New post to avoid getting in trouble over on NU180" - A U180 drop-in replacement for the 3458A
A few questions:
Zondar said:
Where in the machine, or how, is the CAL175 temperature measured? I have a temperature probe on my previous board and find a substantial difference between that value and the temperature at the board (not unexpected). The temperature delta at the board is almost 2x the delta of CAL175 (which is distinctly higher in temperature, too). TCR's on the board would respond to the temperature at the board, obviously, rather than that of CAL175, so I wonder about pertinence.
What is the correct formula to compute temperature drift in ppm/C from CAL72? Is it just (cal72_1-cal72_2)/(temp1-temp2)*1,000,000? Adjust by dividing by cal72_1 as well? Is there another factor of 10 in there due to the 10V scale?
Also, what are people's experiences with the rising clock edge to date? I think most people are using the falling edge right now. Is it clear yet that we should stick with that?
I've wondered the same thing and also why have I not seen someone controlling the temperature instead of letting the tail wag the dog?
My former "parts unit" now workingish, that has an illegal Noctua NF-A6x25 FLX and runs 5 °c cooler than the hotter of the 2 other meters with Papst fans.
(Illegal=Noctua 12V fan running on the 15V unmodified 15V reg out)
Seems like a smart fella could generate a tempco table*, stick it in a micro, and modulate the fan speed (and maybe a gentile heater in the airflow). Seems like Bob's your uncle.
EBM Papst 162L spec 12CFM @12V 26 dBA
Noctua NF-A6x25 FLX spec 17.19 CFM @12V (quieter too) 19.3 dBA
* (if you are fancy, use some coefficients, or other math magic, or God forbid: Analog.)
TEMP? is measured via U21, on pdf page 50 in the CLIP, A1 board, top, left. That's an LM35DZ, please look for yourself, where it's localized on the PCB.
Don't fiddle around with the fan, as the 3458A is very delicate with varying temperature, if you want to go into sub ppm range, where it's intended for. Papst is fine.
You need to keep the room temperature absolutely constant, TEMP? will strictly follow with 12 ..20°C temperature rise, depending on how you set up your 3458A, i.e. on table top (best), or in a rack (worst), Best, because you have free air flow from the intake and out of the slots.
I run my lab mostly between 21 .. 23°C over the seasons, but constant to < 0.2°C over 24h, during a stability measurement. That feature of the basement room is much more important than any thermal experiments with the 3458A itself.
Keep the filter of the fan clean, and TEMP? as low as possible. Mine has always 33 ..35°C, which also gives longer life to these ELANTEC comparators.
Frank