Okay, here are some news. The unit is running with all 4 reference boards. It was "just" a matter of hitting the right buttons at the right time, no independant NOVRAM content required
1. Started the instrument with the fourth board fitted with an empty NOVRAM. The unit stops during boot sequence showing firmware version.
2. Pushed the SYS RST button, unit notice bad NOVRAM content
3. with the CAL key in CAL position and pushing the hidden key (below Last Cal) the instrument is able to repair the "broken" memory content, once the secret password was entered
IMPORTANT: Whenever the unit asks "CONNECTION ?" it expects a connention between SENSE terminals and NULLMETER, otherwhise an error new to me appears that tells you "CANNOT OUTPUT". Afterwards the voltage output doesn't work anymore. Ask me how I know
It took quite some time to figure out how to get away with that error, as it is not clearly described in the manual (missing image too).
Meanwhile, I also adjusted all oven settings, which seems to have affected at least one of the references, but consequent Internal Cal should get away with it at some point once everything has stabilized again. That's at least what the manual tells you.
I repeat my question: who has the possibilty to connect the device to GPIB?
Not here, not at the moment, need to prepare something first, as all our original Agilent GPIB-USB adapters failed. I wonder why
ranadic: could you please post the OCR-version of the Vol. 1_UM.pdf?
The file is too large to post it here
-branadic-