The noise specs of the ADA4530-1 are not much better than for the AD549L. Both are 4 µV_pp for 0.1 Hz-10 Hz. And current noise is only 0.07 compared to 0.11 fA /sqrt(Hz). It is mainly the offset drift that is really much better. Also bias current at higher temperature and thus bias drift is much better.
The input capacitance is quite a lot larger - so if important, the AC performance can differ.
Before a modification, I would test the ADA4530-1 in a separate circuit. At that current level the board design and insulation is important, and the AD549L is likely not in a SOI8 case to allow direct replacement any adapter will add to leakage.
.The noise specs of the ADA4530-1 are not much better than for the AD549L. Both are 4 µV_pp for 0.1 Hz-10 Hz. And current noise is only 0.07 compared to 0.11 fA /sqrt(Hz). It is mainly the offset drift that is really much better. Also bias current at higher temperature and thus bias drift is much better.
The input capacitance is quite a lot larger - so if important, the AC performance can differ.
Before a modification, I would test the ADA4530-1 in a separate circuit. At that current level the board design and insulation is important, and the AD549L is likely not in a SOI8 case to allow direct replacement any adapter will add to leakage.
The AD549 and AD4530 are FET based. So there current noise should not be higher at low frequencies. It is more that beginning from a frequency in the kHz range it goes up.
Besides current noise from the chip itself there could be noise associated with leakage - this might show some 1/f part, but this could be comparable as it does not depend so much on the chip, more on the package / board or standoffs.
It might no be practical to operate the chips at 20 C - to get the humidity down one might want to have the chip more at 10 K above environment. Also for the AD4530 there should be the typical increase in bias with temperature, at least most CMOS chips show it. Only the level is lower so that one might not notice it so much.
Early this morning I've noted that the offset current on my K263 is about 37fA at the room temperature about 20C. It was almost 100fA yesterday evening when the lab was at about 25C.












. Even without opening the calibrator I've guessed that the output stage of the PWM is dead, so both Q2 and Q3 FETs are conducting all the time and the PWM output is essentially Vref divided by R13/R12 resistors. Either the FETs or U6 driver chip is faulty. Today I've opened the unit and it indeed looks like U6 (TSC426) is kaput. I should receive a replacement chip tomorrow and hopefully the 263 will be back in operation. Yesterday morning my (used 24/7) Keithley 263 developed a fault: the voltage source got stack at about half-scale 10.7V/1.07V/107mV depending on the output amp gain setting. Even without opening the calibrator I've guessed that the output stage of the PWM is dead, so both Q2 and Q3 FETs are conducting all the time and the PWM output is essentially Vref divided by R13/R12 resistors. Either the FETs or U6 driver chip is faulty. Today I've opened the unit and it indeed looks like U6 (TSC426) is kaput. I should receive a replacement chip tomorrow and hopefully the 263 will be back in operation.
Cheers
Alex
As the K263 performs rather better than expected, my plan is to improve the weakest point in this unit - the electrometer. The chip used - AD549L - was probably the best available at the time as a single chip solution, however the input current is in the area of 30-50fA and this creates a noticeable offset (and noise) in the active current sourcing mode. Also the AD549 does drift with temperature much more than the latest electrometer chips. From the look of it, the circuit allows for a simple upgrade with the ADA4530-1 . The error current can be reduced to less than 1fA and the temperature drift by over ten times. It also should improve the accuracy of the self-calibration routine and remove the null adjustment requirement (as the initial offset of the ADA chip is less than the temperature drift of the AD549 for 1C).
Cheers
Alex
. I will post results, when I did the update.