Summary before teardown
- The years of release and stop are estimate. Correct me if anyone know the accurate.
- Oven: small dot for chip level, large for board level
- There are evidents indicating that the voltage of most LTZ1000 and LTFLU-1 go down with time while SZA263 go up.
- The order of the teardown is roughly chronological
- The order on the list is according to the annual drift(and the time of release of same)
- The teardown will focus to the most important part: the 7V reference and the 7V to 10V conversion.
- Of course I don't have the teardowns of all listed(the list is not complete on its own). If anyone has, contribute.
- There is no release of new voltage standard for sometimes now, until recently by ADCMT, 'The highest performance 6900 in the world':
http://www.adcmt.com/news/press/press-2015/press20150624.htmlFluke 731AOne of the basic circuit.
If delete R3 and R4, connect neg input of the opamp to base of the refamp, then we get the simplest refamp circuit.
Fluke 731BThere are three major modifications compare to 731A
- output current limiting circuitry added
- use hermetic WW. Those WW became widely used by Fluke later
- use SZA263 as refamp. This refamp became widely used by Fluke later
Fluke 732A(See attachment)
There are some modifications compare to 731B
- board ovenized.
- use lead-acid batteries instead of NiCad.
- small-sized hermetic WW were used
- refamp protected by Q5(for later version of LTFLU-1, the protection is build-in)
- formalize the output as 'series pass'
- use 'binary' scheme on major part of the 10V adjustment, but not all.
The basic circuitry is not changed even for the designator