Author Topic: PX Reference  (Read 52395 times)
Offline ap
    • ab-precision
Re: PX Reference
« Reply #125 on: March 09, 2018, 03:59:35 pm »
After burn in a couple of weeks, i need a calibration of one or some of them.
(as free as possible, except shipping costs of course.  ::) )
Who can I address then, medium term, if possible in Germany or EU?
I know, maybe a difficult question...
Calibrating your LTZ reference after stabilization burn-in against a 3458A does not make much sense. A standard 3458A has a 1 year uncertainty of 8+ ppm, so effectively arround 10ppm. Your LTZ, if properly build and stabilized, should be arround 1 or 2 ppm pa. So if you e.g. do two consecutive CALs, 1 year appart, against calibrated 3458As in their one-year accuracy cycle, to evaluate the stability of your LTZ, you will know nothing. Any drift of your LTZ will be hidden in the 3458As' uncertainties. It deserves a much tighter cal. If you are ready with burn-in feel free to send me a PM.
« Last Edit: March 09, 2018, 04:26:38 pm by ap »
Metrology and test gear and other stuff: www.ab-precision.com
The following users thanked this post: cellularmitosis, hwj-d
Offline TiN
  • Country: ua
    • xDevs.com
Re: PX Reference
« Reply #126 on: March 09, 2018, 04:09:29 pm »
I still hold KX project calibrations for willing participants, and to support community I can offer one time PX ref calibration (DCV value and tempco as received) too, at shipping cost. Calibration vs unknown history 3458A indeed does not provide uncertainty better than 10ppm, but verified known stable 3458A can do 4ppm without much problem. Both of my 3458A drifted <2ppm from last year, however I'd stress out this is NOT an official or legally traceable calibration, and my meters are not standard stock boxes.
I do my verifications for 10V range in my homelab using bank of 6 x LTZ1000 references and 4 x LTZ1000A-based DMMs versus annually calibrated 732B (unc 0.55ppm). So my safe lab guesstimate is 4ppm/annual, 2 ppm if you hit a lucky day.  :) . Good verified stable 3458A is great to do DCV-DCV transfers over a day or even week with sub-ppm uncertainty (I have one verified vs 732B).
Offline cellularmitosisTopic starter
Re: PX Reference
« Reply #127 on: March 09, 2018, 04:57:43 pm »
Very generous TiN!  Time to assemble my KX board!
LTZs: KX FX MX CX PX Frank A9 QX
Offline hwj-d
  • save the children - chase the cabal
Re: PX Reference
« Reply #128 on: March 09, 2018, 05:05:46 pm »
@ap, @TiN, @Echo88, @cellular and all the others here and over the world,
what can i say now,
here with you, i feel in safe hands.  :)
After burn in, i will PM you.
Maybe one goes to ap, one to TiN, if i know, whether there are import export customs problems or not.
Thanks so much to all
« Last Edit: March 09, 2018, 05:11:34 pm by hwj-d »
Offline eurofox
    • Music
Re: PX Reference
« Reply #129 on: March 13, 2018, 05:54:56 pm »
Hi PX builders,
I would like to get your input, what do you think about my measures taken on a Keithley 7 1/2 digit, 100 averages:
Max: 07.065917 V
Min: 07.065843 V
For me it does not look very stable  :palm:
eurofox
There's No Future In the Past.
eurofox
Offline hwj-d
Re: PX Reference
« Reply #130 on: March 13, 2018, 07:51:46 pm »
Hi PX builders,
Max: 07.065917 V
Min: 07.065843 V
eurofox
If you switch it on first time, the voltage drops. That's normal. Let it stabilize a couple of hours, days, weeks without turning it off. Don't connect chassis/PSU ground to measurement -Vz. Use a noiseless PSU. Watch the used current, that should be < 30mA (26-27mA).
€: Measure without greater ambient temperature fluctuations for the first time.
€€: Don't use clips, crokodiles etc. with measurement cables. Twist them.
« Last Edit: March 13, 2018, 08:32:12 pm by hwj-d »
Offline eurofox
    • Music
Re: PX Reference
« Reply #131 on: March 13, 2018, 10:48:12 pm »
Hi PX builders,
Max: 07.065917 V
Min: 07.065843 V
eurofox
It is on my lab desk not connected to ground, powered by an HP lineair power supply and use less than 30mA.
It is only powered for a few hours, have to build it into a "container" with a dedicated power supply.
I would like to add a opamp to raise the voltage to 10V.
eurofox
eurofox
Re: PX Reference
« Reply #132 on: March 13, 2018, 11:07:24 pm »
Hi PX builders,
Max: 07.065917 V
Min: 07.065843 V
eurofox
Hi eurofox,
I have found that my LTZ's have been more sensitive to EMI / waving my hand near the circuit, etc, than my LM399 circuits.
The most quiet results I have obtained so far were:
- With the LTZ board inside of an aluminum Hammond 1590B enclosure
- Measuring the difference between one LTZ and another
- With both LTZ's powered by independent batteries / regulators
I have other boards on-hand (a clone of Dr. Frank's board, and a KX board), but I have not assembled them yet (still characterizing resistors and I'd like to measure the 1/f noise of my LTZ's before soldering them in).  When I get those other boards assembled, I'll be able to compare the EMI / hand-waving sensitivity of the different board designs.
Offline d-smes
Re: PX Reference
« Reply #133 on: March 14, 2018, 11:50:22 am »
Why don't more folks pour a ground fill in their circuit board layouts and stitch them top to bottom with via holes to improve localized shielding against EMI?
Offline Dr. Frank
  • Got a lot of time now - and very low uncertainty
Re: PX Reference
« Reply #134 on: March 14, 2018, 12:29:44 pm »
Why don't folks use more of Andreas RF/EMC blocking capacitors?
Single sided PCB is also fully sufficient, but a Guard / shield enclosure required.
My 'new' circuits / assemblies are completely immune against common hand-waving, SMPSU, or mains disturbances.
Frank
Re: PX Reference
« Reply #135 on: March 14, 2018, 02:16:58 pm »
You mean like this?
The following users thanked this post: d-smes
Offline hwj-d
Re: PX Reference
« Reply #136 on: March 14, 2018, 03:12:51 pm »
That's strange. I've build four PX'es now, and none has this EMI sensitiveness. See attached pictures. Nearby i'm designing/building a little compact 2x15V PSU. That's worstcase  ;) for EMI, normaly. Simultaneously measurement with 10 PLC; 10 µV Y-Achsis.
« Last Edit: March 14, 2018, 03:23:42 pm by hwj-d »
Offline hwj-d
Re: PX Reference
« Reply #137 on: March 14, 2018, 03:28:18 pm »
You mean like this?
Does it go inside a TEKO 371 ;)
Re: PX Reference
« Reply #138 on: March 14, 2018, 03:42:43 pm »
You mean like this?
Nope, a 1590B  ;D
I should put a disclaimer: I haven't actually made nor tried that board yet, and it is also missing one of Andreas' capacitors (his C15, I think), so you'd have to bodge that in.
Offline hwj-d
Re: PX Reference
« Reply #139 on: March 14, 2018, 04:05:08 pm »
You mean like this?
Nope, a 1590B  ;D
Don't be fooled. ;)
Your PX V2.4 is very good, in my opinion.
These EMI issues must be due to external circumstances.  :-+
Offline Kleinstein
Re: PX Reference
« Reply #140 on: March 14, 2018, 04:37:51 pm »
Ground fill is often a poor solution for precision circuits, as one looses the control over the ground connection. A ground plane can be low impedance, but not low enough when it comes to ppm range.
One might use an extra plane just for shielding, e.g. add an extra shield layer to an otherwise single sided layout.
Close to the LTZ1000 one may not want the extra thermal conduction everywhere, so it would end up with a circle around the LTZ without an shield.  One downside with partial shields is that they tend to concentrate interfering fields near the edge. So a partial shield might be worse than non at all.
Offline quarks
Re: PX Reference
« Reply #141 on: March 15, 2018, 01:40:10 pm »
Thanks
because I caught a bad cold I am a little late, but feel free to contact me, if you need another meassurement of your LTZ1000A.
BTW very nice build, so thanks for sharing :-+
The following users thanked this post: splin, hwj-d
Offline hwj-d
Re: PX Reference
« Reply #142 on: March 15, 2018, 04:24:27 pm »
Thank you very much for your offer, will come back to that after burn in a few weeks more.  :)
Get well.  :-+
Re: PX Reference
« Reply #143 on: April 29, 2018, 11:01:52 pm »
Version 2.4.1
Just a few trivial changes:
I found the TO-99 too difficult to solder, so I made the pads a bit larger.
I also added footprints for an input and output capacitor directly on-board (I had been mounting these caps across the binding posts).
https://oshpark.com/shared_projects/hUZKe65r
https://github.com/pepaslabs/px-ref/tree/master/kicad/releases/v2.4.1
Offline WillTurner
Re: PX Reference
« Reply #144 on: June 09, 2019, 06:36:33 am »
@Cellular
  I wasn't quite sure where to put this, but since it's a derivative of your LTZ1000 symbol, I thought you'd most likely find it here.
  Below is a Kicad LTZ1000 schematic symbol broken into heater and zener elements. I'm planning on moving the schematic back to being closer to the reference circuit.
  Here is the heater.
  And here is the rest of the component.
Edit: inserted images inline.
« Last Edit: June 09, 2019, 06:41:04 am by WillTurner »
Offline aronake
Re: PX Reference
« Reply #145 on: May 11, 2022, 02:27:33 pm »
For version 2.5 and 2.6 what connector is intended to be used for the two big areas of the PCB?
The following users thanked this post: bck
Re: PX Reference
« Reply #146 on: September 06, 2026, 01:55:39 pm »
ah, 2.5 and 2.6 were "beta" boards which happened at about the same time that my volt-nut disease went into remission.  They were both designed to fit in the next larger size of TEKO case.
The two big areas on the PCB are for the large tubular ferrite chokes as used in the cal club's px ref, but you can attach the chokes to the board instead of just stuffing them into the case.
If I recall, 2.5 was simply the 2.4 series but with the chokes on-board, and also a couple of footprints for larger 1uF red box wima caps on input and output.  I don't think I ever got around to populating one.
2.6 was an experimental board I made after getting my hands on some OHMTEK 132077 hermetic multi-resistor chips, which I probably found on taobao, and thought "I should build an LTZ around this".  I seem to recall it didn't work correctly, but I never got around to debugging it.
I've attached a photo from 2018.
edit: recalling a bit more of this now: the jumpers on 2.6 (at least in theory) allowed you to make many combinations of R4 and R5, thus dialing in different temperature set points.  I wrote a little script back then to spit out all of the combinations, which I just threw into a gist.
I doubt anyone could get their hands on any OHMTEK 132077 chips today, so that's a bit of a dead end.
« Last Edit: September 06, 2026, 02:09:44 pm by cellularmitosis »
The following users thanked this post: Andreas, doktor pyta, exe, MiDi, Grandchuck
Offline Grandchuck
Re: PX Reference
« Reply #147 on: September 06, 2026, 02:03:21 pm »
hey CM...nice to see you posting again! :)
Offline Conrad Hoffman
    • The Messy Basement
Re: PX Reference
« Reply #148 on: September 11, 2026, 03:27:32 am »
CM, your projects make me feel better about mine. My interests fade in and out over several year cycles. Metrology for a while, photography for a while, health issues for a while, family issues for a while, 3d printing for a while, then around again.  :D
Re: PX Reference
« Reply #149 on: September 11, 2026, 11:32:44 pm »
Rock on, Conrad!  You've certainly been an inspiration to me.