If you have something particular that interests you, I can set up and log something, B2987A is unused anyway.
1. 100MOhm resistance standard (not an ordinary resistor), preferably placed in a thermostat
2. Stable voltage source 10V.
Place 100MOhm standard between voltage source and electrometer input.
Write log measurement from B2987A within about 3..5 hours. Ideally if you also record the measurement log from the analog output (TIA output) of the B2987A
Why such measurements, I will tell a little later.
Out of latest data I used these for was triax connector leakage test with 20V bias.
I had one question, why, with a positive bias, the current went into the negative region. Why this happened, I have some thoughts. Here I just want to focus on this.