For test fixtures, I bought a TH26001A which is a Chinese clone of the much more expensive HP version (see photo).  Works great.  About $70 on eBay.
You might post your question on the Repair board rather than the Metrology board.
« Reply #2 on: October 23, 2021, 01:40:47 pm »
The switch in question is on the A11 board under the black cover inside the unit
I believe the schematic shows it closed. I suppose that I will set it there. There is very little change in readings with it closed or open
I am definately considering the TH fixtures, thanks for the recommendation.