I've just checked the drift of a 33 year old SR1010-10k box. 1993 factory deviations ranged from +10 to +27ppm (ie. a span of 17ppm) for the twelve 10k resistors. Whilst I can't confirm deviations from nominal, I was able to measure a deviation span of about 30ppm, with eight of the twelve resistors showing similar relative deviations, but four had noticeably drifted down in relative value.
The SR1010-1k in this post (
https://www.eevblog.com/forum/testgear/test-equipment-anonymous-(tea)-group-therapy-thread/msg4107907/#msg4107907 ) had been re-calibrated after 9 years, and showed worst-case drifts of +12ppm and -13ppm.
The ESI technique of using shorting bars, and series/parallel, and series connections, for 0.1x, 1x, and 10x transfer ratios is somewhat moot to lowish deviations between individual steps in a box. Choosing a sequence of 9 steps for the 1x transfer, and 10 steps for the 0.1x and 10x transfers may need some selection of which end steps to avoid if they have noticeably drifted away from the pack.
In some of the teardown photos of ESI resistors, and in the photo below of my box, there appear to be 'scuff' patches on some sides. Nothing consistent, and perhaps just blemishes of the enamel coating, but perhaps not. I suppose its not implausible that they are related to a factory selection or testing or fitting procedure.