Hello,
I followed the SNS_LO line, there is no broken line. CR14 and CR15 are in good.
Output values up to AC 33V are good, but when I go above 33V, the range changes and turns off the output.
Output values up to DC 33V are fine. an upper range also gives the value, but it is not stable. When I increase the voltage a little more, it turns off the output.
Do you have a suggestion about this issue?
If output values up to 33V AC/DC are correct then at least the main output amplifier on the A8 Voltage board appears to be working as expected, but isolating the fault to a specific section is going to be a bit tricky.
For voltages above 33V AC/DC the calibrator makes use of various step-up transformers which are located on the A8 assembly, the specific transformer depends on a few parameters.
The big transformer (T3) is used for low frequency ACV (<= 2 kHz if I remember correctly), the smaller potted transformer (T2) is used for high voltage DCV and higher frequency ACV.
The primaries of these transformers are driven by the same amplifier used for voltage in the 33V range, also located on the A8 assembly.
Trying different AC voltage/frequency combinations might aid in isolating the fault, as well as checking all components on the secondaries of the transformers on this board.
I've also had some issues with zener diode VR5 failing on this board, this diode is located in the top left of the board.
Certain parts on the A6 assembly are used differently depending on the output range selected, the Fluke 57/LFC service manual has a very detailed theory of operation. This instrument is a bit newer, but a lot of the details of the A6 assembly are similar.
You mentioned that the output switches off when going above 33V AC, does it throw an error when this happens? And if so, what error?