Author Topic: LM399 based 10 V reference  (Read 548269 times)
Offline guenthert
Re: LM399 based 10 V reference
« Reply #1075 on: January 03, 2020, 05:17:11 pm »
@Andreas: my conspiracy theory would be AD/LT is selecting somehow (before a burn-in) the LM399s based on popcorn or 1/f noise level (or Vref<7V?). They sell parts with "LM399" markings to the general public, and the parts marked with "weird numbers" to the selected customers. Guess who gets what..
  Not quite and reality is not quite as nefarious.  There are companies who offer a selection service and it stands to reason that AD/LT is one of those.  HP (erm, Keysight) is one of those major customers who pay extra to get units with tighter/guaranteed specifications (and a customized label).  Those units are selected from the general pool.  The open market gets those which don't quite meet those tighter limits, but also - and that might be the overwhelming majority - all those units which haven't been tested at all.  You're free to purchase a set and select the best and sell the rest on Fleebay (which I think some do).
Online iMo
  • Country: pm
  • It's important to try new things..
« Reply #1076 on: January 03, 2020, 05:44:41 pm »
An lawyer would say "KS is not using an LM399H/AH in their DMMs".. :)
Offline Andreas
« Reply #1077 on: January 03, 2020, 07:54:53 pm »
@Andreas: my conspiracy theory would be AD/LT is selecting somehow (before a burn-in) the LM399s based on popcorn or 1/f noise level. They sell parts with "LM399" markings to the general public, and the parts marked with "weird numbers" to the selected customers. Guess who gets what..
sounds familiar to me see:
https://www.eevblog.com/forum/metrology/lm399-based-10-v-reference/msg854683/#msg854683
so it must be true somehow.   >:D
with best regards
Andreas
The following users thanked this post: iMo
Offline David Hess
  • DavidH
« Reply #1078 on: January 03, 2020, 08:24:53 pm »
Exactly because the noise and drift tests are very expensive because of the time they take.  So most common devices only had the short functional test.
Online blackdog
  • Please stop pushing bullshit...
« Reply #1079 on: January 03, 2020, 09:15:28 pm »
I want to show the oven that I'm going to use for this LM399 reference, this because I already have quite a few test hours in it and that the "gain" is above 150x.
So 15C temperature increase gives less than 0.1C change in the oven.
How good it really gets depends on how well the insulation is done for the oven.
By placing the furnace in a simple polystyrene box, I got more than 150x gain.
I showed this oven before on this forum, and the controller is equipped with a LM723 IC.
And the NTC is located in the 7mm thick bottom under one of the power MOSFets that heat the box.
The extra TO220 IC is a LM35DT, I thought it would be easy to measure the temperature of the oven, but its a bit of a drama temperaturesensor, noise as hell.  :-DD
The oven is made from an already used HF box from which I removed the connectors.
It is a beautiful box for an oven because, of the very thick walls, the bottom plate on which the MOSFets are mounted is 7.5mm thick.
The battery is for the impression of the size of this oven.
Here you can see the extra 5K NTC measuring the temperature in the oven, this NTC was connected to a 34461A DMM.
It is clearly visible that I made sure that the wiring is mounted against the oven body.
This has been done to prevent you from making wrong measurements by leaking energy from the connection wiring.
During measuring, the lid was closed and the holes covered with tape.
This picture is for the impression how much space there is left for the Reference circuit, the parts around the LT1010 buffer do not come in this box.
That leaves some extra space to place the parts in an optimal way.
This because I'm still thinking about the remarks to leave out the bootstrap capacitor filter and only use the noise reduction of the 4x LM399 parallel.
I can then add more insulation to the LM399 IC in the oven, because no filter capacitors are needed.
I also thought about not connecting the oven in the LM399 IC's, but this oven is not good enough for that.
It requires an oven with a gain well above 1000x.
That's why this oven should be placed in a flask, or you could opt for a double oven.
As far as I'm concerned, both fall outside my project objective.
My first tests with the two types of Nichion VR(M) 470uF capacitors are good.
Eventually the 35V versions look good after first forming the capacitors at 15V and after that I started measuring leakage current around the 7V voltage which is the reference level.
One of these 35V types is at less than 1uV at 7V across the 10K resistor series.
So this is not the bootstrap filter but 1 single capacitor!
I've been thinking about "maat" his remarks about the strange drift of the bootstrap filter he tested.
My experience is that it works well but I also came across some weird effects today.
That wasn't the filter, but the cause was the power sources I used for the tests.
Those were two different power supplies, an old Harrison and the Rigol DP832.
Both can't be used properly because of the 1/F noise of these power supplies.
Today I picked up some different types of batteries I ordered from the wholesaler.
I first had to get these batteries to LAB temperature otherwise the drift of these batteries also caused too many errors...
Now the 4x penlite are quieter than my 10V reference PCB with the LT1021 on it.
This is a picture of the measurement on the 16V 470uF bootstrap filter with my old Philips PM3424 microvolt meter.
The DC value over the 10K resistor is less than 2uV.
Oh yeah, of course I regularly check if the "null" level of this meter hasn't changed.
And this is for the impression of my 10V Quad LT1021 LAB using a capacitor bootstrap filter.
The PostIt note is from June 2018 and the measured value on the 3458A is from this evening with a well warmed up 3458A and after an Autocal.
And below a link to the website of Rubicon with clear information about the properties of capacitors.
If you look at the leakage current between 22C and 42C, that difference is only small.
You have much more advantage in properly forming the capacitor and selecting the one with the lowest leakage currents than to worry about maybe 20% extra leakage current at 42C.
But, uh... still a lot to think about.
For the sake of completeness, I'm not looking for the perfect 10V reference, but try to do the best I can with the stuff I've got.
Kind regards,
Bram
« Last Edit: January 03, 2020, 09:21:45 pm by blackdog »
Necessity is not an established fact, but an interpretation.
The following users thanked this post: 3roomlab, Andreas
Online iMo
  • Country: pm
« Reply #1080 on: January 03, 2020, 09:32:12 pm »
Does an electrolytic (or tantalum) capacitor need a minimum DC bias in order to work properly (EDIT: to be properly "polarized")? I ask while the top capacitor in the bootstrapped combo sees "0V" best case.
« Last Edit: January 03, 2020, 09:43:51 pm by imo »
Online blackdog
« Reply #1081 on: January 03, 2020, 09:37:20 pm »
Hi imo,
Actually, I don't know...
I don't know what manufacturers say about this.
Maybe some forum members have more information on this?
Kind regards,
Bram
Offline David Hess
  • DavidH
« Reply #1082 on: January 03, 2020, 10:35:27 pm »
No, electrolytic capacitors work fine with 0 volts of bias.  Noise and leakage depend on the capacitor so they may need to be tested for it.
Online blackdog
« Reply #1083 on: January 04, 2020, 11:57:25 am »
Hi 3roomlab,
Do you want to tell us where the picture of the drift of the LM399 at different temperatures comes from.
Thanks
Bram
Offline FransW
« Reply #1084 on: January 05, 2020, 08:54:12 pm »
The drift rate at that temperature of 95°C is nowhere specified, but it is an Arrhenius calculation, according to e.g. P J Spreadbury: 'The Ultra-Zener.. is it a portable replacement for the Weston cell?' , Meas. Si. Technol. 1 (1990).
They demonstrated, that the drift of the LTZ1000 doubles with each 10°C increase of the oven temperature. At 55°C they measure typically -2ppm/year. 95°C would yield a 16 times higher drift.
So this mean one can expect about a 2 times higher drift rate for the LTZ1000 A due to the about 10 K higher set point needed. This could be a bit less if good insulation is used around the non A version.
Due to the square law for the heater, half the power needed means about 70% of the current needed. So there is some power saving for the A version possible, but not that much if external insulation is added.  A first point to save power and keep the drift low is choosing an oven temperature that is not too high for the planed use.
Just as a crazy idea: The heat lost from the transistor to drive the heater could in theory also be used to heat the reference - so maybe have the transistor on top of the LTZ instead of somewhere far away. This may need a modified compensation however, to really work at the low power end.
From: https://www.eevblog.com/forum/metrology/ltz1000ch-or-ltz100ach/msg2341581/#msg2341581
PE1CCN, Systems Engineering, HP, Philips, TEK, BRYMAN, Fluke, Keithley
Offline dietert1
  • Country: br
    • CADT Homepage
« Reply #1085 on: January 06, 2020, 08:50:46 am »
The measurements presented are not that conclusive to use them for extrapolation below 25 °C. Aging at 25 °C may be small because the parts used had been on shelf for years at room temperature (20 .. 30 °C). When you go below room temperature, aging may start anew.
Regards, Dieter
Offline Kleinstein
« Reply #1086 on: January 06, 2020, 09:31:15 am »
The normal theory suggests that aging is slower at low temperature. However the data shown for the LTZ refs are not very conclusive.
Another difficulty with aging is that there are different processes / parts that show aging. At a high temperature the faster processes may have reached the final values and one only sees the slow processes, while at low low temperature one would see the effect of the faster processes with the slow part essentially at a stand still. So this can be more of less independent effects, especially after some burn in process.
The LM399 is made with a fixed temperature - if not using the temperature control, one could as well use a different more conventional zener with higher current, like the old 1N829A or the Chinese 2DW232. The main point of the LM399 is that it makes it easy to use a heated zener. That lower temperature slows down aging is no a new finding. So when designing the LM399 they probably had a reason to choose some 95 C and not 70 C or 50 C. If it would be so easy to improve, I would expect a slightly different version with lower temperature setting to be available too.
Offline dietert1
  • Country: br
    • CADT Homepage
« Reply #1087 on: January 06, 2020, 09:35:37 am »
Ease of use is also the reason for choosing a temperature near 100 °C, otherwise you need humidity control.
Regards, Dieter
Online iMo
  • Country: pm
« Reply #1088 on: January 06, 2020, 10:08:41 am »
The LM399 and LTZ1000 have different internal wiring. While LTZ is basically a naked zener, the LM399's zener is embedded into a circuit with many transistors and resistors. That may influence the temp vs. aging too.  The zener's current in LM399 is ~0.25mA, afaik.
PS: the LM329 datasheet shows the Ref's schematics identical as the LM399's one.
PPS: Dobkin's paper shows the 30k and 10k resistors are trimmable, it could be they can set the zero TC for 329 and 399 accordingly.
« Last Edit: January 06, 2020, 11:59:06 am by imo »
Offline stijena1973
  • Country: hr
« Reply #1089 on: January 09, 2020, 04:19:26 pm »
I doubt that one can calculate individual noise from pairs.
Indeed you can :) Click :
www.intersil.com/data/an/an177.pdf
The calculation is not difficult!
The measurement is a pairwise sample DUT noise density over frequency. Because the assumption is that the samples are statistical independend the noise power at sa specific frequency is simply the addition of the noise power of the DUT pair.
If you have n=4 samples you can make n*(n-1)/2=6 pairwise measurements. I will call the reference sample spectral noise power at a specific frequency simply a,b,c,d.
What you measure with the pairs is the sum of the noise powers
a+b
a+c
a+d
b+c
b+d
c+d
You get the noise power of the first sample reference by the following expression:
a=(1/3)*(((a+b)+(a+c)+(a+d))-(1/2)*((b+c)+(b+d)+(c+d)))
a=(1/3)*((3*a+b+c+d)-(1/2)*(2*b+2*c+2*d))
a=(1/3)*((3*a+b+c+d)-(b+c+d))
a=(1/3)*(3*a)
a=a
For the general case n the expression is:
a=1/(n-1)*(sumwith(a)-1/(n-2)*sumnot(a))
The application note from LT does not give a calculation example. It is only stated that you have to use a much better reference as a DUT partner.
This method is called "three-cornered hat" method. Of course it can be four-cornered etc.
 But minimum is three, and pre-condition is that they are statistically indepedent.
Offline Kleinstein
« Reply #1090 on: January 09, 2020, 09:30:18 pm »
Calculating the noise of individual reference from measurements in pairs is rather difficult, as the noise is not constant in time  :-DD. So the individual noise amplitudes will scatter quite a bit. So the cornered hat version is not really practical, as the individual measurement will contain quite scattering.
There are usually 2 practical method to check the noise:
One is to amplify the noise with an AC coupled amplifier. So there will be lower frequency cut off, at maybe some 0.1 Hz.
The other method is using a known lower noise reference to subtract. A first approximation could be using the average of several (e.g. 5) references as the point to compare too. Once really bad units are out, one can assume the average is lower than the individual units.
Online ch_scr
« Reply #1091 on: February 14, 2020, 08:50:57 pm »
I have finally built my east-german-themed burn-in rig for my czech LM399-clones. Some western parts sprinkled in as per tradition.
928404-0
It is based on a B589(AD589 clone) Ref / OP07 / 1k shunt current sink and one quad NPN array / three six-fold p-fet arrays used as current mirrors to supply 15 current sources tied to one reference.
928400-1
The B341 and U105 IC's were made in east germany but there are still some on ebay. I tried my best to have star ground / supply where applicable. Kicad project zip attatched.
Maybe the best thing that has come from this folly is that I have found a socket that really grabs the references. And I mean grab, no voltage fluctuation because the reference does not move after insertion. I believe they are "precision sockets" (like the turned ones but with additional inserts). As they are from the parts bin, I have no name or number, but the pictures of precision sockets look like it. As I tried to match the pitch of the references on the pcb the sockets have to be broken apart, soldered in one-by-one and insulated with heatshrink...
How would you set the measurement cycle on a 34970A? I thought to dwell on one reference(call it #1) for like 8 hours, then switch through all other ref's (#2 to #15), measure voltage there for short amount of time and stop on #15 for the next 8 hours. Now measure #1 - #13 for a short amount and dwell on #14; and so on, to minimize relay actuation while having long and short term data on all references.
The following users thanked this post: Andreas
Offline Kleinstein
« Reply #1092 on: February 14, 2020, 09:09:59 pm »
The current mirrors from discrete transistors usually don't work very well. even for less precision needed, one should add some resistors at the emitter / drain.  Even with transistor arrays the mirrors tend to be not the precision wanted to test references.
The more normal circuit would be to wire one of the LM399 with an OP to create a reasonable stable 10 V and than use resistors from that 10 V level. Even if just an TL431 is used for the 10 V level, this may be more accurate than the current mirrors.
The heater part usually likes to have more than 10 V anyway.
For the reading cycle, I would consider just switching to the next reference at a regular rate, like every 10 or 100 seconds.
It kind of depends on the effort spend in separating the data later and the MUX used. The obvious choice here would be CMOS MUX chips like DG408 (or even old  CD4051), that are good enough for a 7 V level.
Online ch_scr
« Reply #1093 on: February 14, 2020, 09:24:47 pm »
The heaters are supplied by two 7815 . Good point about the emitter / drain resistors. I have some CD4067, that would allow fast cycling without stressing the relays in the 34970. Did not think of that, just took the relays as given.
Edit: Just realised, U105 has the drains connected inside the ic - no resistors there.
« Last Edit: February 14, 2020, 09:49:38 pm by ch_scr »
Offline Andreas
« Reply #1094 on: February 14, 2020, 10:06:15 pm »
Hello,
I am using some MAX4051A (with Kelvin sensing) in my LM399 ageing box:
https://www.eevblog.com/forum/metrology/measurements-on-emf-error-of-switches/msg1254204/#msg1254204
edit: the MAX4051(A) has the advantage that I do not need any level shifters to the 5V logic which would be needed with CD4051/CD4067.
with best regards
Andreas
« Last Edit: February 14, 2020, 10:22:59 pm by Andreas »
Offline notfaded1
« Reply #1095 on: February 15, 2020, 02:04:29 am »
I seem to remember seeing this somewhere on here where someone had 5 LTZ1000 references that were averaged (must be nice).  Is there a multiple LTZ1000 (or even two) reference design pcb on eevblog?
« Last Edit: February 15, 2020, 02:06:38 am by notfaded1 »
.ılılı..ılılı.
notfaded1
Offline Kleinstein
« Reply #1096 on: February 15, 2020, 10:10:11 am »
TiN has a board with 4 x LTZ1000. The circuit is pretty straight forward: separate references and than average with resistors and have a buffer / amplifier after that.  So there is limited need for a special board - just use 2 identical boards and a single buffer / amplifier  (e.g. have one board partially populated).
For the LM399 references one could use a single source for the driving current, so only one amplifier to 10 V needed, that can supply the current to all refs.
Even with extra emitter resistors, the current mirror type current source will likely no be very accurate or stable. A resistor from a higher voltage (even if just the 15 V from an 7815) is  more stable.
The following users thanked this post: notfaded1
Offline niner_007
« Reply #1097 on: May 08, 2020, 07:28:55 am »
Made a few updates to my LM399 design, not tried this version yet
The following users thanked this post: Marco1971
Offline niner_007
« Reply #1098 on: May 17, 2020, 04:57:41 am »
That was terrible :) I forgot everything about LM399, much better now, 3 VCC planes, and 3 GND planes
I'm using the connector that the 3458A reference is using (Molex 22172052), the board size is 50mm x 26mm. I'll be using VPG VHD200 for the gain stage.
« Last Edit: May 17, 2020, 05:11:23 am by niner_007 »
Offline KK6IL
« Reply #1099 on: May 24, 2020, 11:36:40 pm »
R5, 200K looks like too high a value. the LM399 operating current is 0.5-10 mA, so with a 24V supply, R5 should be between 34K  and 1.73K.
The schematic does not show the connection between Z- and GND.