In order to make valid readings on any given parameter, all other parameters should be held as constant as possible. There is no way to measure a given parameter when you have a bunch of variables interacting at the same time, this always leads to inaccurate results and mistaken conclusions. If you want to find out what effect humidity has a on a given component, then all other major parameters that may have a significant effect on the component must be controlled. I have not seen this with most of the measurements that have been posted, this leads to incorrect conclusions about what a given parameter has on a given part.
As noted here, temperature is a major parameter, therefore it is one of the parameters that must be controlled. You should preferably know which parameters affect the type of component you are measuring, for example Manganin (and its derivatives) are known to be a bit sensitive to barometric pressure, Evanohm not so much. Manganin is slightly affected by moisture, Evanohm isn't. The type of construction may also determine a sensitivity to the environment, most molded packaged film/foil resistors are sensitive to humidity, epoxy resistors vary depending on who made them.
Usually the two most significant parameters that resistors respond to are temperature and stress, stress being a component of heat and winding techniques or die attachment as in film/foil. These two parameters are difficult to separate in many cases as they both cause a shift in resistor value. Generally speaking, internal stress tends to produce a non-linear curve in the resistance, unfortunately many resistors do not have an inherently linear TCR so again, it can be difficult to separate the two from each other.
Don't forget, there are also measurement uncertainties which can vary depending on the circumstances, instrument drift and a pile of other small errors that can sneak into your measurements without you noticing them. Measurements at PPM and sub-ppm levels are not trivial to accomplish and just because an instrument may display a measurement to .01PPM resolution doesn't mean it is also accurate, uncertainty wipes out most anything below 0.1- or 0.2- PPM at the very least, there is also source noise, thermals and EMI to consider. Too often the output of a digital instrument is taken as absolute, it isn't and don't forget those inaccuracies are included and amplified in math calculations so be careful what you believe.