Thanks both for your advices, but it's getting even better.
I guess I made an essential, ground breaking discovery and improvement on probably all existing long scale DMMs.
These all have an application flaw, but nobody before has noticed that, neither LT, nor any DMM design engineer.
This effect might only affect the LTZ based DMMs, as this following root cause might destabilize the oven control circuit only.
When making direct comparisons with this absolute method by means of an DMM, I always wondered, why I never could achieve the very same noise figures as the differential method, although both methods are essentially identical: Any DMM directly measures the difference between two (LTZ1000/LTFLU/LM399) references, and the A/D only serves as the differential instrument, contributing no additional noise.

I always observed 160nV @ 10min and 200nV @ 1h time constant.
Please observe these strange humps, very prominent after 33600sec, and being always present, unavoidable.
Please check this on your own stability measurements with your specific 8.5 digit DMMs.

I always assigned these to my specific LTZ1000A on A9, assuming that it suffers from sort of 'random walk desease', maybe being a different effect apart from the usual 'popcorn noise' of the LM399.
After Andreas ingenious invention, I could demonstrate, that his circuit has the effect of effectively suppressing any external ever so slight e.m.c. or similar disturbances.
These reference amplifiers are extremely sensitive, as it's commonly known.
The comparison with the improved A9 board now shows exactly the same noise behavior as the differential method, with the exact same noise levels for 10 min, which is a -33% noise improvement.
The low 1min noise can be assigned to the doubling of the zener current, giving another -33% improvement.
I estimated 50% in total, before.

This reduces my Type B measurement uncertainty greatly, making long term monitoring stable to the physical limit.
I as well always wondered, why all my transfer measurements of 15 references showed 0.2ppm wide fluctuations from month to month, although each sample can be made with 0.02ppm StD, and all references are as well stable to this noise level.

I bet, that this effect as well causes the instability in DATRONs 1271 reference circuitry. This part of the DATRON design is extremely complicated and over-engineered. John R. Pickering generally comfirmed to me over-engineering of his former DATRON engineers,which was the decisive competition disadvantage over the HP3458A development and although the 1271 is really much better designed.
Their reference sensing method, which is basically a very clever idea, creates an instability source with small oscillations directly impacting the LTZ1000, which would normally not create voltage reference instabilities. Btw, DATRON already used Andreas' low pass feedback 10k/22n on the amplifier.
As this sensitivity characteristics is inherent to the 'Reference Amplifier', probably all DMMs have inherited this problem, but this can easily be fixed.
Frank
Andreas, you Wizard of LTZ, now I'm really happy to meet you for our small MM on Saturday!
PS: for obvious reasons I'm just calibrating my HP3458A.
It came to my mind, that this is the first time ever, that it's now possible to make a proper calibration, w/o an error directly from the factory, however small this might has been