Author Topic: USA Cal Club Round 3.1  (Read 20653 times)
Offline CalibrationGuy
Re: USA Cal Club Round 3.1
« Reply #25 on: April 06, 2025, 03:51:18 am »
First Voltage Reference Bank FX assigned value: +9.999956709. Less than 24 hours powered up.
Setup conditions: +/-12Vdc by 4x6Vdc AGM SLA batteries for lowest noise. Actual voltage was 12.8Vdc at start.
70.49 degrees F, 31.49% RH
Hope this is useful to the group. Batteries should last 4-5 days, so I'll update with daily readings and send it back by week's end.
Voltage Bank: 4x732A, 4x732b, 1x732c (characterized at Keysight's Standards Calibration Laboratory) with 34420a using NIST measurement methodology running custom data acquisition and mathematical regression script.
PS - Kudos to my friend Ilya at xDevs, truly remarkable performance!!!
TomG.
« Last Edit: April 06, 2025, 04:07:13 am by CalibrationGuy »
The following users thanked this post: Rax, Grandchuck, bdunham7
« Reply #26 on: April 07, 2025, 04:48:05 am »
Ok, the day 2 number is: +9.999955289Vdc. Temp and humidity within +/-1 of yesterday's values. Next update will be tomorrow evening.
TomG.
The following users thanked this post: Grandchuck
« Reply #27 on: April 07, 2025, 05:04:09 am »
I guess I should also point out that the difference between these numbers is nearing the noise threshold of all the devices involved.
TomG.
« Reply #28 on: April 08, 2025, 03:10:36 am »
OK, day 3 last reading: +9.999955144, day's average: +9.999954989.
TomG.
The following users thanked this post: bdunham7
Offline bdunham7Topic starter
« Reply #29 on: April 08, 2025, 03:14:25 am »
That's a lot of digits!  Is the final result going to be 9.999955 +/- 2µV?  That puts it at -0.9ppm from the previous calibration over almost 18 months and dozens of cross-country trips via USPS.
One question about this--what is your educated opinion on how close your 10V is to xDevs 10V?
« Last Edit: April 08, 2025, 03:26:06 am by bdunham7 »
« Reply #30 on: April 08, 2025, 03:23:41 am »
The digits result from averaging samples over the hours, so we keep the digits for the math. Yes, the value looks like it's going to be +9.999955, well within +/-2µV.
TomG.
« Reply #31 on: April 08, 2025, 03:20:57 pm »
I just noticed your question. xDevs' and our 10V values are extremely close, probably +/-1ppm or less. The master 732c in our lab was compared to Keysight's primary JVS less than a year ago, so our confidence level is very high.
TomG.
The following users thanked this post: Conrad Hoffman, Rax
« Reply #32 on: April 09, 2025, 03:13:38 am »
Ok, day 4 reading: +9.999955408. That's the average. Variations pretty small today.
TomG.
« Reply #33 on: April 09, 2025, 07:29:12 pm »
You guys wanted to know how close I was to Ilia's value. Now you're going to have your answer because I dropped the FX reference off with him and he will characterize it. And then we'll see the difference.
TomG.
« Reply #34 on: April 10, 2025, 04:40:57 am »
So Ilya's first measurement of our 732b transfer standard is less than 0.18ppm away from our value. So very high level of confidence in our measurement setup as well as his. Hope this is useful information to you guys.
TomG.
Offline 1audio
« Reply #35 on: April 10, 2025, 05:25:07 am »
Thats really reassuring, even though its well below anything I could confidently measure.  I now have some confidence in my 732A measured value.
The following users thanked this post: CalibrationGuy
Online Rax
« Reply #36 on: April 11, 2025, 05:02:38 pm »
Second 1audio's point. Many thanks, TomG, all this data is in spot on agreement with everything I've been able to assess across my bench standards (to a scary level of agreement, candidly).
« Reply #37 on: April 11, 2025, 09:47:55 pm »
A request was made for final value and measurement uncertainty for this calibration/characterization. Here it is:
+9.9999554Vdc, uncertainty for this measurement is 0.2ppm.
TomG.
The following users thanked this post: Rax, bdunham7
Offline Zondar
« Reply #38 on: July 06, 2025, 10:50:54 pm »
Hello Forum Members,
I have an Agilent 3458A which has lost its calibration data. CatalinaWOW encouraged me to join the "club" here in the hope that an excellent voltage reference, etc., might come to its aid. So please sign me up or let me know how I can help. I'm in So-Cal if that matters.
Thank you very much!
« Last Edit: July 06, 2025, 11:25:49 pm by Zondar »
Online Rax
« Reply #39 on: July 06, 2025, 11:09:07 pm »
I have an Agilent 3428A
3458A maybe?... Otherwise 3478A?
Welcome to the club!
Offline Zondar
« Reply #40 on: July 06, 2025, 11:25:14 pm »
Ah yes, that was a typo. It's an 3458A, 8.5 digit DVM (fixed).
Thank you!  :)
« Reply #41 on: July 06, 2025, 11:32:48 pm »
Sent you a PM.
What condition is it in and has the reason for losing the cal data been addressed?  When these get old enough to have problems they typically need heroic levels of TLC to get them back into shape.
USA Cal Club Round 4 is coming in a bit, I've been a bit preoccupied and that may be a number of weeks off.  I want to get it reorganized to be as useful as possible to as many people as possible.
Offline cellularmitosis
« Reply #42 on: September 10, 2026, 09:44:40 am »
Heyo  :D
It appears my volt nut disease is flaring up again after being in remission for years.
I was curious if the px ref could make its way to my place?
(edit: bdunham7 shot me a DM, thanks!)
« Last Edit: September 11, 2026, 12:16:04 am by cellularmitosis »
LTZs: KX FX MX CX PX Frank A9 QX
Offline Conrad Hoffman
    • The Messy Basement
« Reply #43 on: September 10, 2026, 06:08:35 pm »
Yes, I've also been curious if the standards might make another round.
« Reply #44 on: September 10, 2026, 07:09:58 pm »
Hit me up if you want it characterized before sending it to the members here.
TomG.
The following users thanked this post: Conrad Hoffman, cellularmitosis
« Reply #45 on: September 10, 2026, 09:54:57 pm »
TomG.
I surely do!  Thanks so much!
Offline 1audio
« Reply #46 on: September 10, 2026, 10:05:40 pm »
I would be up for a recheck as well.