PDVS2mini v2 Tempco measurementI measured the temperature coefficient of my PDVS2mini v2. Already when measuring warm-up behaviour, it looked like that temperature had most influence on the PDVS2mini v2 output voltage. So i wanted to take a closer look.
I put the PDVS2mini in a cooky tin. This cooky tin has wooden bars glued to bottom and sides. So the is content kept on a distance from bottom and walls.
Also in the cookie tin is a LM35 temperature sensor for measuring the ambient temperature.
IMG_4527_PSDVS2mini_v2_tempco_setup-2000pix.jpgThe cookie tin is placed in a cool box, with a 12V Peltier element and fan. It is placed in a way that the air can flow around it, but it is shielded from direct air flow from the fan. The cabling was also shielded / isolated from the air flow, to minimise thermal voltages.
By reversing the power to the Peltier, one can also heat the contents of the cool box.
Full size graph:
PDVS2mini-v2-tempco-26-09-2025-1200pix.pngBlue is the ambient temperature. After some stabilizing time, the temperature goes down to ≈ 14°C. The temperature dwells at ≈ 14°C, to let the PDVS2mini internal temperature (and output) catch up. Then the temperature rises slowly to ≈ 31°C. Unfortunately, by then the battery was almost empty... So I let the temperature dwell at this temp, to get nice stable internal temp and output voltage values.
Green is the internal PDVS2mini temperature, tracking the ambient with a some delay.
Red is the resulting output voltage of the PDVS2mini v2. One vertical division is 10uV = 1ppm.
Battery intermezzoThe grey curve is the PDVS2mini v2 battery voltage. Unintended this measurement was also a battery test. The measurement started with fully charged EBL 6F22 batteries, and after 8:50 hours they were empty. Earlier then promised / expected.
So I ran capacity tests on 2 EBL 6F22 batteries, and got ≈ 450mAh. I posted test & results in the
EBL 9V battery topic.
Back to tempcoEven though I could not complete my temperature sweep because of empty batteries, there is enough info to calculate tempco.
Using the boxed method I calculate a tempco of -0,77ppm/°C related to the outside temperature.
Related to the PDVS2mini internal temperature, I got a tempco of -0,89ppm/°C
This is the same measurement, transferred to Excel:
Full size graph:
PDVS2mini-v2-tempco-26-09-2025-Excel-1500pix.pngHere I have added two traces, showing the PDVS2mini output voltage with the tempco calculated out of it. For each trace, I nudged the tempco figure to get as flat lines as possible.
The added light-blue trace is the PDVS2mini v2 output, with the ambient tempco calculated out of it. I found an optimum using -0,75 ppm/°C tempco.
The reason that this line is not flat is the delay between a change in ambient temperature, and the resulting output change.
The added light-green line is the PDVS2mini v2 output, with the tempco referred to the PDVS2mini internal temperature calculated out of it. This line is much flatter, because of the good coupling between internal temp and output voltage. With this correction of -0,9ppm/°C the PVDS2 mini v2 output would only deviate ±1ppm over the 15°C ….31°C temperature range!
ConclusionsThe Tempco figures found with the boxed method correlate well with the figures found in the Excel graph.
My findings in the warm-up measurement were correct: my PDVS2mini v2 output is more sensitive to temperature then expected.
The bad news is that the tempco I found of -0,75 ppm/°C is much worse than the Wrytech spec of “Typically < 2 ppm across a 10K temperature change”.
The good news is that the light-green trace in the Excel graph shows that it is quite possible to calculate this tempco out in the PDVS2 mini v2 software, and removing it from the output. The temp sensor is already there.
This could be done in the next PDVS2 mini v2 firmware.....
I will run a second tempco measurement to verify my findings.
Regards, Gertjan.