It is about 1.5mV too high when i measure 10V (compared to my Keithley 2700 and HP3457A). 10kohm is about 80 ohm too high.
Congrats for 1st impression on this purchase!
These big deviations, providing:
your DMMs themselves are not too far off
you've properly warmed up and made an ACAL All
used 4W Ohm and OFFSET COMPSENSATION
are 150ppm for 10V and 8000ppm for 10kOhm, much too big for the usual drift.
This indicates, that the cal constants are both lost, that would fit to the age of the nvRAMs.
You can read the 40kohm and 7V reference values by CAL? 1,1, CAL? 2,1.
Maybe these are set to their initial values of 40,000k and 7,2000V.
If you used 2W Ohm only, these +80 Ohm might indicate the usual contamination of the front/rear switch, which you can check by engaging multiple times, and by using the back jacks.
Fingers crossed, that you got a stable one (U180).
Frank