Author Topic: Microphone calibration- GR 1559-B  (Read 3864 times)
Offline Conrad HoffmanTopic starter
    • The Messy Basement
« on: January 14, 2019, 02:17:44 am »
It's metrology, so I'll put it here. I've wanted one of these for a while, both to calibrate microphones and to better understand reciprocity calibration. It's pretty well tuned up, save for one dial vs. potentiometer setting that's a bit difficult. Fortunately that mode isn't essential to operation. The 1965 manual that's available online has no calibration information. I'm hoping somebody might have a later manual that they could scan, if it has the info. I contacted IET, who took over the GR line long ago, and they said in the 20 years they've servicing GR equipment, they've never once seen a 1559!
I did a page on the instrument here- http://www.conradhoffman.com/gr01.htm
The 1559 is a few items down, under the hummer.
Thanks,
Conrad
The following users thanked this post: ManateeMafia, Vgkid, kj7e, Svgeesus
Offline GLouie
Re: Microphone calibration- GR 1559-B
« Reply #1 on: January 14, 2019, 06:12:11 am »
I don't know about the unit specifically, but the red "S" on the mic looks like the old Shure logo. You might try contacting Shure. They have a rather good archive, employing at least 2 archivists. They might have info on supplying GR. Unless someone substituted the mic on your unit.
Shure archives article:
http://blog.shure.com/mysteries-and-treasures-in-the-shure-archives/
« Reply #2 on: January 14, 2019, 01:36:47 pm »
Interesting idea. The mic is original, as the older GR catalogs show exactly that same logo on their mics. The mic is buried in the back of the PZT cavity, so the unit would likely become a doorstop before anybody went that deep to fix it.
Offline SilverSolder
« Reply #3 on: January 14, 2019, 06:59:35 pm »
Interesting test gear for sure!
By coincidence, I bought my first Genrad item the other day - a 1454-AH resistance decade divider.   After cleaning it up, a quick test drive showed that it still comfortably met its specification...  not bad, after 50 - 60 years!  Got to hand it to whoever made those resistors.
And stable! - that was the most surprising thing.  Rock solid output.  I guess when a divider is rated for 700 volts, it isn't even going to notice a modern 10V reference!
Seems to me that Genrad puts a new meaning to the term "over-engineered"...   I have never seen rotary switches this big, coated with enough silver to pay off the national debt (just about)!
Is the GR 1559-B similarly over-engineered inside?
Offline GLouie
« Reply #4 on: January 14, 2019, 07:37:11 pm »
A place like Shure might know all about the device, especially if their mic was in it.
« Reply #5 on: January 14, 2019, 08:20:07 pm »
@SilverSolder- you should look at my GR pages for each instrument if you like GR stuff. The 1559 is certainly built to the GR standard, though there's no reason for big low resistance switches. They used nice ceramic wafers instead. Everything else is rock solid. You should see the insides of their 1608 bridge- a mechanical wonder. Be sure to read the GR history on the IET Labs site. They modeled themselves after Leitz in Germany and over-engineered the heck out of everything. They were also way ahead of their time in terms of employee treatment and benefits. It paid off until it didn't, like so many other classic companies.
edit- I found someone to measure the resistance of their pot at several dial settings, so I'm all "set" now.
« Last Edit: January 15, 2019, 03:24:28 am by Conrad Hoffman »
Offline texaspyro
« Reply #6 on: January 15, 2019, 04:50:43 am »
I have a couple of GR-1567 sound level meter calibrators.   They generate a precise tone (1000 Hz, 90dB?) tone.  They are a cylindrical shape with a metal "cup" at the top where you seat the sond level meter microphone.
Caution,  the cup at the top is removable.  It is held in by three ball bearing detents under a rubber o-ring.  That o-ring can degrade over time and break...  then you get to play "find the frickin' balls".
Offline VintageNut
« Reply #7 on: January 15, 2019, 12:05:13 pm »
Hello Conrad
Interesting acquisition. Great job describing the operation of the instrument on your website.
I own a calibrated capsule and some associated preamp gear.
I had the capsule calibrated at PCB Piezoelectronics. They use something like your acquired instrument but my guess is that it is more up-to-date and more automated. They may be a resource if you find a way to strike up a conversation.
The capsules are not super expensive to acquire used. Many times the capsules are 40+ years old but work perfectly if they were treated gently.
A calibration is also not super expensive.
working instruments :Keithley 260,261,2750,7708, 2000 (calibrated), 2015, 236, 237, 238, 147, 220,  Rigol DG1032  PAR Model 128 Lock-In amplifier, Fluke 332A, Gen Res 4107 KVD, 4107D KVD, Fluke 731B X2 (calibrated), Fluke 5450A (calibrated)
« Reply #8 on: January 15, 2019, 01:59:13 pm »
Yes, I know PCB and they're not far away from me. I talk to them now and then, usually about accelerometers for work, though I also have a collection in the lab at home. They seem to have gotten into acoustics in a big way lately, so they're definitely worth a call. Speaking of accelerometers, I also have the GR accelerometer calibrator. Also the tone calibrator.
Offline VintageNut
« Reply #9 on: January 15, 2019, 09:21:04 pm »
An accelerometer calibrator sounds very cool. I have a small collection of accelerometers used for attaching to a loudspeaker to look at linearity of travel versus current input.
« Reply #10 on: January 21, 2019, 03:30:20 pm »
I've updated the page and included an Excel spreadsheet with some useful information. Enjoy.
CH
« Reply #11 on: January 27, 2019, 05:02:59 pm »
Emailed with Shure and, surprise, they made the mics for General Radio. I don't know for how long or when and if GR might have brought production in-house, but the GR mics are obviously a special version of the Shure 98A108A, that you can find a datasheet for online. Updated my GR 1559-B page with more into on this, photos of the adapter and other stuff. Enjoy.
Offline GLouie
« Reply #12 on: January 28, 2019, 07:14:40 pm »
Nice! It's shame detailed historical information gets lost so often.
(I think there's a typo on Michael Pettersen's name)
« Reply #13 on: January 29, 2019, 01:01:11 am »
Ooops, fix him! I also decided that I don't understand the calibration of the meter yet, so deleted that stuff from the spreadsheet. Calibration is certainly a piece of historical info that would be nice to have.
edit- 1/29 I keep uploading new versions of the spreadsheet and I think the calibration section is close now.
« Last Edit: January 29, 2019, 01:13:57 pm by Conrad Hoffman »