After observing two new LTFLU references for some days, i got a question to the experts.
For the direct comparison of the two voltages i made a 100x preamplifier with a ADA4522 ("null detector"). I am using two 470 Ohm resistors for the inputs and a 47 KOhm resistor for negative feedback, a very basic inverting amplifier circuit. I am measuring the amplified difference voltage using a HP 3456A in its lowest range. Then it has a resolution of 100 nV, so with the preamplifier the resolution is 1 nV. The ADA4522 is specified with 117 nV noise peak-to-peak into 0,1 to 10 Hz. Using the preamplifier i am trying to reduce the difficulties of bringing sub uV voltages to the DVM without thermal EMF.
What would be a realistic expection for the noise in 100 PLC measurements that i am doing every 10 seconds? One measurement takes about 2 seconds, so the Nyquist limit would be 0,25 Hz and i would expect a small fraction of 117 nV, like 20 nVpp, about 2 ppb of a 10 V standard. Plus the noise of the two references.
I am asking this because today i got a sequence of about 300 measurements that exhibited less than 3 ppb noise after subtraction a the thermal drift. Is the reference noise negligible in such a setup?
Maybe i should first measure the preamplfier+DVM noise with a short-circuit on the amplfier input.
Regards, Dieter